Hi Ulrich, Thanks for the insight! I tried manually adjusting the value of CalByte 50 (as per Magnus' suggestion), but it did not make the problem go away. There is interaction between this adjustment and the CalByte 4 adjustment for timebase accuracy, and when all was said and done, I didn't notice any significant difference in the offset.
Where in the manual does it describe the procedure to minimize the bias? It's definitely not where I would have expected it to be! Regards, Stan Message: 1 Date: Mon, 28 Sep 2009 09:08:21 +0200 From: "Ulrich Bangert" <[email protected]> Subject: Re: [time-nuts] Stanford Research SR620 Measurement Bias To: "'Discussion of precise time and frequency measurement'" <[email protected]> Message-ID: <a4d0a2a6239b4798ab28e7193a646...@athlon> Content-Type: text/plain; charset="us-ascii" Stan, I own a SR620 as well and mine behaves the same (while with a different number of "counts") so your's is surely not defective! My understanding a few years ago when I acquired the counter was that this bias is not due to a "adjustment" in the normal sense but to set a parameter in the counter's eeprom to a correct value. Since this parameter is limited to a certain precisison you can make some effort in minimizing this bias but you should not expect to make it exactly zero. The manual explains how to do but you do not find the procedure where you would have expected it. Best regards Ulrich Bangert _______________________________________________ time-nuts mailing list -- [email protected] To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.