RE: Production Line Test Requirements - Medical Devices

2001-05-23 Thread Peter Merguerian

Dear Jon, Dick and All Memebers,

I believe the UL method for Production-Line Grounding Continuity Test is
just to verify that someone did not forget to connect the earthing lead from
the appliance inlet or power corrd to the chassis. Therefore an ohmmeter,
buzzer or the like  is all that is considered necessary. The feeding of an
ac or dc current between the protective earth terminal and other parts of
the unit likely to be energized by a single insulation fault is not
necessary during the production-line tests because these tests have already
been conducted by the lab during the type tests. UL and other NRTLs want to
make sure that assembler did not forget to connect the main earthing lead. 

Also, if you check a UL or NRTL descriptive report used for the quarterly
folow-up inspections, you will note that the construction of the product has
been described in detail, including but not limited to earthing connections,
enclosure, covers, screws and washers, etc.  European test labs for
follow-up purposes, use a Critical Components List but do not describe the
construction of the product in detail. Hence, they require the  feeding on
an ac or dc low voltage current source into the earthing path for each
production unit.


Regards



PETER S. MERGUERIAN
Technical Director
I.T.L. (Product Testing) Ltd.
26 Hacharoshet St., POB 211
Or Yehuda 60251, Israel
Tel: + 972-(0)3-5339022  Fax: + 972-(0)3-5339019
Mobile: + 972-(0)54-838175






-Original Message-
From: Jon Griver [mailto:jo...@medson.com]
Sent: Tuesday, May 22, 2001 9:06 AM
To: Dick Grobner
Cc: emc-p...@majordomo.ieee.org
Subject: RE: Production Line Test Requirements - Medical Devices



Dick,

This is the situation as I understand it. Please note that a few years ago I
did factory inspections for TUV Rheinland, so I'm speaking from a position
of some knowledge (though a bit outdated).

UL includes its production test requirements in its standards. It
traditionally has only required a check of earth continuity under low
current conditions.

The European test labs work on the basis of extracting production tests from
the standards. They require these tests to be performed as a condition of
using their mark on the product. It is the test lab's decision which
production tests are to be performed and it is not mandated by any standard.
Traditionally, the European test labs require the earth continuity test to
be carried out at high current conditions, as in the standard.

The logic behind this is that if there is a fault to earth, it will cause a
high current which may cause any weak links in the earth circuit to heat up
and possibly char and increase in resistance or even burn out completely.
(We've all seen charring around sockets). Thus this test is considered more
effective in identifying loose connections, etc. due to production mistakes
than the UL test, which might miss them.

Presumably if you are selling your medical product in the EU, you have a
Notified Body (unless your product is Group 1). You should agree your
production test plan with them. Be aware that they may insist on the high
current earth continuity test.

Good Luck,

Jon Griver
Medson Ltd


 -Original Message-
 From: Dick Grobner [mailto:dick.grob...@medgraph.com]
 Sent: Monday, May 21, 2001 18:23
 To: 'Jon Griver'
 Cc: IEEE EMC-PSTC E-Mail Forum (E-mail)
 Subject: RE: Production Line Test Requirements - Medical Devices


 Jon
 Thx for the info, however, I re-checked the two known US
 standards (UL2601,
 Appendix D - Manufacturer's Responsibilities, Construction Considerations
 and Requirements for Factory Tests) (NFPA 99, Grounding Circuit
 Continuity -
 Measurement of Resistance). UL defines Production Line Grounding
 Continuity
 Test Equipment as: Any suitable continuity indication device (such as an
 ohmmeter, a battery and buzzer combination, or the like) may be used to
 determine compliance with Grounding Continuity Test requirements. The NFPA
 std is even less definitive on this matter.
 EN60601-1, Appendix B - Testing During Manufacture and/or Installation,
 Not Used. See rational to sub-clause 4.1. Which states - Tests
 described in
 this standard are type tests.

 I checked all of the amendments including EN60601-1 for medical
 systems, and
 I do not see anything changing this.

 My questions remains, where does it state to use the 10-25Amp, 50
 or 60Hz @
 6Volt for 5 seconds test? Is this a carry over from another EN
 standard? Why
 wouldn't a simple continuity tests as defined by UL be appropriate from
 equipment coming off of the production line (assuring that the protective
 earth circuit in intake)?

 Appreciate your feedback Jon!  thanks again.

 Does anyone else on this forum have any input - would appreciate it.



---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your

RE: Production Line Test Requirements - Medical Devices

2001-05-22 Thread Jon Griver

Dick,

This is the situation as I understand it. Please note that a few years ago I
did factory inspections for TUV Rheinland, so I'm speaking from a position
of some knowledge (though a bit outdated).

UL includes its production test requirements in its standards. It
traditionally has only required a check of earth continuity under low
current conditions.

The European test labs work on the basis of extracting production tests from
the standards. They require these tests to be performed as a condition of
using their mark on the product. It is the test lab's decision which
production tests are to be performed and it is not mandated by any standard.
Traditionally, the European test labs require the earth continuity test to
be carried out at high current conditions, as in the standard.

The logic behind this is that if there is a fault to earth, it will cause a
high current which may cause any weak links in the earth circuit to heat up
and possibly char and increase in resistance or even burn out completely.
(We've all seen charring around sockets). Thus this test is considered more
effective in identifying loose connections, etc. due to production mistakes
than the UL test, which might miss them.

Presumably if you are selling your medical product in the EU, you have a
Notified Body (unless your product is Group 1). You should agree your
production test plan with them. Be aware that they may insist on the high
current earth continuity test.

Good Luck,

Jon Griver
Medson Ltd


 -Original Message-
 From: Dick Grobner [mailto:dick.grob...@medgraph.com]
 Sent: Monday, May 21, 2001 18:23
 To: 'Jon Griver'
 Cc: IEEE EMC-PSTC E-Mail Forum (E-mail)
 Subject: RE: Production Line Test Requirements - Medical Devices


 Jon
 Thx for the info, however, I re-checked the two known US
 standards (UL2601,
 Appendix D - Manufacturer's Responsibilities, Construction Considerations
 and Requirements for Factory Tests) (NFPA 99, Grounding Circuit
 Continuity -
 Measurement of Resistance). UL defines Production Line Grounding
 Continuity
 Test Equipment as: Any suitable continuity indication device (such as an
 ohmmeter, a battery and buzzer combination, or the like) may be used to
 determine compliance with Grounding Continuity Test requirements. The NFPA
 std is even less definitive on this matter.
 EN60601-1, Appendix B - Testing During Manufacture and/or Installation,
 Not Used. See rational to sub-clause 4.1. Which states - Tests
 described in
 this standard are type tests.

 I checked all of the amendments including EN60601-1 for medical
 systems, and
 I do not see anything changing this.

 My questions remains, where does it state to use the 10-25Amp, 50
 or 60Hz @
 6Volt for 5 seconds test? Is this a carry over from another EN
 standard? Why
 wouldn't a simple continuity tests as defined by UL be appropriate from
 equipment coming off of the production line (assuring that the protective
 earth circuit in intake)?

 Appreciate your feedback Jon!  thanks again.

 Does anyone else on this forum have any input - would appreciate it.



---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Michael Garretson:pstc_ad...@garretson.org
 Dave Healddavehe...@mediaone.net

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.rcic.com/  click on Virtual Conference Hall,




Re: Production Line Test Requirements - Medical Devices

2001-05-22 Thread John Woodgate

200105212113.oaa14...@epgc196.sdd.hp.com, Rich Nute ri...@sdd.hp.com
inimitably wrote:
Yes, for one strand.  No, for five strands.

I agree with what you say. In Europe, the PEC within equipment may be
16/0.2, so the possibility of the connection being reduced to one or two
strands is, in one sense, 1 in 560, while for your 36 strand wire it is
very much lower.
-- 
Regards, John Woodgate, OOO - Own Opinions Only. Phone +44 (0)1268 747839
Fax +44 (0)1268 777124. http://www.jmwa.demon.co.uk Why not call a vertically-
applied manulo-pedally-operated quasi-planar chernozem-penetrating and 
excavating implement a SPADE?

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Michael Garretson:pstc_ad...@garretson.org
 Dave Healddavehe...@mediaone.net

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.rcic.com/  click on Virtual Conference Hall,




Re: Production Line Test Requirements - Medical Devices

2001-05-21 Thread Rich Nute





Hi John:


   Because continuity at low current does not ensure that the protective
   circuit will carry a large fault current - it might be 'hanging on by
   one strand'.

Yes, for one strand.  No, for five strands.

Some years ago, I did some experiments on what
problems the 25-amp test would detect.  I 
simulated broken strands by cutting them one
at a time.  With five strands intact, the circuit 
passed the 25-amp, 2-minute test.  It failed at 
4 strands and 1 minute.

(The tested wire was 18 AWG comprised of 36 
strands of 34 AWG.)

The ability of a few strands to carry the 25-amp
current depends on the free length of the 
those few strands, which in turn determines the
heat-sinking provided to those strands.  The 
free length was on the order of 3 mm.  

My experiment assumed the problem was caused by 
an incorrectly set wire stripper, that cut a 
number of strands.  So there was a very small 
free length of strands.

I published this study in the Product Safety
Newsletter, Vol. 10, No.1, January-March, 1997.


Best regards,
Rich





---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Michael Garretson:pstc_ad...@garretson.org
 Dave Healddavehe...@mediaone.net

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.rcic.com/  click on Virtual Conference Hall,




Re: Production Line Test Requirements - Medical Devices

2001-05-21 Thread John Woodgate

36BDBCA75E0FD411A80100104B93ABF202C2C079@MGCMAIL, Dick Grobner
dick.grob...@medgraph.com inimitably wrote:
Why
wouldn't a simple continuity tests as defined by UL be appropriate from
equipment coming off of the production line (assuring that the protective
earth circuit in intake)? 

Because continuity at low current does not ensure that the protective
circuit will carry a large fault current - it might be 'hanging on by
one strand'.
-- 
Regards, John Woodgate, OOO - Own Opinions Only. Phone +44 (0)1268 747839
Fax +44 (0)1268 777124. http://www.jmwa.demon.co.uk Why not call a vertically-
applied manulo-pedally-operated quasi-planar chernozem-penetrating and 
excavating implement a SPADE?

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Michael Garretson:pstc_ad...@garretson.org
 Dave Healddavehe...@mediaone.net

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.rcic.com/  click on Virtual Conference Hall,




RE: Production Line Test Requirements - Medical Devices

2001-05-21 Thread Ned Devine

Hi Dick,

The requirement for using 10 to 25 Amps comes from the test house.  It is
their mark.  If you want to use it, they can require you to do anything they
want.  Is it fair, no.  Does the requirement make engineering sense, no (I
believe that Rich Nute did an article on how this requirement did not
identify anymore failures than the simple ohm meter did).  Do you still have
to do it, yes!

If you are just declaring conformity for the MDD, use your risk analysis
(i.e. EN 1441) to show that the risk of using a ohm meter is an acceptable
one.  Your NB might review it, but you should be able to switch to the ohm
meter.

Ned


Ned Devine
Program Manager III
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI  49548

616 248 9671 Phone
616 574 9752 Fax
ndev...@entela.com  e-mail




---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Michael Garretson:pstc_ad...@garretson.org
 Dave Healddavehe...@mediaone.net

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.rcic.com/  click on Virtual Conference Hall,




RE: Production Line Test Requirements - Medical Devices

2001-05-21 Thread Dick Grobner

Jon
Thx for the info, however, I re-checked the two known US standards (UL2601,
Appendix D - Manufacturer's Responsibilities, Construction Considerations
and Requirements for Factory Tests) (NFPA 99, Grounding Circuit Continuity -
Measurement of Resistance). UL defines Production Line Grounding Continuity
Test Equipment as: Any suitable continuity indication device (such as an
ohmmeter, a battery and buzzer combination, or the like) may be used to
determine compliance with Grounding Continuity Test requirements. The NFPA
std is even less definitive on this matter.
EN60601-1, Appendix B - Testing During Manufacture and/or Installation,
Not Used. See rational to sub-clause 4.1. Which states - Tests described in
this standard are type tests.

I checked all of the amendments including EN60601-1 for medical systems, and
I do not see anything changing this.

My questions remains, where does it state to use the 10-25Amp, 50 or 60Hz @
6Volt for 5 seconds test? Is this a carry over from another EN standard? Why
wouldn't a simple continuity tests as defined by UL be appropriate from
equipment coming off of the production line (assuring that the protective
earth circuit in intake)? 

Appreciate your feedback Jon!  thanks again.
 
Does anyone else on this forum have any input - would appreciate it.


-Original Message-
From: Jon Griver [mailto:jo...@medson.com]
Sent: Monday, May 21, 2001 9:16 AM
To: Dick Grobner
Subject: RE: Production Line Test Requirements - Medical Devices


Dick,

The requirements for production tests of medical equipment generally are:

1. Hipot at 1500AC - between live and neutral connected together and earth.
2. Earth continuity at 25A between earth pin on plug and enclosure (choose a
point that may be problematic if there are manufacturing problems)
3. Earth and/or patient leakage. Whether you do both, one or neither of
these tests depends on the product.

As the production tests are intended to find production problems (as opposed
to type tests which are intended to find design problems), choose the tests
that you think make sense from the production engineering point of view.
Document the reasoning behind your choice.

IT equipment only requires Hipot and earth continuity tests.

Hope this helps,


Jon


 -Original Message-
 From: Dick Grobner [mailto:dick.grob...@medgraph.com]
 Sent: Monday, May 21, 2001 16:36
 To: 'Jon Griver'
 Subject: RE: Production Line Test Requirements - Medical Devices


 Jon
 Thanks, you have confirmed what I thought I knew. We just had our
 annual ISO
 9001 audit (2 weeks ago). One issue that one of the two auditors
 raised was
 why are you not doing the ground integrity test (25 Amp test) on 100% of
 your production units? I asked do you mean ground continuity
 (Ohm test) and
 he said no, the ground integrity test. That's when I started digging into
 the standards. UL (and the former ETL) states high potential and ground
 continuity tests on all production units. EN60601-1 states that all tests
 within this document are type tests. The auditor gave me no reference to
 any EN, etc. when I asked. So, I have a suspicion that this is
 his wish and
 not stated in any EN (at least that I know of so far). SO - I
 will continue
 to pursue with other outside sources (but not this one auditor!)
 Thanks for the reply back!
 PS - does ITE equipment require this test (ground integrity ((25
 Amp)) test?


 -Original Message-
 From: Jon Griver [mailto:jo...@medson.com]
 Sent: Sunday, May 20, 2001 1:34 AM
 To: emc-p...@majordomo.ieee.org
 Subject: FW: Production Line Test Requirements - Medical Devices



 Dick,

 To the best of my knowledge there is no EN standard or guidance on
 production tests for medical equipment (There is a standard, EN 50116, for
 IT equipment).
 Your best bet is to discuss this with your Notified Body and to come to a
 mutually acceptable set of tests.

 Regards,

 Jon Griver
 Medson Ltd.


  Good Day Everyone
  Question I have - Does anyone out there know if an EN standard
 or guidance
  document exists that deals with production line test requirements of
  finished medical devices.
  Reading in EN60601-1 it states that the test (ground resistance, high
  potential, etc.) are type test only (See appendix B, and then paragraph
  4.1). Thus - they are not identified as production line tests.
 I know that
  the NRTL's in the USA (UL/ETL) specifically call out what tests
 are to be
  performed after the production build of equipment.
  Any input would be appreciated.
  Thx
 
  Dick Grobner
  Medical Graphics Corporation
  350 Oak Grove Parkway
  St Paul MN 55127
  651-766-3395
  651-766-3389 (fax)
  dick.grob...@medgraph.com
 
 
 
  ---
  This message is from the IEEE EMC Society Product Safety
  Technical Committee emc-pstc discussion list.
 
  Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/
 
  To cancel your subscription, send mail to:
   majord...@ieee.org

FW: Production Line Test Requirements - Medical Devices

2001-05-20 Thread Jon Griver

Dick,

To the best of my knowledge there is no EN standard or guidance on
production tests for medical equipment (There is a standard, EN 50116, for
IT equipment).
Your best bet is to discuss this with your Notified Body and to come to a
mutually acceptable set of tests.

Regards,

Jon Griver
Medson Ltd.


 Good Day Everyone
 Question I have - Does anyone out there know if an EN standard or guidance
 document exists that deals with production line test requirements of
 finished medical devices.
 Reading in EN60601-1 it states that the test (ground resistance, high
 potential, etc.) are type test only (See appendix B, and then paragraph
 4.1). Thus - they are not identified as production line tests. I know that
 the NRTL's in the USA (UL/ETL) specifically call out what tests are to be
 performed after the production build of equipment.
 Any input would be appreciated.
 Thx

 Dick Grobner
 Medical Graphics Corporation
 350 Oak Grove Parkway
 St Paul MN 55127
 651-766-3395
 651-766-3389 (fax)
 dick.grob...@medgraph.com



 ---
 This message is from the IEEE EMC Society Product Safety
 Technical Committee emc-pstc discussion list.

 Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

 To cancel your subscription, send mail to:
  majord...@ieee.org
 with the single line:
  unsubscribe emc-pstc

 For help, send mail to the list administrators:
  Michael Garretson:pstc_ad...@garretson.org
  Dave Healddavehe...@mediaone.net

 For policy questions, send mail to:
  Richard Nute:   ri...@ieee.org
  Jim Bacher: j.bac...@ieee.org

 All emc-pstc postings are archived and searchable on the web at:
 http://www.rcic.com/  click on Virtual Conference Hall,




---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Michael Garretson:pstc_ad...@garretson.org
 Dave Healddavehe...@mediaone.net

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.rcic.com/  click on Virtual Conference Hall,




Production Line Test Requirements - Medical Devices

2001-05-17 Thread Dick Grobner

Good Day Everyone
Question I have - Does anyone out there know if an EN standard or guidance
document exists that deals with production line test requirements of
finished medical devices.
Reading in EN60601-1 it states that the test (ground resistance, high
potential, etc.) are type test only (See appendix B, and then paragraph
4.1). Thus - they are not identified as production line tests. I know that
the NRTL's in the USA (UL/ETL) specifically call out what tests are to be
performed after the production build of equipment.
Any input would be appreciated.
Thx

Dick Grobner
Medical Graphics Corporation
350 Oak Grove Parkway
St Paul MN 55127
651-766-3395
651-766-3389 (fax)
dick.grob...@medgraph.com



---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Michael Garretson:pstc_ad...@garretson.org
 Dave Healddavehe...@mediaone.net

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.rcic.com/  click on Virtual Conference Hall,