Re: [volt-nuts] AC Voltage Measurement Standards

2014-07-10 Thread ben
Hello,   For my 5cents worth, I also found Fluke 8506A's to be satisfactory 
- so long as they work. I only have one good remaining unit that does AC 
volts as well as DC.  I like the HP 3458A for AC volts, but always need to 
apply corrections for serious AC volts work.A better solution I now use 
is the Datron 4920 units which are an order of magnitude more accurate than 
the 3458A across their full bandwidth. However, be aware of their input 
impedance and selection of cables - always use short coaxial. The 4920 
impedance is around 100k to 400kohm (mostly frequency dependant) and this 
also varies during its sampling period. Otherwise they are great units, in 
the same ball park as a Fluke 5790 but way cheaper.  For an AC source I 
thoroughly recommend the Datron 4200 or 4700 series, which are also way 
cheaper than the near equivalent Fluke 5700. 
Ben





From: acb...@gmx.de
Sent: Wednesday, July 09, 2014 9:29 PM
To: Discussion of precise voltage measurement volt-nuts@febo.com
Subject: Re: [volt-nuts] AC Voltage Measurement Standards 

J.L.,
it really depends on what level of accuracy you are looking for. The gold 
standard obviously is the Fluke 792A. That is probably not an option. The 
Fluke 8606 is more on the lower end side (dont get me wrong, compared to 
the 792A), it does not really bring much improvements compared to the 3458A 
operated in the AC SYNC mode, but maybe it is good enogh for what you want 
to do. In that range are also e.g. the RS ac voltmeters, but their focus 
is more in the higher frequency area.
Very good accuracies are achieved by thermal converters (Ballantine...). 
You can do DC to AC comparison measurements with a few ppm accuracy. But 
measurements are slow, and you need a nanovolt meter.

I personally did the following: I got a Ballantine 1605A transfer 
voltmeter. This is comparable to the 792A in a way, except it was much 
cheaper. It is automatic, much easier to use than the Fluke 540 and goes up 
to I think 100MHz. This can be used for percision calibrations as a working 
standard. The calibration of this meter as well as others (e.g. the 3458A 
in its AC mode) I am doing with a set of thermal converters (0.5V to 100V). 
One of which (10V) has been externally calibrated up to 30MHz, cal of the 
others are derived from it. That way I am deriving everything from a very 
precisely (few ppm) calibrated 10V TVC. Overall, this saves cost on the 
calibration side, allows for high accuracy and measurement speed is good.

 Gesendet: Mittwoch, 09. Juli 2014 um 06:20 Uhr
 Von: John Phillips 
 An: mi...@vincentelectronics.com, Discussion of precise voltage 
measurement 
 Betreff: Re: [volt-nuts] AC Voltage Measurement Standards

 J. L. ,
 I do not know what level you are willing to pay for. I do not work in a
 real high end lab but we seem to get by with checking our secondary
 standards with a 3458A before and after we send it to gary to be
 calibrated. We use our 10 volts and 10K resistor to calibrate the 3458A
 before we send it in and then after. We look at the before and after 
data
 as well to come up with new values for the units. Comes down to using 
the
 3458A as the primary standard filtered by our history. The different AC
 measurement methods in statistical mode can give you a good idea where 
your
 AC values really are.
 
 I would like to know how that compares with what the extra hardware can 
do.
 
 
 
 On Tue, Jul 8, 2014 at 7:52 PM, Mitch Van Ochten  
mi...@vincentelectronics.com wrote:
 
  Joe,
 
  I used a Fluke 540 for a while, and also the Fluke 8506A.  The Fluke 
540 is
  MUCH more difficult to use. You need to flip the switch between two
  positions rapidly and the results you get depend somewhat on how 
rapidly
  you make the transition. For AC volts the 8506A seems very nice.
 
 
  Best regards,
 
  mitch
 
  -Original Message-
  From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] 
On
  Behalf Of J. L. Trantham
  Sent: Tuesday, July 08, 2014 9:46 PM
  To: Discussion of precise voltage measurement
  Subject: [volt-nuts] AC Voltage Measurement Standards
 
  I've been thinking about adding an AC Voltage Measurement Standard to 
my
  shop.
 
 
 
  It would appear that most of these have to do with thermal converters.
   Does
  anyone have any thoughts about this?
 
 
 
  I've been thinking about a Fluke 540B, 8506A, or a collection of A55
  Thermal
  Converters.  I have accurate DC measurement tools and DC standards.
 
 
 
  I would appreciate any thoughts.
 
 
 
  Thanks in advance.
 
 
 
  Joe
 
  ___
  volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
  https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
  and follow the instructions there.
 
  ___
  volt-nuts mailing list -- volt-nuts@febo.com
  To unsubscribe, go to
  https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
  

[volt-nuts] fluke 5200a repair

2014-07-10 Thread Kgoodhew
Hi Gerd and Charles,
Thanks for the input it has given me other things to think about.
I checked the output of pin 6 of U1 and I get a voltage variation of 10.2 mv
from +15 to -15v adjustment on R6, so that appears to be working correctly.
Now the fun starts :- using my HP 3478A to measure the dc on the output
terminals I get  a reading of 0.9mv on the 1 volt range, 0.3 mv on the 10
volt range, and 1.8 mv on the 100volt range.
As a quick check of wether C8 may be leaky I pulled Q38 out of it's socket
and got a constant DC out on the output terminals of 0.22mv on all three
ranges, so maybe C8 is leaky?
Using a fluke 77 3.5 digit multimeter yields vastly different DC readings,
up to 5.5 v on the 100 volt range, so I suspect the Fluke 77 is reading some
of the AC component on the output terminals even though DC is selected.
Last night I did check the input and output on the amplifier, and on the 1 v
range I was getting an input of 1v and an output of 3 volts, so the gain was
correct, I did not check it on the 100v range though to see if the gain was
30.
The input and output was a good clean sine wave on the amplifier.
I have previously checked the 190 volt + and - supplies and they were spot
on, but I will check them again before I pull the amplifier board apart
again to remove C1 and C8 to test with a higher voltage other than the
battery voltage of the multimeter, to see if they may be leaking at a higher
voltage.
Thanks,
Ken

-

Message: 2
Date: Wed, 09 Jul 2014 14:21:13 -0400
From: Charles Steinmetz csteinm...@yandex.com
To: Discussion of precise voltage measurement volt-nuts@febo.com
Subject: Re: [volt-nuts] Fluke 5200A repair
Message-ID: 20140709222123.lmbgv...@smtp2o.mail.yandex.net
Content-Type: text/plain; charset=us-ascii; format=flowed

Ken wrote:

R6 slider to U1 varies from +15v to -15v but the output of U1 only 
varies by mv's, I put this down to the loading of R9 to earth on the 
input to pin3 of U1 (221 ohm),only allowing a small variation to the 
input on pin
3 of U1, maybe this needs further investigation!

Look at the circuit.  The 100k potentiometer (R6) feeds a voltage divider
composed of R8 and R9 (464k and 221 ohms, respectively).  The tap of this
divider feeds the reference input of U1.  R8 and R9 attenuate the +/- 15v
range of R6 by a factor of 210, so the divider tap voltage (i.e., the
reference voltage for U1) only varies +/- 7.14 mV from ground for the full
range of R6.

U1 is an integrator, so it has lots of gain at DC (Q38 switches the
integrating capacitor to give two time constants).  If one of these
capacitors is leaky or shorted (or there is another leakage path from Pin 6
to Pin 2 of U1), it would reduce the gain of the integrator and could
produce the symptom you are having.

Best regards,

Charles





--

Message: 3
Date: Wed, 09 Jul 2014 14:29:40 -0400
From: Charles Steinmetz csteinm...@yandex.com
To: Discussion of precise voltage measurement volt-nuts@febo.com
Subject: Re: [volt-nuts] Fluke 5200A repair
Message-ID: 20140709222949.tnbeu...@smtp2o.mail.yandex.net
Content-Type: text/plain; charset=us-ascii; format=flowed

Oops!  the divider ratio of R8/R9 is 2100, not 210.

Best regards,

Charles





Message: 4
Date: Thu, 10 Jul 2014 18:25:32 +1000
From: Gerd ad...@controlelectronics.com.au
To: Discussion of precise voltage measurement volt-nuts@febo.com
Subject: [volt-nuts]  Fluke 5200A repair
Message-ID: 53be4dfc.7090...@controlelectronics.com.au
Content-Type: text/plain; charset=ISO-8859-1; format=flowed

Hello Ken,

The offset adjustment is about +-7mV to the input of U1. The amplifier has a
gain of 3 or 30 so adjustment is small. Is the problem the same on both the
10V and 100V ranges? Is the amplifier producing the full output voltage? Is
it distortion free?
Are both the 190V regulators working correctly?

Regards



---
This email is free from viruses and malware because avast! Antivirus protection 
is active.
http://www.avast.com

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] AC Voltage Measurement Standards

2014-07-10 Thread Stephen Grady
Todd and anyone else would is interested,

Measuring a Thermal Converter against another Thermal Converter is a bit of
a black art. The main problem is thermal converters are a square law device
that is if you change the input voltage by a factor of 2 the output voltage
will change by a factor of approximately 4. Now I say approximately because
for most thermal converters like the Fluke 540B, A55, Ballantine and Holts
the actual factor can be anywhere from 1.4 to 1.8 due to losses in the
thermal converter. In the literature you often see this factor refer to as
the N factors. Each thermal converter will have its own N factor which must
be measured to make sense of the measurements.

It even becomes more difficult in that the AC-DC difference of a thermal
converter is defined as (Vac-Vdc)/Vdc where Vac and Vdc are the inputs to a
thermal converter which give an equal output from the converter. Also Vdc is
the mean of the forward and reverse DC voltages. The problem is that when
you have two converters connected in parallel you cannot balance both
converters AC and DC inputs to produce equal voltage out of the converters
at the same time because each converter has its own AC-DC difference and its
own N factor.

It anyone is interested I can send them a technical paper that describes
this process and the appropriate math to use but I cannot send it to the
list due to copyright issues.

Kind Regards,

Stephen Grady
Sydney Australia


-Original Message-
From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
Behalf Of Todd Micallef
Sent: Thursday, 10 July 2014 2:36 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] AC Voltage Measurement Standards

I would like to know more about your setup. Which source(s) are you using
for the input and which nanovoltmeter(s) are you using to transfer the 10V
TVC to your other converters? I remember you asking on PMEL forum about the
accuracy of using a 34420A nanovoltmeter. I did not see a response as to
whether you opted for an alternative like Keithley 2182(A) or a low noise
preamplifier connected to a 3458A.

I have done some reading about how NIST transfers their calibrations using
two TVC's in parallel and I am guessing that is what you are doing.

http://www.nist.gov/pml/div684/acdc/tcs.cfm

To transfer the accuracy up/down to other TVC's at different rated voltages
appears to be a difficult task since they typically need at least half the
rated max voltage to be within spec. It would be similar to starting with a
SR104 standard and transferring its value through a set of SR1010 and
SR1050 resistors using an ESI 242.

I have a few AC sources, and I would like to be able to verify my TVC's
without sending all of them out for cal. Ballantine quoted $600+ per TVC and
I haven't checked what Fluke would charge for each A55.

Todd


I personally did the following: I got a Ballantine 1605A transfer
 voltmeter. This is comparable to the 792A in a way, except it was much 
 cheaper. It is automatic, much easier to use than the Fluke 540 and 
 goes up to I think 100MHz. This can be used for percision calibrations 
 as a working standard. The calibration of this meter as well as others 
 (e.g. the 3458A in its AC mode) I am doing with a set of thermal
converters (0.5V to 100V).
 One of which (10V) has been externally calibrated up to 30MHz, cal of 
 the others are derived from it. That way I am deriving everything from 
 a very precisely (few ppm) calibrated 10V TVC. Overall, this saves 
 cost on the calibration side, allows for high accuracy and measurement
speed is good.





___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


---
This email is free from viruses and malware because avast! Antivirus protection 
is active.
http://www.avast.com

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] AC Voltage Measurement Standards

2014-07-10 Thread Todd Micallef
Ben,

Do you have the 4921 Current Shunt Adapter option for the 4920? I have not
found any information on it other than what is listed in the 4920 user's
handbook.
It mentions the unit simulates the impedance of a 540B. The guide says that
the instructions are included with the 4921. Do you happen to have a copy
of that document?

The 4920 user's guide is on the KO4BB website.

Thanks,

Todd


On Thu, Jul 10, 2014 at 5:17 AM, ben b...@veritechmeasurements.com.au
wrote:

 Hello,   For my 5cents worth, I also found Fluke 8506A's to be satisfactory
 - so long as they work. I only have one good remaining unit that does AC
 volts as well as DC.  I like the HP 3458A for AC volts, but always need to
 apply corrections for serious AC volts work.A better solution I now use
 is the Datron 4920 units which are an order of magnitude more accurate than
 the 3458A across their full bandwidth. However, be aware of their input
 impedance and selection of cables - always use short coaxial. The 4920
 impedance is around 100k to 400kohm (mostly frequency dependant) and this
 also varies during its sampling period. Otherwise they are great units, in
 the same ball park as a Fluke 5790 but way cheaper.  For an AC source I
 thoroughly recommend the Datron 4200 or 4700 series, which are also way
 cheaper than the near equivalent Fluke 5700.
 Ben





___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


[volt-nuts] Desperately seeking OPTION 10 for Datron 4708

2014-07-10 Thread Chris
I am looking to purchase OPTION 10 slide in board for the Datron 4708
Multifunction Calibrator. If anybody has a parted out unit and can sell me
just the option 10 board i would appreciate it dearly! Thank you for
considering!

Chris
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] AC Voltage Measurement Standards

2014-07-10 Thread acbern
I have a set of 7 TVCs, from 1 to 100V. having them all calibrated externally 
is just too expensive. There is no doubt that having all 7 TVCs calibrated at 
NIST or PTB, wherever you are, is much more precise, I will just not spend it 
and live with the accuracy I get. 
As source I am using a Datron 4808, as nanovolt meter an Agilent 34420. Problem 
with this is, you pointed it out, Agilent does not specify transfer accuracy, I 
am in contact with them, so far no outcome though, not sure there will be any. 
So I have not yet completed my error calculation for the TVC cal. I will 
probably end up doing some calculations based on data sheet an then vaildate by 
measurement. Need that on the 10mv range only, so effort is limited. 
additionally I also have a ratio transformer, which is good to about 10kHz, 
that also allows for validating the TVC results by comparison in lower 
frequency ranges and narrow down the tolerances.
I do have a 182-M, but do not trust it, I saw a lot of drift doing some other 
measurements although it passes performance verification, and I just got a 
2182, which has an issue that needs to be fixed first. I need to say the 34420a 
is very stable with the digital filter switched on and you can see the 
nanovolts walk until the TVC stabilizes. vice versa, if you have an issue in 
the setup (thermal drifts...), you also seethat very well, so the filter is not 
hiding this.
as far as your comment re. calibrating a set of resistor cal from one 
calibrated one, whats wrong with it? if you do that with a stable voltage and 
by voltage divider measurement using the linearity of the 3458A, you end up 
with pretty decent accuracies. I cannot comment on the 242, just did some quick 
checks some time ago and my quick assessment then was it is worse/not better 
than the 3458a method.
 

 Gesendet: Mittwoch, 09. Juli 2014 um 16:35 Uhr
 Von: Todd Micallef tmical...@gmail.com
 An: Discussion of precise voltage measurement volt-nuts@febo.com
 Betreff: Re: [volt-nuts] AC Voltage Measurement Standards

 I would like to know more about your setup. Which source(s) are you using
 for the input and which nanovoltmeter(s) are you using to transfer the 10V
 TVC to your other converters? I remember you asking on PMEL forum about the
 accuracy of using a 34420A nanovoltmeter. I did not see a response as to
 whether you opted for an alternative like Keithley 2182(A) or a low noise
 preamplifier connected to a 3458A.
 
 I have done some reading about how NIST transfers their calibrations using
 two TVC's in parallel and I am guessing that is what you are doing.
 
 http://www.nist.gov/pml/div684/acdc/tcs.cfm
 
 To transfer the accuracy up/down to other TVC's at different rated voltages
 appears to be a difficult task since they typically need at least half the
 rated max voltage to be within spec. It would be similar to starting with a
 SR104 standard and transferring its value through a set of SR1010 and
 SR1050 resistors using an ESI 242.
 
 I have a few AC sources, and I would like to be able to verify my TVC's
 without sending all of them out for cal. Ballantine quoted $600+ per TVC
 and I haven't checked what Fluke would charge for each A55.
 
 Todd
 
 
 I personally did the following: I got a Ballantine 1605A transfer
  voltmeter. This is comparable to the 792A in a way, except it was much
  cheaper. It is automatic, much easier to use than the Fluke 540 and goes up
  to I think 100MHz. This can be used for percision calibrations as a working
  standard. The calibration of this meter as well as others (e.g. the 3458A
  in its AC mode) I am doing with a set of thermal converters (0.5V to 100V).
  One of which (10V) has been externally calibrated up to 30MHz, cal of the
  others are derived from it. That way I am deriving everything from a very
  precisely (few ppm) calibrated 10V TVC. Overall, this saves cost on the
  calibration side, allows for high accuracy and measurement speed is good.
 
 
 
 
 
 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.
 
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] AC Voltage Measurement Standards

2014-07-10 Thread acbern
in a nutshell, what i am doing is that I first establish the dc (+/-) output of 
the ref. TVC at nominal and then determine the ac voltage from a precision, 
highly linear (datron 4808) ac source that generates that default output 
voltage at the key frequencies. that establishes a set of ac voltage settings. 
for these the deviations of the TVC are known from the calibration. then I do 
the same with the TVC to be calibrated. (factor of 2 nom. voltage is important 
to stay within allowable range). that way you can link both.
sure, sending all to cal is more precise, if you get the right lab at least, 
but also very expensive.

would sure be interested in your tech paper



 Gesendet: Donnerstag, 10. Juli 2014 um 13:27 Uhr
 Von: Stephen Grady grady.st...@gmail.com
 An: 'Discussion of precise voltage measurement' volt-nuts@febo.com
 Betreff: Re: [volt-nuts] AC Voltage Measurement Standards

 Todd and anyone else would is interested,
 
 Measuring a Thermal Converter against another Thermal Converter is a bit of
 a black art. The main problem is thermal converters are a square law device
 that is if you change the input voltage by a factor of 2 the output voltage
 will change by a factor of approximately 4. Now I say approximately because
 for most thermal converters like the Fluke 540B, A55, Ballantine and Holts
 the actual factor can be anywhere from 1.4 to 1.8 due to losses in the
 thermal converter. In the literature you often see this factor refer to as
 the N factors. Each thermal converter will have its own N factor which must
 be measured to make sense of the measurements.
 
 It even becomes more difficult in that the AC-DC difference of a thermal
 converter is defined as (Vac-Vdc)/Vdc where Vac and Vdc are the inputs to a
 thermal converter which give an equal output from the converter. Also Vdc is
 the mean of the forward and reverse DC voltages. The problem is that when
 you have two converters connected in parallel you cannot balance both
 converters AC and DC inputs to produce equal voltage out of the converters
 at the same time because each converter has its own AC-DC difference and its
 own N factor.
 
 It anyone is interested I can send them a technical paper that describes
 this process and the appropriate math to use but I cannot send it to the
 list due to copyright issues.
 
 Kind Regards,
 
 Stephen Grady
 Sydney Australia
 
 
 -Original Message-
 From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
 Behalf Of Todd Micallef
 Sent: Thursday, 10 July 2014 2:36 AM
 To: Discussion of precise voltage measurement
 Subject: Re: [volt-nuts] AC Voltage Measurement Standards
 
 I would like to know more about your setup. Which source(s) are you using
 for the input and which nanovoltmeter(s) are you using to transfer the 10V
 TVC to your other converters? I remember you asking on PMEL forum about the
 accuracy of using a 34420A nanovoltmeter. I did not see a response as to
 whether you opted for an alternative like Keithley 2182(A) or a low noise
 preamplifier connected to a 3458A.
 
 I have done some reading about how NIST transfers their calibrations using
 two TVC's in parallel and I am guessing that is what you are doing.
 
 http://www.nist.gov/pml/div684/acdc/tcs.cfm
 
 To transfer the accuracy up/down to other TVC's at different rated voltages
 appears to be a difficult task since they typically need at least half the
 rated max voltage to be within spec. It would be similar to starting with a
 SR104 standard and transferring its value through a set of SR1010 and
 SR1050 resistors using an ESI 242.
 
 I have a few AC sources, and I would like to be able to verify my TVC's
 without sending all of them out for cal. Ballantine quoted $600+ per TVC and
 I haven't checked what Fluke would charge for each A55.
 
 Todd
 
 
 I personally did the following: I got a Ballantine 1605A transfer
  voltmeter. This is comparable to the 792A in a way, except it was much 
  cheaper. It is automatic, much easier to use than the Fluke 540 and 
  goes up to I think 100MHz. This can be used for percision calibrations 
  as a working standard. The calibration of this meter as well as others 
  (e.g. the 3458A in its AC mode) I am doing with a set of thermal
 converters (0.5V to 100V).
  One of which (10V) has been externally calibrated up to 30MHz, cal of 
  the others are derived from it. That way I am deriving everything from 
  a very precisely (few ppm) calibrated 10V TVC. Overall, this saves 
  cost on the calibration side, allows for high accuracy and measurement
 speed is good.
 
 
 
 
 
 ___
 volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.
 
 
 ---
 This email is free from viruses and malware because avast! Antivirus 
 protection is active.
 http://www.avast.com
 
 ___
 volt-nuts 

Re: [volt-nuts] AC Voltage Measurement Standards

2014-07-10 Thread Tom Knox

The ultimate AC standard would be NIST's AC Josephson junction array. Sam Benz 
and his team continue to make advances, pushing the accuracy and output level. 
They have also substantially reduced the needed electronics placing much of the 
peripheral electronics in a custom built two-U enclosure. NIST has also 
recently introduced a fully programmable JJA that automates most of the set up 
and operation. It won't be long before we need 10 1/2 digit DMM's 
Cheers;
Thomas Knox



 From: acb...@gmx.de
 To: volt-nuts@febo.com
 Date: Fri, 11 Jul 2014 01:13:44 +0200
 Subject: Re: [volt-nuts] AC Voltage Measurement Standards
 
 in a nutshell, what i am doing is that I first establish the dc (+/-) output 
 of the ref. TVC at nominal and then determine the ac voltage from a 
 precision, highly linear (datron 4808) ac source that generates that default 
 output voltage at the key frequencies. that establishes a set of ac voltage 
 settings. for these the deviations of the TVC are known from the calibration. 
 then I do the same with the TVC to be calibrated. (factor of 2 nom. voltage 
 is important to stay within allowable range). that way you can link both.
 sure, sending all to cal is more precise, if you get the right lab at least, 
 but also very expensive.
 
 would sure be interested in your tech paper
 
 
 
  Gesendet: Donnerstag, 10. Juli 2014 um 13:27 Uhr
  Von: Stephen Grady grady.st...@gmail.com
  An: 'Discussion of precise voltage measurement' volt-nuts@febo.com
  Betreff: Re: [volt-nuts] AC Voltage Measurement Standards
 
  Todd and anyone else would is interested,
  
  Measuring a Thermal Converter against another Thermal Converter is a bit of
  a black art. The main problem is thermal converters are a square law device
  that is if you change the input voltage by a factor of 2 the output voltage
  will change by a factor of approximately 4. Now I say approximately because
  for most thermal converters like the Fluke 540B, A55, Ballantine and Holts
  the actual factor can be anywhere from 1.4 to 1.8 due to losses in the
  thermal converter. In the literature you often see this factor refer to as
  the N factors. Each thermal converter will have its own N factor which must
  be measured to make sense of the measurements.
  
  It even becomes more difficult in that the AC-DC difference of a thermal
  converter is defined as (Vac-Vdc)/Vdc where Vac and Vdc are the inputs to a
  thermal converter which give an equal output from the converter. Also Vdc is
  the mean of the forward and reverse DC voltages. The problem is that when
  you have two converters connected in parallel you cannot balance both
  converters AC and DC inputs to produce equal voltage out of the converters
  at the same time because each converter has its own AC-DC difference and its
  own N factor.
  
  It anyone is interested I can send them a technical paper that describes
  this process and the appropriate math to use but I cannot send it to the
  list due to copyright issues.
  
  Kind Regards,
  
  Stephen Grady
  Sydney Australia
  
  
  -Original Message-
  From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
  Behalf Of Todd Micallef
  Sent: Thursday, 10 July 2014 2:36 AM
  To: Discussion of precise voltage measurement
  Subject: Re: [volt-nuts] AC Voltage Measurement Standards
  
  I would like to know more about your setup. Which source(s) are you using
  for the input and which nanovoltmeter(s) are you using to transfer the 10V
  TVC to your other converters? I remember you asking on PMEL forum about the
  accuracy of using a 34420A nanovoltmeter. I did not see a response as to
  whether you opted for an alternative like Keithley 2182(A) or a low noise
  preamplifier connected to a 3458A.
  
  I have done some reading about how NIST transfers their calibrations using
  two TVC's in parallel and I am guessing that is what you are doing.
  
  http://www.nist.gov/pml/div684/acdc/tcs.cfm
  
  To transfer the accuracy up/down to other TVC's at different rated voltages
  appears to be a difficult task since they typically need at least half the
  rated max voltage to be within spec. It would be similar to starting with a
  SR104 standard and transferring its value through a set of SR1010 and
  SR1050 resistors using an ESI 242.
  
  I have a few AC sources, and I would like to be able to verify my TVC's
  without sending all of them out for cal. Ballantine quoted $600+ per TVC and
  I haven't checked what Fluke would charge for each A55.
  
  Todd
  
  
  I personally did the following: I got a Ballantine 1605A transfer
   voltmeter. This is comparable to the 792A in a way, except it was much 
   cheaper. It is automatic, much easier to use than the Fluke 540 and 
   goes up to I think 100MHz. This can be used for percision calibrations 
   as a working standard. The calibration of this meter as well as others 
   (e.g. the 3458A in its AC mode) I am doing with a set of thermal
  converters (0.5V to 

Re: [volt-nuts] AC Voltage Measurement Standards

2014-07-10 Thread Todd Micallef
The remark about calibrating the resistors was merely an example of how I
thought you may have been transferring the calibration of the 10V TVC. My
first thought was that you compared the 10V TVC to the next closest model
at fixed points. The inter-comparison would then continue from that model
to the next value down (or up).
I had a hard time trying to understand how that could be accurate, and the
thought of using a ratio transformer didn't cross my mind. Thanks for the
explanation.

I can see how the costs of calibrating a TVC can be so high. I am guessing
few labs possess the equipment to calibrate them and places like Ballantine
probably use NIST for their cals.

Todd





 as far as your comment re. calibrating a set of resistor cal from one
 calibrated one, whats wrong with it? if you do that with a stable voltage
 and by voltage divider measurement using the linearity of the 3458A, you
 end up with pretty decent accuracies. I cannot comment on the 242, just did
 some quick checks some time ago and my quick assessment then was it is
 worse/not better than the 3458a method.



___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


[volt-nuts] Need info on LN 4214 temp vs resistance plot..

2014-07-10 Thread new

I'm guessing that the 4214 10K standard has a resistance curve
that peaks around 73F??? I've searched for old pix of data and can't find
any. The 4214 I got this week had no old, original data with it in its box.

Over the past few years there have been some resistance standards
sold on fleabay with pictures of temp vs. resistor value charts pasted
on the top cover.  They may not have been 4214s, though. I also just
discovered that fleabay drops its pix on items sold past a certain date.

So, volt-nuts any help??? Alphas, Betas, charts?

Willy

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.