Re: [volt-nuts] RE new 3458A

2014-08-17 Thread R.Phillips

Randy
I would guess that your display is almost certainly 'on its way out'.  I had 
a new display fitted to my 3458A earlier this year and the difference is 
astounding. The display is of the vacuum-fluorescent type, and they will 
deteriorate over time, especially if your instrument was used in a Lab. for 
24/7 - and many had been. As I live in the UK, I am not able to give you a 
price as it was a 'package deal' - but I'm sure you get better deal in the 
US.

Roy


-Original Message- 
From: Randy Evans

Sent: Sunday, August 17, 2014 6:36 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] RE new 3458A

Interesting note.  After the room cooled down from about 79F to 73F, and
another ACAL, the meter now reads +000.00035 mVDC, a more reasonable value,
although it does bounce around a couple of tenths of a uV.

Maybe that is OK?  If so, then the only issue would seem to be the  display
has some faint pixels, which a new display should fix.

Randy


On Sat, Aug 16, 2014 at 9:59 PM, Randy Evans randyevans2...@gmail.com
wrote:



The unit seems to be working so far except for one issue.  After doing an
ACAL, and making sure the Auto Zero is ON,  I short the input leads with a
copper wire shunt across the inputs and the reading is approximately
 -000.0023 mVDC.  That seems rather high.  I would expect the unit to 
short

the input leads internally and force a zero reading during the ACAL.
Anyone have any comments on this reading?

Thanks,

Randy


On Sat, Aug 16, 2014 at 3:37 PM, Richard Moore richiem5...@gmail.com
wrote:


Hi Randy -- sounds like your unit is in cal, based on your measurements
of DCV and precision 10k resistor.

Using autocal all is recommended before doing precision measurements, and
I do that if it's been more than a day or two since last use. The autocal
uses the internal Vref and an internal 10K resistor to do cal on 
everything
else, so that tells you what the basic cal procedure is. I just got my 
3458

back from Loveland, and that's what they did for me -- warmed it up, then
ran autocal, then measured everything against a Fluke 5700, aided by an 
HP

3325, and another 3458.

It has been 5 years since I replaced the display board (no exchange
deal was available then AFAIK, so I don't know what's changed) and also 
the

NVRAM board, which was dead, with one with the Snap-cap RAM chips. I did
those replacements, then sent it home for cal, which was complete, since
all the RAM was new. Now after 5 years, the unit passed all incoming
performance tests and was sent back to me without a cal process of any
kind. This tells me that an old, well-aged Vref module is a good thing. 
The

10VDC test had changed by a bit under 5ppm, or roughly 1ppm/year.

They have a cal deal -- use code 1.090 -- press them for it -- and that
saved me 30% off the normal price. I think this deal lasts until
mid-September, so my recent cal ended up at just under $400 including
shipping. I'm not sure the deal is available on new or first-time cals; 
my

unit was in their data bank.

But this is a long way of saying I don't think you need to send it for
cal -- just push Auto Cal and Enter and wait about 10 minutes and you
should be good to go.
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.





___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there. 



___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] RE new 3458A

2014-08-17 Thread Bill Gold
Randy:

Look at page #23 of the 3458A Calibration Manual.  Make yourself a 4
Terminal Short as shown and put it in as shown.

Before I did an ACAL I had around -000.00025 mVDC.  After ACAL I now
read +000.2 mVDC.  I did the CAL 0 myself so I would expect that the
meter should return to a low value, and it does.  I do get a variation of
+/- 30 nVDC using 100 PLC and just observing the variations.  As I remember
I have never seen a spec on the ZERO stability over temperature.

If I turn on the MATH function and then do 40 measurements with 100 PLC
the statistics show:

Low reading-70 nVDC
Mean reading   -28 nVDC
High reading+3.5 nVDC
Total Variation73 nVDC

So that correlates with my visual observation of 60 nVDC.  After an hour
the room had gone up around 1 degree C.  Then I observed -000.00023 mVDC.
After another ACAL the reading was again +000.2 mVDC.  This particular
meter has a negative tempco as the room temp goes up.

Obviously do an ACAL before any precision measurements requiring low
nanovolts.

Go to the Keysight website and go to Technical Support and choose
Parts.  Then enter in the Part Number  03458-66517 which is the
replacement 03458-66507 assembly and you will see the replacement part
number on the right hand side.  Click on that and you will get the
information about the exchange program and so on.  Looks like you can just
order this part online and pay for it with a credit card, but you have to
create or use an existing login account.

I needed a new display a few years ago.  At that time you could order
just the display for around $80.  Being extremely good at removing and then
inserting and soldering I ordered the part.  The problem was that the
spacing from top to bottom of the pins had changed.  It went from around 1.3
inches to around 1.5 inches.  So I had to bend the pins to fit my display
board and then get something like 72 pins into the holes on the PC Board.
It took hours.  This change is probably why HP/Agilent/Keysight doesn't let
you just get the display anymore but wants you to get the whole PC Board
assembly.  It did work just fine once installed.

Hope this helps your decision to keep or not.

Bill



- Original Message - 
From: Randy Evans randyevans2...@gmail.com
To: Discussion of precise voltage measurement volt-nuts@febo.com
Sent: Saturday, August 16, 2014 10:36 PM
Subject: Re: [volt-nuts] RE new 3458A


 Interesting note.  After the room cooled down from about 79F to 73F, and
 another ACAL, the meter now reads +000.00035 mVDC, a more reasonable
value,
 although it does bounce around a couple of tenths of a uV.

 Maybe that is OK?  If so, then the only issue would seem to be the
display
 has some faint pixels, which a new display should fix.

 Randy


 On Sat, Aug 16, 2014 at 9:59 PM, Randy Evans randyevans2...@gmail.com
 wrote:

 
  The unit seems to be working so far except for one issue.  After doing
an
  ACAL, and making sure the Auto Zero is ON,  I short the input leads with
a
  copper wire shunt across the inputs and the reading is approximately
   -000.0023 mVDC.  That seems rather high.  I would expect the unit to
short
  the input leads internally and force a zero reading during the ACAL.
  Anyone have any comments on this reading?
 
  Thanks,
 
  Randy
 
 
  On Sat, Aug 16, 2014 at 3:37 PM, Richard Moore richiem5...@gmail.com
  wrote:
 
  Hi Randy -- sounds like your unit is in cal, based on your measurements
  of DCV and precision 10k resistor.
 
  Using autocal all is recommended before doing precision measurements,
and
  I do that if it's been more than a day or two since last use. The
autocal
  uses the internal Vref and an internal 10K resistor to do cal on
everything
  else, so that tells you what the basic cal procedure is. I just got my
3458
  back from Loveland, and that's what they did for me -- warmed it up,
then
  ran autocal, then measured everything against a Fluke 5700, aided by an
HP
  3325, and another 3458.
 
  It has been 5 years since I replaced the display board (no exchange
  deal was available then AFAIK, so I don't know what's changed) and also
the
  NVRAM board, which was dead, with one with the Snap-cap RAM chips. I
did
  those replacements, then sent it home for cal, which was complete,
since
  all the RAM was new. Now after 5 years, the unit passed all incoming
  performance tests and was sent back to me without a cal process of any
  kind. This tells me that an old, well-aged Vref module is a good thing.
The
  10VDC test had changed by a bit under 5ppm, or roughly 1ppm/year.
 
  They have a cal deal -- use code 1.090 -- press them for it -- and that
  saved me 30% off the normal price. I think this deal lasts until
  mid-September, so my recent cal ended up at just under $400 including
  shipping. I'm not sure the deal is available on new or first-time cals;
my
  unit was in their data bank.
 
  But this is a long way of saying I don't think you need to send

Re: [volt-nuts] RE new 3458A

2014-08-17 Thread Randy Evans
Bill,

You are very helpful and I really appreciate it.

I entered BLUE-C-03458 and I still get the ERR message (106, OUT OF RANGE
-- CAL secured).  That would be consistent with what you are saying.  I
guess its no reason to panic yet.

I'm not sure how to use the SECURE command.   I guess it's time to open it
up but I will probably have to get some PosiDrive screwdrivers.

Thanks,

Randy


On Sun, Aug 17, 2014 at 11:39 AM, Bill Gold wpgold3...@att.net wrote:

 Randy:

 You are doing a ZERO calibration on the meter rather than a ACAL.  You
 have to give it a password after the 0 and then it will do the ZERO
 calibration.  Usually the password is 3458 and comes from the factory
 that
 way.  But someone could have changed it in the past.  You can do this from
 the front panel menu using the SECURE command.  But there is a jumper
 inside
 the 3458A which disables the request for a password so that you can do any
 CAL or reset the password to what you want.  So everything is working as
 expected.  The jumper is JM600 on the outguard processor board 66505 or
 66515 on the left of the meter inside.  Try 3458 first as most are set to
 this password.  Read the Calibration Manual pg. #8 for how to get inside of
 the meter.  You will need two sizes of a PosiDrive type screwdrivers to
 accomplish this.  Again in the Cal Manual.

 Bill

 - Original Message -
 From: Randy Evans randyevans2...@gmail.com
 To: Discussion of precise voltage measurement volt-nuts@febo.com
 Sent: Sunday, August 17, 2014 11:22 AM
 Subject: Re: [volt-nuts] RE new 3458A


  Bill,
 
  I installed the U-short and executed the BLUE-C-0 front panel command and
 I
  get an ERR on the display.  I assume this is not good.  Did I do this
  correctly?
 
  Thanks,
 
  Randy
 
 
  On Sun, Aug 17, 2014 at 10:49 AM, Bill Gold wpgold3...@att.net wrote:
 
   Randy:
  
   Look at page #23 of the 3458A Calibration Manual.  Make yourself a
 4
   Terminal Short as shown and put it in as shown.
  
   Before I did an ACAL I had around -000.00025 mVDC.  After ACAL I
 now
   read +000.2 mVDC.  I did the CAL 0 myself so I would expect that
 the
   meter should return to a low value, and it does.  I do get a variation
 of
   +/- 30 nVDC using 100 PLC and just observing the variations.  As I
 remember
   I have never seen a spec on the ZERO stability over temperature.
  
   If I turn on the MATH function and then do 40 measurements with 100
 PLC
   the statistics show:
  
   Low reading-70 nVDC
   Mean reading   -28 nVDC
   High reading+3.5 nVDC
   Total Variation73 nVDC
  
   So that correlates with my visual observation of 60 nVDC.  After an
   hour
   the room had gone up around 1 degree C.  Then I observed -000.00023
 mVDC.
   After another ACAL the reading was again +000.2 mVDC.  This
 particular
   meter has a negative tempco as the room temp goes up.
  
   Obviously do an ACAL before any precision measurements requiring
 low
   nanovolts.
  
   Go to the Keysight website and go to Technical Support and choose
   Parts.  Then enter in the Part Number  03458-66517 which is the
   replacement 03458-66507 assembly and you will see the replacement
 part
   number on the right hand side.  Click on that and you will get the
   information about the exchange program and so on.  Looks like you can
 just
   order this part online and pay for it with a credit card, but you have
 to
   create or use an existing login account.
  
   I needed a new display a few years ago.  At that time you could
 order
   just the display for around $80.  Being extremely good at removing and
 then
   inserting and soldering I ordered the part.  The problem was that the
   spacing from top to bottom of the pins had changed.  It went from
 around
   1.3
   inches to around 1.5 inches.  So I had to bend the pins to fit my
 display
   board and then get something like 72 pins into the holes on the PC
 Board.
   It took hours.  This change is probably why HP/Agilent/Keysight doesn't
 let
   you just get the display anymore but wants you to get the whole PC
 Board
   assembly.  It did work just fine once installed.
  
   Hope this helps your decision to keep or not.
  
   Bill
  
  
  
   - Original Message -
   From: Randy Evans randyevans2...@gmail.com
   To: Discussion of precise voltage measurement volt-nuts@febo.com
   Sent: Saturday, August 16, 2014 10:36 PM
   Subject: Re: [volt-nuts] RE new 3458A
  
  
Interesting note.  After the room cooled down from about 79F to 73F,
 and
another ACAL, the meter now reads +000.00035 mVDC, a more reasonable
   value,
although it does bounce around a couple of tenths of a uV.
   
Maybe that is OK?  If so, then the only issue would seem to be the
   display
has some faint pixels, which a new display should fix.
   
Randy
   
   
On Sat, Aug 16, 2014 at 9:59 PM, Randy Evans
 randyevans2...@gmail.com
wrote:
   

 The unit seems

Re: [volt-nuts] RE new 3458A

2014-08-17 Thread Randy Evans
That worked! Now it reads 000.00030mV.  So far it looks good except for the
display.

Thanks,

Randy


On Sun, Aug 17, 2014 at 12:54 PM, Bill Gold wpgold3...@att.net wrote:

 You need a comma , between the 0 and the 3458.  So blu CCAL
 0,3458 Enter.

 See pg. # 64 of the CAL Manual for the syntax.

 Bill

 - Original Message -
 From: Randy Evans randyevans2...@gmail.com
 To: Discussion of precise voltage measurement volt-nuts@febo.com
 Sent: Sunday, August 17, 2014 12:25 PM
 Subject: Re: [volt-nuts] RE new 3458A


  Bill,
 
  You are very helpful and I really appreciate it.
 
  I entered BLUE-C-03458 and I still get the ERR message (106, OUT OF
 RANGE
  -- CAL secured).  That would be consistent with what you are saying.  I
  guess its no reason to panic yet.
 
  I'm not sure how to use the SECURE command.   I guess it's time to open
 it
  up but I will probably have to get some PosiDrive screwdrivers.
 
  Thanks,
 
  Randy
 
 
  On Sun, Aug 17, 2014 at 11:39 AM, Bill Gold wpgold3...@att.net wrote:
 
   Randy:
  
   You are doing a ZERO calibration on the meter rather than a ACAL.
 You
   have to give it a password after the 0 and then it will do the ZERO
   calibration.  Usually the password is 3458 and comes from the factory
   that
   way.  But someone could have changed it in the past.  You can do this
 from
   the front panel menu using the SECURE command.  But there is a jumper
   inside
   the 3458A which disables the request for a password so that you can do
 any
   CAL or reset the password to what you want.  So everything is working
 as
   expected.  The jumper is JM600 on the outguard processor board 66505 or
   66515 on the left of the meter inside.  Try 3458 first as most are
 set
 to
   this password.  Read the Calibration Manual pg. #8 for how to get
 inside
 of
   the meter.  You will need two sizes of a PosiDrive type screwdrivers to
   accomplish this.  Again in the Cal Manual.
  
   Bill
  
   - Original Message -
   From: Randy Evans randyevans2...@gmail.com
   To: Discussion of precise voltage measurement volt-nuts@febo.com
   Sent: Sunday, August 17, 2014 11:22 AM
   Subject: Re: [volt-nuts] RE new 3458A
  
  
Bill,
   
I installed the U-short and executed the BLUE-C-0 front panel command
 and
   I
get an ERR on the display.  I assume this is not good.  Did I do this
correctly?
   
Thanks,
   
Randy
   
   
On Sun, Aug 17, 2014 at 10:49 AM, Bill Gold wpgold3...@att.net
 wrote:
   
 Randy:

 Look at page #23 of the 3458A Calibration Manual.  Make
 yourself
 a
   4
 Terminal Short as shown and put it in as shown.

 Before I did an ACAL I had around -000.00025 mVDC.  After ACAL
 I
   now
 read +000.2 mVDC.  I did the CAL 0 myself so I would expect
 that
   the
 meter should return to a low value, and it does.  I do get a
 variation
   of
 +/- 30 nVDC using 100 PLC and just observing the variations.  As I
   remember
 I have never seen a spec on the ZERO stability over temperature.

 If I turn on the MATH function and then do 40 measurements with
 100
   PLC
 the statistics show:

 Low reading-70 nVDC
 Mean reading   -28 nVDC
 High reading+3.5 nVDC
 Total Variation73 nVDC

 So that correlates with my visual observation of 60 nVDC.
 After
 an
 hour
 the room had gone up around 1 degree C.  Then I observed -000.00023
   mVDC.
 After another ACAL the reading was again +000.2 mVDC.  This
   particular
 meter has a negative tempco as the room temp goes up.

 Obviously do an ACAL before any precision measurements
 requiring
   low
 nanovolts.

 Go to the Keysight website and go to Technical Support and
 choose
 Parts.  Then enter in the Part Number  03458-66517 which is
 the
 replacement 03458-66507 assembly and you will see the replacement
   part
 number on the right hand side.  Click on that and you will get the
 information about the exchange program and so on.  Looks like you
 can
   just
 order this part online and pay for it with a credit card, but you
 have
   to
 create or use an existing login account.

 I needed a new display a few years ago.  At that time you could
   order
 just the display for around $80.  Being extremely good at removing
 and
   then
 inserting and soldering I ordered the part.  The problem was that
 the
 spacing from top to bottom of the pins had changed.  It went from
   around
 1.3
 inches to around 1.5 inches.  So I had to bend the pins to fit my
   display
 board and then get something like 72 pins into the holes on the PC
   Board.
 It took hours.  This change is probably why HP/Agilent/Keysight
 doesn't
   let
 you just get the display anymore but wants you to get the whole PC
   Board
 assembly.  It did work just fine once installed.

 Hope this helps

Re: [volt-nuts] RE new 3458A

2014-08-17 Thread Todd Micallef
Randy,

Do you mean that you can select it, but it won't give you a temperature? Or
do you mean that you can't find it in the menu? If you go to the MENU
option, select FULL or something like that. It should enable all the extra
menus.

If you can see it, I believe all you have to do is find it under the menu
and hit Enter. I have tested mine and have never had an issue with getting
a temperature measurement.

Todd


On Sun, Aug 17, 2014 at 11:48 PM, Randy Evans randyevans2...@gmail.com
wrote:

 Can someone explain to me how to read the internal temperature using the
 front panel  TEMP? command.  I have played around but I am doing something
 wrong.

 thanks,

 Randy


 On Sun, Aug 17, 2014 at 5:13 PM, Randy Evans randyevans2...@gmail.com
 wrote:

  That worked! Now it reads 000.00030mV.  So far it looks good except for
  the display.
 
  Thanks,
 
  Randy
 
 
  On Sun, Aug 17, 2014 at 12:54 PM, Bill Gold wpgold3...@att.net wrote:
 
  You need a comma , between the 0 and the 3458.  So blu CCAL
  0,3458 Enter.
 
  See pg. # 64 of the CAL Manual for the syntax.
 
  Bill
 
  - Original Message -
  From: Randy Evans randyevans2...@gmail.com
  To: Discussion of precise voltage measurement volt-nuts@febo.com
  Sent: Sunday, August 17, 2014 12:25 PM
  Subject: Re: [volt-nuts] RE new 3458A
 
 
   Bill,
  
   You are very helpful and I really appreciate it.
  
   I entered BLUE-C-03458 and I still get the ERR message (106, OUT OF
  RANGE
   -- CAL secured).  That would be consistent with what you are
 saying.  I
   guess its no reason to panic yet.
  
   I'm not sure how to use the SECURE command.   I guess it's time to
 open
  it
   up but I will probably have to get some PosiDrive screwdrivers.
  
   Thanks,
  
   Randy
  
  
   On Sun, Aug 17, 2014 at 11:39 AM, Bill Gold wpgold3...@att.net
 wrote:
  
Randy:
   
You are doing a ZERO calibration on the meter rather than a
 ACAL.
  You
have to give it a password after the 0 and then it will do the
 ZERO
calibration.  Usually the password is 3458 and comes from the
  factory
that
way.  But someone could have changed it in the past.  You can do
 this
  from
the front panel menu using the SECURE command.  But there is a
 jumper
inside
the 3458A which disables the request for a password so that you can
 do
  any
CAL or reset the password to what you want.  So everything is
 working
  as
expected.  The jumper is JM600 on the outguard processor board 66505
  or
66515 on the left of the meter inside.  Try 3458 first as most are
  set
  to
this password.  Read the Calibration Manual pg. #8 for how to get
  inside
  of
the meter.  You will need two sizes of a PosiDrive type screwdrivers
  to
accomplish this.  Again in the Cal Manual.
   
Bill
   
- Original Message -
From: Randy Evans randyevans2...@gmail.com
To: Discussion of precise voltage measurement volt-nuts@febo.com
 
Sent: Sunday, August 17, 2014 11:22 AM
Subject: Re: [volt-nuts] RE new 3458A
   
   
 Bill,

 I installed the U-short and executed the BLUE-C-0 front panel
  command
  and
I
 get an ERR on the display.  I assume this is not good.  Did I do
  this
 correctly?

 Thanks,

 Randy


 On Sun, Aug 17, 2014 at 10:49 AM, Bill Gold wpgold3...@att.net
  wrote:

  Randy:
 
  Look at page #23 of the 3458A Calibration Manual.  Make
  yourself
  a
4
  Terminal Short as shown and put it in as shown.
 
  Before I did an ACAL I had around -000.00025 mVDC.  After
  ACAL I
now
  read +000.2 mVDC.  I did the CAL 0 myself so I would
 expect
  that
the
  meter should return to a low value, and it does.  I do get a
  variation
of
  +/- 30 nVDC using 100 PLC and just observing the variations.
 As I
remember
  I have never seen a spec on the ZERO stability over temperature.
 
  If I turn on the MATH function and then do 40 measurements
  with
  100
PLC
  the statistics show:
 
  Low reading-70 nVDC
  Mean reading   -28 nVDC
  High reading+3.5 nVDC
  Total Variation73 nVDC
 
  So that correlates with my visual observation of 60 nVDC.
  After
  an
  hour
  the room had gone up around 1 degree C.  Then I observed
  -000.00023
mVDC.
  After another ACAL the reading was again +000.2 mVDC.  This
particular
  meter has a negative tempco as the room temp goes up.
 
  Obviously do an ACAL before any precision measurements
  requiring
low
  nanovolts.
 
  Go to the Keysight website and go to Technical Support and
  choose
  Parts.  Then enter in the Part Number  03458-66517 which
 is
  the
  replacement 03458-66507 assembly and you will see the
  replacement
part
  number on the right hand side.  Click on that and you will get
 the
  information about