Re: [time-nuts] Stanford Research SR620 Measurement Bias

2009-10-04 Thread Stan
One last follow-on to this thread: In order to speed up the calibration
process by being able to see the measurement results in real time while
adjusting the two CalByte values, I connected the SR620 scope outputs and
monitored the graphical output on my oscilloscope. It turns out that the
graphical output still works while adjusting the CalBytes and makes the
setup simpler (for me, anyway) than using a terminal emulator on the SR620
serial port to make the adjustments.

Stan

Message: 1
Date: Wed, 30 Sep 2009 16:05:46 +0200
From: J?rg K?gel j.koeg...@gmail.com
Subject: Re: [time-nuts] Stanford Research SR620 Measurement Bias
To: Discussion of precise time and frequency measurement
time-nuts@febo.com
Message-ID:
390229150909300705v1808d748h5376cf05e7ae6...@mail.gmail.com
Content-Type: text/plain; charset=ISO-8859-1

Make the adjustment in the correct sequence:

1. Set the Cal jumper to Cal Enable
2. Connect the reference to Ext Ref (rear) and Input A
3. Switch the counter to Ext Ref
4. Set the CalByte 50 for the best display (this is a very fine adjustement)
5. Switch the counter to Int Ref
6. Set the CalByte 4 for the best display (this adjustement is coarse,
optimize!)
7. Set the Cal jumper to Cal Disable

After this calibration my SR620 is

with the external reference  +/- 3 counts
(9'999'999.9997x10'000'000,0003)
previous value   -210 counts
with the internal reference  +/- 6 counts
previous value   -25 counts

Best regards

J?rg K?gel



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Stanford Research SR620 Measurement Bias

2009-10-01 Thread Stan
Thanks for the tip! Your method eliminates the interaction between the
effects of the CalByte 50 and CalByte 4 adjustments in a way that makes
perfect sense. After some trial and error, I found that it's easier to get
the CalByte 50 adjustment right using the short (0.01 s) timebase setting
rather than the long (1 s). My SR620 is now a whole lot more accurate and
the bias is essentially gone.

My one complaint is that I can't see the actual counter readings while I'm
adjusting the CalByte values. I haven't looked into this yet, but are the
readings being outputted from the scope or printer outputs on the rear panel
during the cal adjustments?

Stan

Message: 1
Date: Wed, 30 Sep 2009 16:05:46 +0200
From: J?rg K?gel j.koeg...@gmail.com
Subject: Re: [time-nuts] Stanford Research SR620 Measurement Bias
To: Discussion of precise time and frequency measurement
time-nuts@febo.com
Message-ID:
390229150909300705v1808d748h5376cf05e7ae6...@mail.gmail.com
Content-Type: text/plain; charset=ISO-8859-1

Make the adjustment in the correct sequence:

1. Set the Cal jumper to Cal Enable
2. Connect the reference to Ext Ref (rear) and Input A
3. Switch the counter to Ext Ref
4. Set the CalByte 50 for the best display (this is a very fine adjustement)
5. Switch the counter to Int Ref
6. Set the CalByte 4 for the best display (this adjustement is coarse,
optimize!)
7. Set the Cal jumper to Cal Disable

After this calibration my SR620 is

with the external reference  +/- 3 counts
(9'999'999.9997x10'000'000,0003)
previous value   -210 counts
with the internal reference  +/- 6 counts
previous value   -25 counts

Best regards

J?rg K?gel


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Stanford Research SR620 Measurement Bias

2009-10-01 Thread Magnus Danielson

Stan,

Stan wrote:

Thanks for the tip! Your method eliminates the interaction between the
effects of the CalByte 50 and CalByte 4 adjustments in a way that makes
perfect sense. After some trial and error, I found that it's easier to get
the CalByte 50 adjustment right using the short (0.01 s) timebase setting
rather than the long (1 s). My SR620 is now a whole lot more accurate and
the bias is essentially gone.


Wonderfull!


My one complaint is that I can't see the actual counter readings while I'm
adjusting the CalByte values. I haven't looked into this yet, but are the
readings being outputted from the scope or printer outputs on the rear panel
during the cal adjustments?


Use the serial port. You can adjust the calibration values using the 
BYTE and WORD commands. See page 40.


read word 50:
WORD?50
write word 50 to 1234:
WORD50,1234

That way you can keep the reading on the display while using a terminal 
emulator to adjust it.


It should not be too hard to make a little program to automize this 
calibration.


Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Stanford Research SR620 Measurement Bias

2009-09-30 Thread Jürg Kögel
Make the adjustment in the correct sequence:

1. Set the Cal jumper to Cal Enable
2. Connect the reference to Ext Ref (rear) and Input A
3. Switch the counter to Ext Ref
4. Set the CalByte 50 for the best display (this is a very fine adjustement)
5. Switch the counter to Int Ref
6. Set the CalByte 4 for the best display (this adjustement is coarse,
optimize!)
7. Set the Cal jumper to Cal Disable

After this calibration my SR620 is

with the external reference  +/- 3 counts  (9'999'999.9997x10'000'000,0003)
previous value   -210 counts
with the internal reference  +/- 6 counts
previous value   -25 counts

Best regards

Jürg Kögel


2009/9/28 Magnus Danielson mag...@rubidium.dyndns.org:
 Ulrich,

 Ulrich Bangert wrote:

 Stan,

 I believe I made it with calbyte 4 as described on page 71 of the manual.

 How does that explains the shift with time-base?

 I think it is the wrong way around for a time-bias like this. Tweaking the
 time-base will maybe fix it for one of the time-bases, but not for the
 others.

 I'm digging deeper into this, I want to know how to fix it if I see it on
 mine.

 Cheers,
 Magnus

 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Stanford Research SR620 Measurement Bias

2009-09-30 Thread Magnus Danielson

Hi Jürg,

Jürg Kögel wrote:

Make the adjustment in the correct sequence:

1. Set the Cal jumper to Cal Enable
2. Connect the reference to Ext Ref (rear) and Input A
3. Switch the counter to Ext Ref
4. Set the CalByte 50 for the best display (this is a very fine adjustement)
5. Switch the counter to Int Ref
6. Set the CalByte 4 for the best display (this adjustement is coarse,
optimize!)
7. Set the Cal jumper to Cal Disable

After this calibration my SR620 is

with the external reference  +/- 3 counts  (9'999'999.9997x10'000'000,0003)
previous value   -210 counts
with the internal reference  +/- 6 counts
previous value   -25 counts


Thanks for that clarification. It seems very reasnoble. I obviously 
never bothered to do a carefull calibration of my SR620, so I haven't 
picked up the fine grain details involved with these procedures. This is 
a good reminder that I should pay attention to those details. I think I 
can say the same for most of my gear.


Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Stanford Research SR620 Measurement Bias

2009-09-28 Thread Ulrich Bangert
Stan,

I own a SR620 as well and mine behaves the same (while with a different
number of counts) so your's is surely not defective! My understanding a
few years ago when I acquired the counter was that this bias is not due to a
adjustment in the normal sense but to set a parameter in the counter's
eeprom to a correct value. Since this parameter is limited to a certain
precisison you can make some effort in minimizing this bias but you should
not expect to make it exactly zero. The manual explains how to do but you do
not find the procedure where you would have expected it.

Best regards
Ulrich Bangert

 -Ursprungliche Nachricht-
 Von: time-nuts-boun...@febo.com 
 [mailto:time-nuts-boun...@febo.com] Im Auftrag von Stan
 Gesendet: Sonntag, 27. September 2009 18:15
 An: time-nuts@febo.com
 Betreff: [time-nuts] Stanford Research SR620 Measurement Bias
 
 
 I just acquired an SR620 that looks to be in excellent 
 condition. I've noticed something in its operation and I'm 
 not sure if it represents a problem, a calibration issue, or 
 if it's just a feature.
 
  
 
 Yesterday, after letting the counter warm up for an hour, I 
 calibrated the 10 MHz timebase and did a couple of Auto-Cals 
 (just to be sure), but even after all of that, there's still 
 a bias of somewhere between 40 and 70 counts in the frequency 
 readout, independent of the gate time. For example, if I 
 measure the 5 MHz output of my 5065A, and using a sample size 
 of 1, here's what I see (these are representative 
 readings-the values are bouncing up and down by 10 or 20 
 counts and I'm visually averaging them). Increasing the 
 sample size has no effect other than to reduce the 
 variability of the measurements, but the bias is still the 
 same magnitude:
 
  
 
 1 sec Gate: 5,000,000.00054 Hz
 
 0.1 sec Gate:   5,000,000.0060 Hz
 
 0.01 sec Gate:  5,000,000.055 Hz
 
 1 Period Gate:  5,004,7__ Hz
 
  
 
 I've had several HP 5370A/B counters, but never one of these. 
 As I recall-and it's been a while-there's an adjustment in 
 the 5370A/B for the interpolators that can remove this sort 
 of bias. Is there a similar adjustment in the SR620? I've 
 been through the manual and don't see anything obvious.
 
  
 
 Thanks for any advice!
 
  
 
 Stan
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Stanford Research SR620 Measurement Bias

2009-09-28 Thread Stan
Hi Ulrich,

Thanks for the insight! I tried manually adjusting the value of CalByte 50
(as per Magnus' suggestion), but it did not make the problem go away. There
is interaction between this adjustment and the CalByte 4 adjustment for
timebase accuracy, and when all was said and done, I didn't notice any
significant difference in the offset. 

Where in the manual does it describe the procedure to minimize the bias?
It's definitely not where I would have expected it to be!

Regards,
Stan


Message: 1
Date: Mon, 28 Sep 2009 09:08:21 +0200
From: Ulrich Bangert df...@ulrich-bangert.de
Subject: Re: [time-nuts] Stanford Research SR620 Measurement Bias
To: 'Discussion of precise time and frequency measurement'
time-nuts@febo.com
Message-ID: a4d0a2a6239b4798ab28e7193a646...@athlon
Content-Type: text/plain;   charset=us-ascii

Stan,

I own a SR620 as well and mine behaves the same (while with a different
number of counts) so your's is surely not defective! My understanding a
few years ago when I acquired the counter was that this bias is not due to a
adjustment in the normal sense but to set a parameter in the counter's
eeprom to a correct value. Since this parameter is limited to a certain
precisison you can make some effort in minimizing this bias but you should
not expect to make it exactly zero. The manual explains how to do but you do
not find the procedure where you would have expected it.

Best regards
Ulrich Bangert




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Stanford Research SR620 Measurement Bias

2009-09-28 Thread Ulrich Bangert
Stan,

I believe I made it with calbyte 4 as described on page 71 of the manual.

Best regards
Ulrich

 -Ursprungliche Nachricht-
 Von: time-nuts-boun...@febo.com 
 [mailto:time-nuts-boun...@febo.com] Im Auftrag von Stan
 Gesendet: Montag, 28. September 2009 14:46
 An: time-nuts@febo.com
 Betreff: Re: [time-nuts] Stanford Research SR620 Measurement Bias
 
 
 Hi Ulrich,
 
 Thanks for the insight! I tried manually adjusting the value 
 of CalByte 50 (as per Magnus' suggestion), but it did not 
 make the problem go away. There is interaction between this 
 adjustment and the CalByte 4 adjustment for timebase 
 accuracy, and when all was said and done, I didn't notice any 
 significant difference in the offset. 
 
 Where in the manual does it describe the procedure to 
 minimize the bias? It's definitely not where I would have 
 expected it to be!
 
 Regards,
 Stan
 
 
 Message: 1
 Date: Mon, 28 Sep 2009 09:08:21 +0200
 From: Ulrich Bangert df...@ulrich-bangert.de
 Subject: Re: [time-nuts] Stanford Research SR620 Measurement Bias
 To: 'Discussion of precise time and frequency measurement'
   time-nuts@febo.com
 Message-ID: a4d0a2a6239b4798ab28e7193a646...@athlon
 Content-Type: text/plain; charset=us-ascii
 
 Stan,
 
 I own a SR620 as well and mine behaves the same (while with a 
 different number of counts) so your's is surely not 
 defective! My understanding a few years ago when I acquired 
 the counter was that this bias is not due to a adjustment 
 in the normal sense but to set a parameter in the counter's 
 eeprom to a correct value. Since this parameter is limited to 
 a certain precisison you can make some effort in minimizing 
 this bias but you should not expect to make it exactly zero. 
 The manual explains how to do but you do not find the 
 procedure where you would have expected it.
 
 Best regards
 Ulrich Bangert
 
 
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Stanford Research SR620 Measurement Bias

2009-09-28 Thread Magnus Danielson

Ulrich,

Ulrich Bangert wrote:

Stan,

I believe I made it with calbyte 4 as described on page 71 of the manual.


How does that explains the shift with time-base?

I think it is the wrong way around for a time-bias like this. Tweaking 
the time-base will maybe fix it for one of the time-bases, but not for 
the others.


I'm digging deeper into this, I want to know how to fix it if I see it 
on mine.


Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] Stanford Research SR620 Measurement Bias

2009-09-27 Thread Stan
I just acquired an SR620 that looks to be in excellent condition. I've
noticed something in its operation and I'm not sure if it represents a
problem, a calibration issue, or if it's just a feature.

 

Yesterday, after letting the counter warm up for an hour, I calibrated the
10 MHz timebase and did a couple of Auto-Cals (just to be sure), but even
after all of that, there's still a bias of somewhere between 40 and 70
counts in the frequency readout, independent of the gate time. For example,
if I measure the 5 MHz output of my 5065A, and using a sample size of 1,
here's what I see (these are representative readings-the values are bouncing
up and down by 10 or 20 counts and I'm visually averaging them).
Increasing the sample size has no effect other than to reduce the
variability of the measurements, but the bias is still the same magnitude:

 

1 sec Gate: 5,000,000.00054 Hz

0.1 sec Gate:   5,000,000.0060 Hz

0.01 sec Gate:  5,000,000.055 Hz

1 Period Gate:  5,004,7__ Hz

 

I've had several HP 5370A/B counters, but never one of these. As I
recall-and it's been a while-there's an adjustment in the 5370A/B for the
interpolators that can remove this sort of bias. Is there a similar
adjustment in the SR620? I've been through the manual and don't see anything
obvious.

 

Thanks for any advice!

 

Stan

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] Stanford Research SR620 Measurement Bias

2009-09-27 Thread Corby Dawson
   Stan. 
Calibrating the units timebase per page 76 of the manual you can
manually
adjust the calbyte 4 word until the counter averages zero offset plus
minus around the 5Mhz.
You will still have the .0060 bias but it will plus and minus 30.
Corby Dawson

Best Weight Loss Program - Click Here!
http://thirdpartyoffers.juno.com/TGL2141/fc/BLSrjpTFoYckJJTSz9ZWtSYYnIdcHHnuViYytOdHmKkw6a10JBXXussDDPq/

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Stanford Research SR620 Measurement Bias

2009-09-27 Thread Magnus Danielson

Corby,

Corby Dawson wrote:
   Stan. 
Calibrating the units timebase per page 76 of the manual you can

manually
adjust the calbyte 4 word until the counter averages zero offset plus
minus around the 5Mhz.
You will still have the .0060 bias but it will plus and minus 30.


This will not solve his issue. Notice how the error increases in size by 
about a decade as the time-base steps down by a decade. This is not a 
time-base issue, this is a time-delay issue, a static delay between 
start and stop events. I am preparing a more detailed answer.


Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Stanford Research SR620 Measurement Bias

2009-09-27 Thread Magnus Danielson

Stan,

Stan wrote:

I just acquired an SR620 that looks to be in excellent condition. I've
noticed something in its operation and I'm not sure if it represents a
problem, a calibration issue, or if it's just a feature.

Yesterday, after letting the counter warm up for an hour, I calibrated the
10 MHz timebase and did a couple of Auto-Cals (just to be sure), but even
after all of that, there's still a bias of somewhere between 40 and 70
counts in the frequency readout, independent of the gate time. For example,
if I measure the 5 MHz output of my 5065A, and using a sample size of 1,
here's what I see (these are representative readings-the values are bouncing
up and down by 10 or 20 counts and I'm visually averaging them).
Increasing the sample size has no effect other than to reduce the
variability of the measurements, but the bias is still the same magnitude:

 


1 sec Gate: 5,000,000.00054 Hz

0.1 sec Gate:   5,000,000.0060 Hz

0.01 sec Gate:  5,000,000.055 Hz

1 Period Gate:  5,004,7__ Hz


Notice how the bias shift up one decade as the time-base shifts down a 
decade. Consider that this is a reciprocal counter which measures the 
number of events and amount of time (in its time-base):


freq = Events / Time

As we step Time a decade down, the error increases by a decade, but an 
offset in number of Events should produce a much larger error (5 Hz, 50 
Hz and 500 Hz respectively) so we must look at the time side of the 
division.


To facilitate this, I make up estimates for the Event counter and Gate 
Counter and runs the calculation backwards:


freq = Events / Time = Events / (GateCount * 90 MHz - Start + Stop)

Events = freq * Time

Time = Events / freq

tauEventsGate Count  -Start + Stop
  1 s  5.000.000 90.000.000  -107,999 ps
100 ms   500.000  9.000.000  -119,999 ps
 10 ms50.000900.000  -109,999 ps
500 ns 1 18  -199,999 ps

This is fairly consistent. Notice that the 1 cycle path is different 
from the internally gated path.


However, the cause of this time error needs some more thought.

If it where an inconsistens between the start and stop interpolators, 
the error would shift as the internal time-base beats against the 
internal 10 MHz oscillator (which is stepped up to the 90 MHz coarse 
counter clock).


However, a voltage error between start and stop trigger voltage would 
produce a stable offset. Since the start and stop input selection is the 
same in the SR620 when doing frequency measurments, this flaw is 
canceled out.


The ST620 uses a little dedicated circuit to post-process the frequency 
measurements and produce the start and stop signals being sent to 
measurement channel muxes. After this mux signals goes to the event and 
gate counter setups and also to the interpolator logics. Any systematic 
time-offset due to propagation delay variations in the start and stop 
delay of that circuit will introduce a time-bias into frequency 
measurements.


Looking at the overview of calibration bytes, byte 50 looks like the 
byte of interest. However, it is claimed that byte 50 is among those 
that is trimmed by AutoCal, which makes the peculiar statement on page 
78 that Frequency does not need calibration. It should say it does not 
need manual calibration. It may be that you could tweak this value and 
see how it changes your readings.



I've had several HP 5370A/B counters, but never one of these. As I
recall-and it's been a while-there's an adjustment in the 5370A/B for the
interpolators that can remove this sort of bias. Is there a similar
adjustment in the SR620? I've been through the manual and don't see anything
obvious.


It is hidden in the details. I think you with the brief analysis above 
should be able to come to the same conclusion.



Thanks for any advice!


Hope I got you onto the track.

Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts