Re: Instruments and UL101/IEC 60065 Leakage Current measurements - hi speed

2005-05-05 Thread owner-emc-p...@listserv.ieee.org
Pete Perkins peperkin...@cs.com wrote (in 000801c551a2$7fc843f0$0400a8c0@Pete97219Compaq) about 'Instruments and UL101/IEC 60065 Leakage Current measurements - hi speed', on Thu, 5 May 2005: In the context of IEC 60990, hi-speed means circuits that will work up to 1 MHz - as described

RE: Instruments and UL101/IEC 60065 Leakage Current measurements - hi speed

2005-05-05 Thread owner-emc-p...@listserv.ieee.org
John, et al, In the context of IEC 60990, hi-speed means circuits that will work up to 1 MHz - as described in the frequency response curves. :) br, Pete Peter E Perkins, PE Principle Product Safety Engineer PO Box 23427 Tigard, ORe 97281-3427 503/452-1201 fone/fax

Re: Instruments and UL101/IEC 60065 Leakage Current measurements

2005-05-03 Thread owner-emc-p...@listserv.ieee.org
Pete Perkins peperkin...@cs.com wrote (in 004a01c54f48$33aa3700$0400a8c0@Pete97219Compaq) about 'Instruments and UL101/IEC 60065 Leakage Current measurements', on Mon, 2 May 2005: Yes, it is straightforward to build the circuits described in the standard, but not so trivial for anyone

RE: Instruments and UL101/IEC 60065 Leakage Current measurements

2005-05-02 Thread owner-emc-p...@listserv.ieee.org
Brian, et al, Perhaps you would be willing to share some of your DSoscillograms - albeit privately, since the IEEE server strips away attachments. (or some other clever scheme to get them thru). Regarding the construction of the IEC 60990 TC circuits. Yes, it is straightforward to

Re: Leakage Current Measurements

2000-10-04 Thread Randolf Keller/Jpn/TUV
...@majordomo.ieee.org Sent by: Subject: Re: Leakage Current Measurements owner-emc-pst c

Re: Leakage Current Measurements

2000-10-04 Thread David Gelfand
If anyone is curious, I have a graph of the frequency response of Annex D using an excel spreadsheet. Contact me off-line. Regards, David. David Gelfand Regulatory Approvals Memotec Communications Inc. Montreal Canada --- This message is from the IEEE

Re: Leakage Current Measurements

2000-10-04 Thread KBJ
MATSUDAK@curtiTo: emc-p...@majordomo.ieee.org sinst.com cc: Sent by: Subject: Leakage Current

Re: Leakage Current Measurements

2000-10-04 Thread Robert Macy
Cc: emc-p...@majordomo.ieee.org emc-p...@majordomo.ieee.org List-Post: emc-pstc@listserv.ieee.org Date: Tuesday, October 03, 2000 3:04 PM Subject: Re: Leakage Current Measurements Hi Ken: I found something interesting today. While measuring leakage current with a digital multimeter, I

Re: Leakage Current Measurements

2000-10-04 Thread Ralph Cameron
Sent: Tuesday, October 03, 2000 3:44 PM Subject: Leakage Current Measurements I found something interesting today. While measuring leakage current with a digital multimeter, I noticed a high amount of leakage. The unit incorporates a switching power supply, with some EMI/EMC circuitry. Anyhow

Re: Leakage Current Measurements

2000-10-03 Thread Rich Nute
Hi Ken: I found something interesting today. While measuring leakage current with a digital multimeter, I noticed a high amount of leakage. The unit incorporates a switching power supply, with some EMI/EMC circuitry. Anyhow, I read somewhere that DMMs are some times inaccurate

RE: Leakage Current Measurements

2000-10-03 Thread Jim Eichner
Subject: Leakage Current Measurements I found something interesting today. While measuring leakage current with a digital multimeter, I noticed a high amount of leakage. The unit incorporates a switching power supply, with some EMI/EMC circuitry. Anyhow, I read somewhere that DMMs are some times

Leakage Current Measurements

2000-10-03 Thread Matsuda, Ken
I found something interesting today. While measuring leakage current with a digital multimeter, I noticed a high amount of leakage. The unit incorporates a switching power supply, with some EMI/EMC circuitry. Anyhow, I read somewhere that DMMs are some times inaccurate with high frequency line

RE: Leakage Current Measurements

1999-03-19 Thread POWELL, DOUG
To: emc-p...@ieee.org Subject: RE: Leakage Current Measurements Interesting. During an audit of the facilities of a former employer, the auditor specifically requested that we didn't use the Simpson 228 for EN60950 because of the roll-off of the analog movement above 100Hz. The suggested fix

Re: Leakage Current Measurements

1999-03-18 Thread Rich Nute
Hi Frank: Interesting. During an audit of the facilities of a former employer, the auditor specifically requested that we didn't use the Simpson 228 for EN60950 because of the roll-off of the analog movement above 100Hz. The suggested fix was to build the network as described

RE: Leakage Current Measurements

1999-03-18 Thread FRANK GOTO
== -Original Message- From: pe...@itl.co.il [mailto:pe...@itl.co.il] Sent: Thursday, March 11, 1999 3:28 PM To: emc-p...@ieee.org Subject: Leakage Current Measurements Dear All Safety Engineers/Technicians, I would like to make a survey of what test equipment (multimeter

RE: Leakage Current Measurements

1999-03-17 Thread POWELL, DOUG
...@ieee.org Subject: Leakage Current Measurements Dear All Safety Engineers/Technicians, I would like to make a survey of what test equipment (multimeter, scope, etc.) you or your friendly test house use to measure the voltage during the leakage current measurements in UL1950/EN 60 950. WE USE

Re: Leakage Current Measurements

1999-03-17 Thread Rich Nute
A meter for IEC 60950 may not be adequate for IEC 61010, especially if the unit under test has a reasonably high frequency switch mode power supply. The high frequencies generated in a switching mode power supply are negligible in the presence of the mains-frequency leakage

RE: Leakage Current Measurements

1999-03-17 Thread Mis, Daniel J
is www.ergonomicsusa.com. Daniel Mis -Original Message- From: me...@aol.com [mailto:me...@aol.com] Sent: Monday, March 15, 1999 7:37 AM To: ri...@sdd.hp.com; jo...@exchange.scotland.ncr.com Cc: pe...@itl.co.il; emc-p...@ieee.org Subject: Re: Leakage Current Measurements EDD Inc. also

RE: Leakage Current Measurements

1999-03-17 Thread Peter Merguerian
to: pe...@itl.co.il, emc-p...@ieee.org Subject:RE: Leakage Current Measurements Date sent: Mon, 15 Mar 1999 11:50:33 -0500 Send reply to: Russell, Ray ray_russ...@gastmfg.com Greetings, I have used the EDD leakage current for my previous 2

Re: Leakage Current Measurements

1999-03-13 Thread Rich Nute
1999 04:38:08 -0500 (EST) Message-ID: 286215c2ca1ad211a13500a024535b58424...@eukscpo3.dundee.ncr.com From: Crabb, John jo...@exchange.scotland.ncr.com To: 'pe...@itl.co.il' pe...@itl.co.il, emc-p...@ieee.org Subject: RE: Leakage Current Measurements Date: Fri, 12 Mar 1999 09:35:40

RE: Leakage Current Measurements

1999-03-12 Thread Crabb, John
networks for 950 and medical equipment measurements. From time to time when I am doing leakage current measurements, I check that the results I get ARE VERY SIMILAR to the reading I get if I just put the meter in series with the ground conductor of the product. (And they are always almost

Leakage Current Measurements

1999-03-11 Thread Peter Merguerian
Dear All Safety Engineers/Technicians, I would like to make a survey of what test equipment (multimeter, scope, etc.) you or your friendly test house use to measure the voltage during the leakage current measurements in UL1950/EN 60 950. WE USE: OUR THIRD PARTY TEST HOUSE USES: PETER S