Thanks Charles.  That makes sense, but at the expense of adding unwanted 
complexity.  As I've been moving the setpoint around this morning, I think I 
see a way to characterize what it's doing.  Maybe I can come up with a small 
correction table or formula that's good enough for my purposes.

Bob






 From: Charles Steinmetz csteinm...@yandex.com
To: Discussion of precise time and frequency measurement time-nuts@febo.com 
Sent: Wednesday, March 26, 2014 12:27 AM
Subject: Re: [time-nuts] RC TIC linearity correction?
 

Bob wrote:

I hadn't given any thought to correcting the linearity of the TIC I 
built, but my PLL plots tell me I should do it now.

You are using a resistor to charge the integrating capacitance, so it 
charges with the classic exponential curve and you get a nonlinear 
time-to-voltage conversion.  You need to charge the integrating 
capacitance with a constant current if you want a linear 
time-to-voltage function.  The current source will probably need to 
be connected to a supply that is higher than 5v, because it needs 
some headroom.

There may be secondary errors, as well, due to the leakage of the 
tri-state buffers in their hi-Z state and/or nonlinearity in the 
ADC's internal capacitors.  Often you can improve things by using 
sufficient external capacitance to swamp the ADC's internal 
capacitance and increasing the charging current.

Best regards,

Charles


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] RC TIC linearity correction?

2014-03-26 Thread Stanley



I hadn't given any thought to correcting the linearity of the TIC I
built, but my PLL plots tell me I should do it now.


One method would be to calibrate with a series of  buckets that you fill by 
sampling a random source, the more samples in a bucket the more range in 
phase for that bucket. For example 1 1 1 2 2 2 3 2 2 1 0 samples sum equal 
17 so each unit equal 17/(total range of  TIC) and the bucket with 3 samples 
would be 3 * 17/(total range of TIC).


Stanley 


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] RC TIC linearity correction?

2014-03-26 Thread Chris Albertson
On Wed, Mar 26, 2014 at 8:41 AM, Bob Stewart b...@evoria.net wrote:

 Thanks Charles.  That makes sense, but at the expense of adding unwanted
 complexity.  As I've been moving the setpoint around this morning, I think
 I see a way to characterize what it's doing.  Maybe I can come up with a
 small correction table or formula that's good enough for my purposes.


Yes a lookup table would be easy.  But how to create the table?  I've been
thinking about a way to do self calibration.   The controller purposely
runs the DAC and of course the OCXO through some range and watches the
phase.  This gives you a rough DAC vs. Phase function.   Re-running the
calibration could make up for some component aging.   It would take some
time (hours) to wait for everything to warm up and then you'd have to move
the EFV voltage slowly



-- 

Chris Albertson
Redondo Beach, California
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] RC TIC linearity correction?

2014-03-26 Thread Bob Stewart
If I were to try to do this automatically, I think I'd move the PLL set point 
in n steps from near the bottom to near the top and look at the width of the 
PPS signal at each step; perhaps using the bucket scheme that Stanley mentioned 
and using some count per bucket to decide how wide is wide.  I don't have any 
sort of phase wrapping code, though, so I have to be careful how close to a 
phase point I get.

Bob




 From: Chris Albertson albertson.ch...@gmail.com
To: Bob Stewart b...@evoria.net; Discussion of precise time and frequency 
measurement time-nuts@febo.com 
Sent: Wednesday, March 26, 2014 11:25 AM
Subject: Re: [time-nuts] RC TIC linearity correction?
 







On Wed, Mar 26, 2014 at 8:41 AM, Bob Stewart b...@evoria.net wrote:

Thanks Charles.  That makes sense, but at the expense of adding unwanted 
complexity.  As I've been moving the setpoint around this morning, I think I 
see a way to characterize what it's doing.  Maybe I can come up with a small 
correction table or formula that's good enough for my purposes.



Yes a lookup table would be easy.  But how to create the table?  I've been 
thinking about a way to do self calibration.   The controller purposely runs 
the DAC and of course the OCXO through some range and watches the phase.  This 
gives you a rough DAC vs. Phase function.   Re-running the calibration could 
make up for some component aging.   It would take some time (hours) to wait 
for everything to warm up and then you'd have to move the EFV voltage slowly





-- 

Chris Albertson
Redondo Beach, California 


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] RC TIC linearity correction?

2014-03-26 Thread Bruce Griffiths
A random source will need a lot more samples in each bucket to reduce 
the noise to an acceptable level.
To determine the relative bucket width with a 10% error requires at 
least 100 samples per bucket.

For 1% error at least 10,000 samples per bucket.

All thats really required is a sufficiently noisy oscillator that isn't 
phase locked to your clock.


Bruce

Stanley wrote:



I hadn't given any thought to correcting the linearity of the TIC I
built, but my PLL plots tell me I should do it now.


One method would be to calibrate with a series of  buckets that you 
fill by sampling a random source, the more samples in a bucket the 
more range in phase for that bucket. For example 1 1 1 2 2 2 3 2 2 1 0 
samples sum equal 17 so each unit equal 17/(total range of  TIC) and 
the bucket with 3 samples would be 3 * 17/(total range of TIC).


Stanley


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] RC TIC linearity correction?

2014-03-26 Thread Bob Camp
Hi

Having done this - it gets really boring to sit there for 10,000 seconds and 
collect data. Best to automate the process. 

In reality you want to run three groups of 10,000 samples and see how they 
relate to each other. With some approaches you can find some disturbing things 
going on.

Bob

On Mar 26, 2014, at 1:22 PM, Bruce Griffiths bruce.griffi...@xtra.co.nz wrote:

 A random source will need a lot more samples in each bucket to reduce the 
 noise to an acceptable level.
 To determine the relative bucket width with a 10% error requires at least 100 
 samples per bucket.
 For 1% error at least 10,000 samples per bucket.
 
 All thats really required is a sufficiently noisy oscillator that isn't phase 
 locked to your clock.
 
 Bruce
 
 Stanley wrote:
 
 I hadn't given any thought to correcting the linearity of the TIC I
 built, but my PLL plots tell me I should do it now.
 
 One method would be to calibrate with a series of  buckets that you fill by 
 sampling a random source, the more samples in a bucket the more range in 
 phase for that bucket. For example 1 1 1 2 2 2 3 2 2 1 0 samples sum equal 
 17 so each unit equal 17/(total range of  TIC) and the bucket with 3 samples 
 would be 3 * 17/(total range of TIC).
 
 Stanley
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] RC TIC linearity correction?

2014-03-26 Thread Bob Stewart
I finally got my PLL running today, and I think I can manage with just a small 
table of 10 values to be used as the width of my PPS jitter corrector.  That 
takes care of it for this project, but obviously not for the general case.

Bob





 From: Chris Albertson albertson.ch...@gmail.com
To: Bob Stewart b...@evoria.net; Discussion of precise time and frequency 
measurement time-nuts@febo.com 
Sent: Wednesday, March 26, 2014 11:25 AM
Subject: Re: [time-nuts] RC TIC linearity correction?
 


Yes a lookup table would be easy.  But how to create the table?  I've been 
thinking about a way to do self calibration.   The controller purposely runs 
the DAC and of course the OCXO through some range and watches the phase.  This 
gives you a rough DAC vs. Phase function.   Re-running the calibration could 
make up for some component aging.   It would take some time (hours) to wait 
for everything to warm up and then you'd have to move the EFV voltage slowly




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] RC TIC linearity correction?

2014-03-25 Thread Bob Stewart
I hadn't given any thought to correcting the linearity of the TIC I built, but 
my PLL plots tell me I should do it now.  Explanation: when I arrange things so 
that the phase point is near the top of my TIC's range, it requires a smaller 
movement than when the phase point is in the middle:  Presumably the difference 
is even greater near the bottom.  Can anyone give me a reference of some type 
for doing this?  I looked around a few weeks ago, but my google-foo wasn't up 
to the challenge.

The schematic is essentially this, except that C1 doesn't really exist.  It was 
a place-holder on my board in case the caps in the PIC weren't up to it by 
themselves - which they were.  And, not that it matters, but the level shifter 
(Q1,Q2) was also not needed.

http://www.evoria.net/AE6RV/TIC/TIC2.bmp

Bob - AE6RV
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] RC TIC linearity correction?

2014-03-25 Thread Charles Steinmetz

Bob wrote:

I hadn't given any thought to correcting the linearity of the TIC I 
built, but my PLL plots tell me I should do it now.


You are using a resistor to charge the integrating capacitance, so it 
charges with the classic exponential curve and you get a nonlinear 
time-to-voltage conversion.  You need to charge the integrating 
capacitance with a constant current if you want a linear 
time-to-voltage function.  The current source will probably need to 
be connected to a supply that is higher than 5v, because it needs 
some headroom.


There may be secondary errors, as well, due to the leakage of the 
tri-state buffers in their hi-Z state and/or nonlinearity in the 
ADC's internal capacitors.  Often you can improve things by using 
sufficient external capacitance to swamp the ADC's internal 
capacitance and increasing the charging current.


Best regards,