I used a Fluke 5100B and 8920A for my SCAL calibration. I got lucky and it
passed at Agilent. I think if you use a 3325A , you should get results that are
close.
Todd
Sent from my iPad
On May 20, 2015, at 15:46, Electronics and Books via volt-nuts
volt-nuts@febo.com wrote:
It is stated that the ac calibration of a 3458 is so difficult. at least the
equipment mentioned is so difficult to obtain. but look what you need. a hp
signal generator 3325 where i have a pile of (any other good generator will
do to), an other hp 3458 and 3 different thermocouples. very expensive
nowhere to be found and very delicate. hp not procduced them for 40 years or
ballantine which cost more then the meter your calibrating.
but now for the accuracy. the signal generator has only 3 digits of amplitude
resolution. if you see the accuracy in the hp protocol what to measure it is
not possible to generate that.
and for the thermocouples. they give only millivolts. you need an other 3458
with no offset or better a nanovolt meter to measure them.
a far easier solution is a hp powersensor 8482a on a hp power meter. it is a
thermocouple of the correct frequency range and can be measured accurate
enough for the generator.
the ac calibation is not an exact value as the volt or resistor but only a
deviation in the frequency range. it is even difficult to obtain the
deviation in 3 digits.
ok it took me an hour but then i was done.
Regards
Frans
i...@electronicsandbooks.com
Netherlands
Discere ne cesses
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.