Re: [Freesurfer] Major problems with T2-based pial removal

2013-09-30 Thread Michael Waskom
Hi, quick question

I'm just going to rerun these without the -T2 stuff for now so I can get
going on analysis.

What stages of recon-all do I have to run to fix everything the -T2pial
mishap broke? The process flow tables on the website seem a bit out of date.

mw


On Sat, Sep 28, 2013 at 3:24 PM, Bruce Fischl fis...@nmr.mgh.harvard.eduwrote:

 hmmm, maybe Nick or Doug can investigate, although we are crazy busy at
 the moment

 On Sat, 28 Sep 2013, Michael Waskom wrote:


 On Sat, Sep 28, 2013 at 5:28 AM, Bruce Fischl fis...@nmr.mgh.harvard.edu
 
 wrote:
   hmmm. That would be strange and would result in complete failure


 I agree, but

 $ grep bbregister recon_log_withT2.txt
 $ grep fslregister recon_log_withT2.txt

 both turn up empty





 The information in this e-mail is intended only for the person to whom it
 is
 addressed. If you believe this e-mail was sent to you in error and the
 e-mail
 contains patient information, please contact the Partners Compliance
 HelpLine at
 http://www.partners.org/**compliancelinehttp://www.partners.org/complianceline.
  If the e-mail was sent to you in error
 but does not contain patient information, please contact the sender and
 properly
 dispose of the e-mail.


___
Freesurfer mailing list
Freesurfer@nmr.mgh.harvard.edu
https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer


The information in this e-mail is intended only for the person to whom it is
addressed. If you believe this e-mail was sent to you in error and the e-mail
contains patient information, please contact the Partners Compliance HelpLine at
http://www.partners.org/complianceline . If the e-mail was sent to you in error
but does not contain patient information, please contact the sender and properly
dispose of the e-mail.


Re: [Freesurfer] Major problems with T2-based pial removal

2013-09-28 Thread Bruce Fischl
hmmm. That would be strange and would result in complete failure (which I 
guess is what you have seen). For what it's worth, Matt is right and we 
haven't tested the code on many datasets, but at least so far it has been 
independent of contrast (that is, it works on both T2 and FLAIR)


cheers
Bruce
On Fri, 27 
Sep 2013, Michael Waskom wrote:



Hmm, Ok. Thanks Matt.
I still think it's weird that the T2 didn't seem to ever get registered to
the FS volume, but it's not a huge loss if we can't use them. With that
said, it might help to clarify the docs on this point a bit, as it sort of
sounded like any reasonably-sized T2-weighted image would be useful, and I
would have tried to collect a different image if I had known!


On Fri, Sep 27, 2013 at 12:58 PM, Matt Glasser m...@ma-tea.com wrote:
  So the T2-SPACE is a variable flip angle sequence that actually
  has a fair amount of T1 and MT contrast in it and relatively
  less T2 contrast.  As a result, you can clearly see the
  difference between grey and white matter, and CSF is not
  massively brighter than grey and white matter.  If you are using
  a different kind of T2w sequence where the distance between CSF
  intensity and grey matter intensity is much greater, some of the
  internal settings to the -T2pial flag will likely not be
  optimal.  I don't know if recon-all allows you to modify those
  with expert options or not.  Bruce, is it possible for the user
  to modify the –nsigma_above and -nsigma_below flags for
  mris_make_surfaces when using -T2pial?

Peace,

Matt.

From: Michael Waskom mwas...@stanford.edu
Date: Friday, September 27, 2013 2:20 PM
To: Matt Glasser m...@ma-tea.com
Cc: Bruce Fischl fis...@nmr.mgh.harvard.edu,
Freesurfer@nmr.mgh.harvard.edu Freesurfer@nmr.mgh.harvard.edu
Subject: Re: [Freesurfer] Major problems with T2-based pial removal

Hi Matt,
It's this
sequence: http://cni.stanford.edu/wiki/MR_Protocols#T2_weighted_scans (the
3D T2)

You can click the console screenshots for a closer view of the
parameters -- let me know if you need me to track down more info.

mw


On Fri, Sep 27, 2013 at 11:38 AM, Matt Glasser m...@ma-tea.com
wrote:
  Someone sent me a different kind of T2w image and it
  didn't work so well,
  so we need to know what kind of image this is.

  Peace,

  Matt.

  On 9/27/13 1:33 PM, Bruce Fischl
  fis...@nmr.mgh.harvard.edu wrote:

  nope, pretty much only with T2-space with and without a
  flair pulse
  On Fri,
  27 Sep 2013, Matt Glasser wrote:
  
   Also, what kind of T2w sequence is this?  I think we've
  mostly tested it
   with the T2-SPACE (unless Bruce has tested with other
  things).
  
   Peace,
  
   Matt.
  
   On 9/27/13 12:58 PM, Bruce Fischl
  fis...@nmr.mgh.harvard.edu wrote:
  
   Hi Michael
  
   that looks pretty bad! What is the resolution of the
  T2? The T1 that
  you
   overlay on also looks somewhat blurry - what was it?
  
   Bruce
  
  
   On Fri, 27 Sep
   2013, Michael Waskom wrote:
  
   Hi, just to make the problem a bit more transparent,
  I took a second
   screenshot of the slices with the pial surface from
  the -T2pial run in
   red
   and the pial surface from the normal run in
   cyan:
  https://dl.dropbox.com/s/jv8rcxnq50tdac6/bad_slice_both_pial.png
  
   On Fri, Sep 27, 2013 at 10:37 AM, Michael Waskom
  mwas...@stanford.edu
   wrote:
         Hi,
   I'm using FS 5.3 and trying to use hires T2-weighted
  scans to refine
   the pial mesh.
  
   However, I've found that using the T2 is causing
  rather dramatic
   problems with the pial surface.
  
   I processed the same subject's data with and without
  the -T2pial
   option. Here are links to a) pial surface image b)
  slices through the
   volume and c) the recon-all.log:
  
   With -T2pial
   
  
  https://dl.dropbox.com/s/rtjgplajfptr4xb/bad_surf_withT2.png
  
  https://dl.dropbox.com/s/d7so5qtujgg8j80/bad_slice_withT2.png
  
  https://dl.dropbox.com/s/7o00jy014hihrgu/recon_log_withT2.txt
  
   Without -T2pial
   
  
  https://dl.dropbox.com/s/51y15kogecltfbi/bad_surf_noT2.png
  
  https://dl.dropbox.com/s/3gps45kp0ngkuli/bad_slice_noT2.png
  
  https://dl.dropbox.com/s/6jih6f8fh4nwh0k/recon_log_noT2.txt
  
   The release notes for 5.3 claim that the T2 is
  registered to the T1
   before refining the surface, but I actually can't
  find any evidence of
   that sequence in the log file. For this subject, the
  T1 and T2 came
   from different scan sessions, although I have other
  data where they
   were acquired in the same session that have similar,
  if less dramatic

Re: [Freesurfer] Major problems with T2-based pial removal

2013-09-28 Thread Michael Waskom
On Sat, Sep 28, 2013 at 5:28 AM, Bruce Fischl fis...@nmr.mgh.harvard.eduwrote:

 hmmm. That would be strange and would result in complete failure


I agree, but

$ grep bbregister recon_log_withT2.txt
$ grep fslregister recon_log_withT2.txt

both turn up empty
___
Freesurfer mailing list
Freesurfer@nmr.mgh.harvard.edu
https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer


The information in this e-mail is intended only for the person to whom it is
addressed. If you believe this e-mail was sent to you in error and the e-mail
contains patient information, please contact the Partners Compliance HelpLine at
http://www.partners.org/complianceline . If the e-mail was sent to you in error
but does not contain patient information, please contact the sender and properly
dispose of the e-mail.


Re: [Freesurfer] Major problems with T2-based pial removal

2013-09-28 Thread Bruce Fischl
hmmm, maybe Nick or Doug can investigate, although we are crazy busy at 
the moment
On Sat, 28 Sep 2013, Michael Waskom wrote:

 
 On Sat, Sep 28, 2013 at 5:28 AM, Bruce Fischl fis...@nmr.mgh.harvard.edu
 wrote:
   hmmm. That would be strange and would result in complete failure
 
 
 I agree, but
 
 $ grep bbregister recon_log_withT2.txt
 $ grep fslregister recon_log_withT2.txt
 
 both turn up empty
 
 

___
Freesurfer mailing list
Freesurfer@nmr.mgh.harvard.edu
https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer


The information in this e-mail is intended only for the person to whom it is
addressed. If you believe this e-mail was sent to you in error and the e-mail
contains patient information, please contact the Partners Compliance HelpLine at
http://www.partners.org/complianceline . If the e-mail was sent to you in error
but does not contain patient information, please contact the sender and properly
dispose of the e-mail.



Re: [Freesurfer] Major problems with T2-based pial removal

2013-09-27 Thread Michael Waskom
Hi, just to make the problem a bit more transparent, I took a second
screenshot of the slices with the pial surface from the -T2pial run in red
and the pial surface from the normal run in cyan:
https://dl.dropbox.com/s/jv8rcxnq50tdac6/bad_slice_both_pial.png


On Fri, Sep 27, 2013 at 10:37 AM, Michael Waskom mwas...@stanford.eduwrote:

 Hi,

 I'm using FS 5.3 and trying to use hires T2-weighted scans to refine the
 pial mesh.

 However, I've found that using the T2 is causing rather dramatic problems
 with the pial surface.

 I processed the same subject's data with and without the -T2pial option.
 Here are links to a) pial surface image b) slices through the volume and c)
 the recon-all.log:

 With -T2pial
 
 https://dl.dropbox.com/s/rtjgplajfptr4xb/bad_surf_withT2.png
 https://dl.dropbox.com/s/d7so5qtujgg8j80/bad_slice_withT2.png
 https://dl.dropbox.com/s/7o00jy014hihrgu/recon_log_withT2.txt

 Without -T2pial
 
 https://dl.dropbox.com/s/51y15kogecltfbi/bad_surf_noT2.png
 https://dl.dropbox.com/s/3gps45kp0ngkuli/bad_slice_noT2.png
 https://dl.dropbox.com/s/6jih6f8fh4nwh0k/recon_log_noT2.txt

 The release notes for 5.3 claim that the T2 is registered to the T1 before
 refining the surface, but I actually can't find any evidence of that
 sequence in the log file. For this subject, the T1 and T2 came from
 different scan sessions, although I have other data where they were
 acquired in the same session that have similar, if less dramatic, problems
 with the pial surface.

 Any ideas?

 Michael

___
Freesurfer mailing list
Freesurfer@nmr.mgh.harvard.edu
https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer


The information in this e-mail is intended only for the person to whom it is
addressed. If you believe this e-mail was sent to you in error and the e-mail
contains patient information, please contact the Partners Compliance HelpLine at
http://www.partners.org/complianceline . If the e-mail was sent to you in error
but does not contain patient information, please contact the sender and properly
dispose of the e-mail.


Re: [Freesurfer] Major problems with T2-based pial removal

2013-09-27 Thread Bruce Fischl

Hi Michael

that looks pretty bad! What is the resolution of the T2? The T1 that you 
overlay on also looks somewhat blurry - what was it?


Bruce


On Fri, 27 Sep 
2013, Michael Waskom wrote:



Hi, just to make the problem a bit more transparent, I took a second
screenshot of the slices with the pial surface from the -T2pial run in red
and the pial surface from the normal run in
cyan: https://dl.dropbox.com/s/jv8rcxnq50tdac6/bad_slice_both_pial.png

On Fri, Sep 27, 2013 at 10:37 AM, Michael Waskom mwas...@stanford.edu
wrote:
  Hi,
I'm using FS 5.3 and trying to use hires T2-weighted scans to refine
the pial mesh.

However, I've found that using the T2 is causing rather dramatic
problems with the pial surface.

I processed the same subject's data with and without the -T2pial
option. Here are links to a) pial surface image b) slices through the
volume and c) the recon-all.log:

With -T2pial

https://dl.dropbox.com/s/rtjgplajfptr4xb/bad_surf_withT2.png
https://dl.dropbox.com/s/d7so5qtujgg8j80/bad_slice_withT2.png
https://dl.dropbox.com/s/7o00jy014hihrgu/recon_log_withT2.txt

Without -T2pial

https://dl.dropbox.com/s/51y15kogecltfbi/bad_surf_noT2.png
https://dl.dropbox.com/s/3gps45kp0ngkuli/bad_slice_noT2.png
https://dl.dropbox.com/s/6jih6f8fh4nwh0k/recon_log_noT2.txt

The release notes for 5.3 claim that the T2 is registered to the T1
before refining the surface, but I actually can't find any evidence of
that sequence in the log file. For this subject, the T1 and T2 came
from different scan sessions, although I have other data where they
were acquired in the same session that have similar, if less dramatic,
problems with the pial surface.

Any ideas?

Michael



___
Freesurfer mailing list
Freesurfer@nmr.mgh.harvard.edu
https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer


The information in this e-mail is intended only for the person to whom it is
addressed. If you believe this e-mail was sent to you in error and the e-mail
contains patient information, please contact the Partners Compliance HelpLine at
http://www.partners.org/complianceline . If the e-mail was sent to you in error
but does not contain patient information, please contact the sender and properly
dispose of the e-mail.


Re: [Freesurfer] Major problems with T2-based pial removal

2013-09-27 Thread Matt Glasser
Also, what kind of T2w sequence is this?  I think we've mostly tested it
with the T2-SPACE (unless Bruce has tested with other things).

Peace,

Matt.

On 9/27/13 12:58 PM, Bruce Fischl fis...@nmr.mgh.harvard.edu wrote:

Hi Michael

that looks pretty bad! What is the resolution of the T2? The T1 that you
overlay on also looks somewhat blurry - what was it?

Bruce


On Fri, 27 Sep 
2013, Michael Waskom wrote:

 Hi, just to make the problem a bit more transparent, I took a second
 screenshot of the slices with the pial surface from the -T2pial run in
red
 and the pial surface from the normal run in
 cyan: https://dl.dropbox.com/s/jv8rcxnq50tdac6/bad_slice_both_pial.png
 
 On Fri, Sep 27, 2013 at 10:37 AM, Michael Waskom mwas...@stanford.edu
 wrote:
   Hi,
 I'm using FS 5.3 and trying to use hires T2-weighted scans to refine
 the pial mesh.
 
 However, I've found that using the T2 is causing rather dramatic
 problems with the pial surface.
 
 I processed the same subject's data with and without the -T2pial
 option. Here are links to a) pial surface image b) slices through the
 volume and c) the recon-all.log:
 
 With -T2pial
 
 https://dl.dropbox.com/s/rtjgplajfptr4xb/bad_surf_withT2.png
 https://dl.dropbox.com/s/d7so5qtujgg8j80/bad_slice_withT2.png
 https://dl.dropbox.com/s/7o00jy014hihrgu/recon_log_withT2.txt
 
 Without -T2pial
 
 https://dl.dropbox.com/s/51y15kogecltfbi/bad_surf_noT2.png
 https://dl.dropbox.com/s/3gps45kp0ngkuli/bad_slice_noT2.png
 https://dl.dropbox.com/s/6jih6f8fh4nwh0k/recon_log_noT2.txt
 
 The release notes for 5.3 claim that the T2 is registered to the T1
 before refining the surface, but I actually can't find any evidence of
 that sequence in the log file. For this subject, the T1 and T2 came
 from different scan sessions, although I have other data where they
 were acquired in the same session that have similar, if less dramatic,
 problems with the pial surface.
 
 Any ideas?
 
 Michael
 
 
 
___
Freesurfer mailing list
Freesurfer@nmr.mgh.harvard.edu
https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer


The information in this e-mail is intended only for the person to whom it
is
addressed. If you believe this e-mail was sent to you in error and the
e-mail
contains patient information, please contact the Partners Compliance
HelpLine at
http://www.partners.org/complianceline . If the e-mail was sent to you in
error
but does not contain patient information, please contact the sender and
properly
dispose of the e-mail.


___
Freesurfer mailing list
Freesurfer@nmr.mgh.harvard.edu
https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer


Re: [Freesurfer] Major problems with T2-based pial removal

2013-09-27 Thread Bruce Fischl
nope, pretty much only with T2-space with and without a flair pulse
On Fri, 
27 Sep 2013, Matt Glasser wrote:

 Also, what kind of T2w sequence is this?  I think we've mostly tested it
 with the T2-SPACE (unless Bruce has tested with other things).

 Peace,

 Matt.

 On 9/27/13 12:58 PM, Bruce Fischl fis...@nmr.mgh.harvard.edu wrote:

 Hi Michael

 that looks pretty bad! What is the resolution of the T2? The T1 that you
 overlay on also looks somewhat blurry - what was it?

 Bruce


 On Fri, 27 Sep
 2013, Michael Waskom wrote:

 Hi, just to make the problem a bit more transparent, I took a second
 screenshot of the slices with the pial surface from the -T2pial run in
 red
 and the pial surface from the normal run in
 cyan: https://dl.dropbox.com/s/jv8rcxnq50tdac6/bad_slice_both_pial.png

 On Fri, Sep 27, 2013 at 10:37 AM, Michael Waskom mwas...@stanford.edu
 wrote:
   Hi,
 I'm using FS 5.3 and trying to use hires T2-weighted scans to refine
 the pial mesh.

 However, I've found that using the T2 is causing rather dramatic
 problems with the pial surface.

 I processed the same subject's data with and without the -T2pial
 option. Here are links to a) pial surface image b) slices through the
 volume and c) the recon-all.log:

 With -T2pial
 
 https://dl.dropbox.com/s/rtjgplajfptr4xb/bad_surf_withT2.png
 https://dl.dropbox.com/s/d7so5qtujgg8j80/bad_slice_withT2.png
 https://dl.dropbox.com/s/7o00jy014hihrgu/recon_log_withT2.txt

 Without -T2pial
 
 https://dl.dropbox.com/s/51y15kogecltfbi/bad_surf_noT2.png
 https://dl.dropbox.com/s/3gps45kp0ngkuli/bad_slice_noT2.png
 https://dl.dropbox.com/s/6jih6f8fh4nwh0k/recon_log_noT2.txt

 The release notes for 5.3 claim that the T2 is registered to the T1
 before refining the surface, but I actually can't find any evidence of
 that sequence in the log file. For this subject, the T1 and T2 came
 from different scan sessions, although I have other data where they
 were acquired in the same session that have similar, if less dramatic,
 problems with the pial surface.

 Any ideas?

 Michael



 ___
 Freesurfer mailing list
 Freesurfer@nmr.mgh.harvard.edu
 https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer


 The information in this e-mail is intended only for the person to whom it
 is
 addressed. If you believe this e-mail was sent to you in error and the
 e-mail
 contains patient information, please contact the Partners Compliance
 HelpLine at
 http://www.partners.org/complianceline . If the e-mail was sent to you in
 error
 but does not contain patient information, please contact the sender and
 properly
 dispose of the e-mail.




___
Freesurfer mailing list
Freesurfer@nmr.mgh.harvard.edu
https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer


Re: [Freesurfer] Major problems with T2-based pial removal

2013-09-27 Thread Matt Glasser
Someone sent me a different kind of T2w image and it didn't work so well,
so we need to know what kind of image this is.

Peace,

Matt.

On 9/27/13 1:33 PM, Bruce Fischl fis...@nmr.mgh.harvard.edu wrote:

nope, pretty much only with T2-space with and without a flair pulse
On Fri, 
27 Sep 2013, Matt Glasser wrote:

 Also, what kind of T2w sequence is this?  I think we've mostly tested it
 with the T2-SPACE (unless Bruce has tested with other things).

 Peace,

 Matt.

 On 9/27/13 12:58 PM, Bruce Fischl fis...@nmr.mgh.harvard.edu wrote:

 Hi Michael

 that looks pretty bad! What is the resolution of the T2? The T1 that
you
 overlay on also looks somewhat blurry - what was it?

 Bruce


 On Fri, 27 Sep
 2013, Michael Waskom wrote:

 Hi, just to make the problem a bit more transparent, I took a second
 screenshot of the slices with the pial surface from the -T2pial run in
 red
 and the pial surface from the normal run in
 cyan: https://dl.dropbox.com/s/jv8rcxnq50tdac6/bad_slice_both_pial.png

 On Fri, Sep 27, 2013 at 10:37 AM, Michael Waskom
mwas...@stanford.edu
 wrote:
   Hi,
 I'm using FS 5.3 and trying to use hires T2-weighted scans to refine
 the pial mesh.

 However, I've found that using the T2 is causing rather dramatic
 problems with the pial surface.

 I processed the same subject's data with and without the -T2pial
 option. Here are links to a) pial surface image b) slices through the
 volume and c) the recon-all.log:

 With -T2pial
 
 https://dl.dropbox.com/s/rtjgplajfptr4xb/bad_surf_withT2.png
 https://dl.dropbox.com/s/d7so5qtujgg8j80/bad_slice_withT2.png
 https://dl.dropbox.com/s/7o00jy014hihrgu/recon_log_withT2.txt

 Without -T2pial
 
 https://dl.dropbox.com/s/51y15kogecltfbi/bad_surf_noT2.png
 https://dl.dropbox.com/s/3gps45kp0ngkuli/bad_slice_noT2.png
 https://dl.dropbox.com/s/6jih6f8fh4nwh0k/recon_log_noT2.txt

 The release notes for 5.3 claim that the T2 is registered to the T1
 before refining the surface, but I actually can't find any evidence of
 that sequence in the log file. For this subject, the T1 and T2 came
 from different scan sessions, although I have other data where they
 were acquired in the same session that have similar, if less dramatic,
 problems with the pial surface.

 Any ideas?

 Michael



 ___
 Freesurfer mailing list
 Freesurfer@nmr.mgh.harvard.edu
 https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer


 The information in this e-mail is intended only for the person to whom
it
 is
 addressed. If you believe this e-mail was sent to you in error and the
 e-mail
 contains patient information, please contact the Partners Compliance
 HelpLine at
 http://www.partners.org/complianceline . If the e-mail was sent to you
in
 error
 but does not contain patient information, please contact the sender and
 properly
 dispose of the e-mail.






___
Freesurfer mailing list
Freesurfer@nmr.mgh.harvard.edu
https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer


Re: [Freesurfer] Major problems with T2-based pial removal

2013-09-27 Thread Michael Waskom
Hi Matt,

It's this sequence:
http://cni.stanford.edu/wiki/MR_Protocols#T2_weighted_scans (the 3D T2)

You can click the console screenshots for a closer view of the parameters
-- let me know if you need me to track down more info.

mw


On Fri, Sep 27, 2013 at 11:38 AM, Matt Glasser m...@ma-tea.com wrote:

 Someone sent me a different kind of T2w image and it didn't work so well,
 so we need to know what kind of image this is.

 Peace,

 Matt.

 On 9/27/13 1:33 PM, Bruce Fischl fis...@nmr.mgh.harvard.edu wrote:

 nope, pretty much only with T2-space with and without a flair pulse
 On Fri,
 27 Sep 2013, Matt Glasser wrote:
 
  Also, what kind of T2w sequence is this?  I think we've mostly tested it
  with the T2-SPACE (unless Bruce has tested with other things).
 
  Peace,
 
  Matt.
 
  On 9/27/13 12:58 PM, Bruce Fischl fis...@nmr.mgh.harvard.edu wrote:
 
  Hi Michael
 
  that looks pretty bad! What is the resolution of the T2? The T1 that
 you
  overlay on also looks somewhat blurry - what was it?
 
  Bruce
 
 
  On Fri, 27 Sep
  2013, Michael Waskom wrote:
 
  Hi, just to make the problem a bit more transparent, I took a second
  screenshot of the slices with the pial surface from the -T2pial run in
  red
  and the pial surface from the normal run in
  cyan:
 https://dl.dropbox.com/s/jv8rcxnq50tdac6/bad_slice_both_pial.png
 
  On Fri, Sep 27, 2013 at 10:37 AM, Michael Waskom
 mwas...@stanford.edu
  wrote:
Hi,
  I'm using FS 5.3 and trying to use hires T2-weighted scans to refine
  the pial mesh.
 
  However, I've found that using the T2 is causing rather dramatic
  problems with the pial surface.
 
  I processed the same subject's data with and without the -T2pial
  option. Here are links to a) pial surface image b) slices through the
  volume and c) the recon-all.log:
 
  With -T2pial
  
  https://dl.dropbox.com/s/rtjgplajfptr4xb/bad_surf_withT2.png
  https://dl.dropbox.com/s/d7so5qtujgg8j80/bad_slice_withT2.png
  https://dl.dropbox.com/s/7o00jy014hihrgu/recon_log_withT2.txt
 
  Without -T2pial
  
  https://dl.dropbox.com/s/51y15kogecltfbi/bad_surf_noT2.png
  https://dl.dropbox.com/s/3gps45kp0ngkuli/bad_slice_noT2.png
  https://dl.dropbox.com/s/6jih6f8fh4nwh0k/recon_log_noT2.txt
 
  The release notes for 5.3 claim that the T2 is registered to the T1
  before refining the surface, but I actually can't find any evidence of
  that sequence in the log file. For this subject, the T1 and T2 came
  from different scan sessions, although I have other data where they
  were acquired in the same session that have similar, if less dramatic,
  problems with the pial surface.
 
  Any ideas?
 
  Michael
 
 
 
  ___
  Freesurfer mailing list
  Freesurfer@nmr.mgh.harvard.edu
  https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer
 
 
  The information in this e-mail is intended only for the person to whom
 it
  is
  addressed. If you believe this e-mail was sent to you in error and the
  e-mail
  contains patient information, please contact the Partners Compliance
  HelpLine at
  http://www.partners.org/complianceline . If the e-mail was sent to you
 in
  error
  but does not contain patient information, please contact the sender and
  properly
  dispose of the e-mail.
 
 
 
 



___
Freesurfer mailing list
Freesurfer@nmr.mgh.harvard.edu
https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer


The information in this e-mail is intended only for the person to whom it is
addressed. If you believe this e-mail was sent to you in error and the e-mail
contains patient information, please contact the Partners Compliance HelpLine at
http://www.partners.org/complianceline . If the e-mail was sent to you in error
but does not contain patient information, please contact the sender and properly
dispose of the e-mail.


Re: [Freesurfer] Major problems with T2-based pial removal

2013-09-27 Thread Matt Glasser
So the T2-SPACE is a variable flip angle sequence that actually has a fair
amount of T1 and MT contrast in it and relatively less T2 contrast.  As a
result, you can clearly see the difference between grey and white matter,
and CSF is not massively brighter than grey and white matter.  If you are
using a different kind of T2w sequence where the distance between CSF
intensity and grey matter intensity is much greater, some of the internal
settings to the -T2pial flag will likely not be optimal.  I don't know if
recon-all allows you to modify those with expert options or not.  Bruce, is
it possible for the user to modify the ­nsigma_above and -nsigma_below flags
for mris_make_surfaces when using -T2pial?

Peace,

Matt.

From:  Michael Waskom mwas...@stanford.edu
Date:  Friday, September 27, 2013 2:20 PM
To:  Matt Glasser m...@ma-tea.com
Cc:  Bruce Fischl fis...@nmr.mgh.harvard.edu,
Freesurfer@nmr.mgh.harvard.edu Freesurfer@nmr.mgh.harvard.edu
Subject:  Re: [Freesurfer] Major problems with T2-based pial removal

Hi Matt,

It's this sequence:
http://cni.stanford.edu/wiki/MR_Protocols#T2_weighted_scans (the 3D T2)

You can click the console screenshots for a closer view of the parameters --
let me know if you need me to track down more info.

mw


On Fri, Sep 27, 2013 at 11:38 AM, Matt Glasser m...@ma-tea.com wrote:
 Someone sent me a different kind of T2w image and it didn't work so well,
 so we need to know what kind of image this is.
 
 Peace,
 
 Matt.
 
 On 9/27/13 1:33 PM, Bruce Fischl fis...@nmr.mgh.harvard.edu wrote:
 
 nope, pretty much only with T2-space with and without a flair pulse
 On Fri,
 27 Sep 2013, Matt Glasser wrote:
 
  Also, what kind of T2w sequence is this?  I think we've mostly tested it
  with the T2-SPACE (unless Bruce has tested with other things).
 
  Peace,
 
  Matt.
 
  On 9/27/13 12:58 PM, Bruce Fischl fis...@nmr.mgh.harvard.edu wrote:
 
  Hi Michael
 
  that looks pretty bad! What is the resolution of the T2? The T1 that
 you
  overlay on also looks somewhat blurry - what was it?
 
  Bruce
 
 
  On Fri, 27 Sep
  2013, Michael Waskom wrote:
 
  Hi, just to make the problem a bit more transparent, I took a second
  screenshot of the slices with the pial surface from the -T2pial run
in
  red
  and the pial surface from the normal run in
  cyan: 
 https://dl.dropbox.com/s/jv8rcxnq50tdac6/bad_slice_both_pial.png
 
  On Fri, Sep 27, 2013 at 10:37 AM, Michael Waskom
 mwas...@stanford.edu
  wrote:
Hi,
  I'm using FS 5.3 and trying to use hires T2-weighted scans to refine
  the pial mesh.
 
  However, I've found that using the T2 is causing rather dramatic
  problems with the pial surface.
 
  I processed the same subject's data with and without the -T2pial
  option. Here are links to a) pial surface image b) slices through the
  volume and c) the recon-all.log:
 
  With -T2pial
  
  https://dl.dropbox.com/s/rtjgplajfptr4xb/bad_surf_withT2.png
  https://dl.dropbox.com/s/d7so5qtujgg8j80/bad_slice_withT2.png
  https://dl.dropbox.com/s/7o00jy014hihrgu/recon_log_withT2.txt
 
  Without -T2pial
  
  https://dl.dropbox.com/s/51y15kogecltfbi/bad_surf_noT2.png
  https://dl.dropbox.com/s/3gps45kp0ngkuli/bad_slice_noT2.png
  https://dl.dropbox.com/s/6jih6f8fh4nwh0k/recon_log_noT2.txt
 
  The release notes for 5.3 claim that the T2 is registered to the T1
  before refining the surface, but I actually can't find any evidence
of
  that sequence in the log file. For this subject, the T1 and T2 came
  from different scan sessions, although I have other data where they
  were acquired in the same session that have similar, if less
 dramatic,
  problems with the pial surface.
 
  Any ideas?
 
  Michael
 
 
 
  ___
  Freesurfer mailing list
  Freesurfer@nmr.mgh.harvard.edu
  https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer
 
 
  The information in this e-mail is intended only for the person to whom
 it
  is
  addressed. If you believe this e-mail was sent to you in error and the
  e-mail
  contains patient information, please contact the Partners Compliance
  HelpLine at
  http://www.partners.org/complianceline . If the e-mail was sent to you
 in
  error
  but does not contain patient information, please contact the sender and
  properly
  dispose of the e-mail.
 
 
 
 
 
 



___
Freesurfer mailing list
Freesurfer@nmr.mgh.harvard.edu
https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer


The information in this e-mail is intended only for the person to whom it is
addressed. If you believe this e-mail was sent to you in error and the e-mail
contains patient information, please contact the Partners Compliance HelpLine at
http://www.partners.org/complianceline . If the e-mail was sent to you in error
but does not contain patient information, please contact the sender and properly
dispose of the e-mail.


Re: [Freesurfer] Major problems with T2-based pial removal

2013-09-27 Thread Michael Waskom
Hmm, Ok. Thanks Matt.

I still think it's weird that the T2 didn't seem to ever get registered to
the FS volume, but it's not a huge loss if we can't use them. With that
said, it might help to clarify the docs on this point a bit, as it sort of
sounded like any reasonably-sized T2-weighted image would be useful, and I
would have tried to collect a different image if I had known!


On Fri, Sep 27, 2013 at 12:58 PM, Matt Glasser m...@ma-tea.com wrote:

 So the T2-SPACE is a variable flip angle sequence that actually has a fair
 amount of T1 and MT contrast in it and relatively less T2 contrast.  As a
 result, you can clearly see the difference between grey and white matter,
 and CSF is not massively brighter than grey and white matter.  If you are
 using a different kind of T2w sequence where the distance between CSF
 intensity and grey matter intensity is much greater, some of the internal
 settings to the -T2pial flag will likely not be optimal.  I don't know if
 recon-all allows you to modify those with expert options or not.  Bruce, is
 it possible for the user to modify the –nsigma_above and -nsigma_below
 flags for mris_make_surfaces when using -T2pial?

 Peace,

 Matt.

 From: Michael Waskom mwas...@stanford.edu
 Date: Friday, September 27, 2013 2:20 PM
 To: Matt Glasser m...@ma-tea.com
 Cc: Bruce Fischl fis...@nmr.mgh.harvard.edu, 
 Freesurfer@nmr.mgh.harvard.edu Freesurfer@nmr.mgh.harvard.edu
 Subject: Re: [Freesurfer] Major problems with T2-based pial removal

 Hi Matt,

 It's this sequence:
 http://cni.stanford.edu/wiki/MR_Protocols#T2_weighted_scans (the 3D T2)

 You can click the console screenshots for a closer view of the parameters
 -- let me know if you need me to track down more info.

 mw


 On Fri, Sep 27, 2013 at 11:38 AM, Matt Glasser m...@ma-tea.com wrote:

 Someone sent me a different kind of T2w image and it didn't work so well,
 so we need to know what kind of image this is.

 Peace,

 Matt.

 On 9/27/13 1:33 PM, Bruce Fischl fis...@nmr.mgh.harvard.edu wrote:

 nope, pretty much only with T2-space with and without a flair pulse
 On Fri,
 27 Sep 2013, Matt Glasser wrote:
 
  Also, what kind of T2w sequence is this?  I think we've mostly tested
 it
  with the T2-SPACE (unless Bruce has tested with other things).
 
  Peace,
 
  Matt.
 
  On 9/27/13 12:58 PM, Bruce Fischl fis...@nmr.mgh.harvard.edu
 wrote:
 
  Hi Michael
 
  that looks pretty bad! What is the resolution of the T2? The T1 that
 you
  overlay on also looks somewhat blurry - what was it?
 
  Bruce
 
 
  On Fri, 27 Sep
  2013, Michael Waskom wrote:
 
  Hi, just to make the problem a bit more transparent, I took a second
  screenshot of the slices with the pial surface from the -T2pial run
 in
  red
  and the pial surface from the normal run in
  cyan:
 https://dl.dropbox.com/s/jv8rcxnq50tdac6/bad_slice_both_pial.png
 
  On Fri, Sep 27, 2013 at 10:37 AM, Michael Waskom
 mwas...@stanford.edu
  wrote:
Hi,
  I'm using FS 5.3 and trying to use hires T2-weighted scans to refine
  the pial mesh.
 
  However, I've found that using the T2 is causing rather dramatic
  problems with the pial surface.
 
  I processed the same subject's data with and without the -T2pial
  option. Here are links to a) pial surface image b) slices through the
  volume and c) the recon-all.log:
 
  With -T2pial
  
  https://dl.dropbox.com/s/rtjgplajfptr4xb/bad_surf_withT2.png
  https://dl.dropbox.com/s/d7so5qtujgg8j80/bad_slice_withT2.png
  https://dl.dropbox.com/s/7o00jy014hihrgu/recon_log_withT2.txt
 
  Without -T2pial
  
  https://dl.dropbox.com/s/51y15kogecltfbi/bad_surf_noT2.png
  https://dl.dropbox.com/s/3gps45kp0ngkuli/bad_slice_noT2.png
  https://dl.dropbox.com/s/6jih6f8fh4nwh0k/recon_log_noT2.txt
 
  The release notes for 5.3 claim that the T2 is registered to the T1
  before refining the surface, but I actually can't find any evidence
 of
  that sequence in the log file. For this subject, the T1 and T2 came
  from different scan sessions, although I have other data where they
  were acquired in the same session that have similar, if less
 dramatic,
  problems with the pial surface.
 
  Any ideas?
 
  Michael
 
 
 
  ___
  Freesurfer mailing list
  Freesurfer@nmr.mgh.harvard.edu
  https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer
 
 
  The information in this e-mail is intended only for the person to whom
 it
  is
  addressed. If you believe this e-mail was sent to you in error and the
  e-mail
  contains patient information, please contact the Partners Compliance
  HelpLine at
  http://www.partners.org/complianceline . If the e-mail was sent to
 you
 in
  error
  but does not contain patient information, please contact the sender
 and
  properly
  dispose of the e-mail.
 
 
 
 




___
Freesurfer mailing list
Freesurfer@nmr.mgh.harvard.edu
https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer

Re: [Freesurfer] Major problems with T2-based pial removal

2013-09-27 Thread Matt Glasser
I'm sure it was registered to the T1w with BBR, though I don't know the
internals of that in recon-all.  Given that you might only be the second
person to try this with non T2-SPACE T2w images (and I unfortunately had not
told Bruce about my bad experience with this), you can forgive the
FreeSurfer docs for not mentioning it yet. :)

That being said, it isn't guaranteed not to work, you just might have to do
some fiddling with the nsigma parameters to get it to work, as the defaults
for T2-SPACE are probably not good for your images.

Peace,

Matt.

From:  Michael Waskom mwas...@stanford.edu
Date:  Friday, September 27, 2013 10:06 PM
To:  Matt Glasser m...@ma-tea.com
Cc:  Bruce Fischl fis...@nmr.mgh.harvard.edu,
Freesurfer@nmr.mgh.harvard.edu Freesurfer@nmr.mgh.harvard.edu
Subject:  Re: [Freesurfer] Major problems with T2-based pial removal

Hmm, Ok. Thanks Matt.

I still think it's weird that the T2 didn't seem to ever get registered to
the FS volume, but it's not a huge loss if we can't use them. With that
said, it might help to clarify the docs on this point a bit, as it sort of
sounded like any reasonably-sized T2-weighted image would be useful, and I
would have tried to collect a different image if I had known!


On Fri, Sep 27, 2013 at 12:58 PM, Matt Glasser m...@ma-tea.com wrote:
 So the T2-SPACE is a variable flip angle sequence that actually has a fair
 amount of T1 and MT contrast in it and relatively less T2 contrast.  As a
 result, you can clearly see the difference between grey and white matter, and
 CSF is not massively brighter than grey and white matter.  If you are using a
 different kind of T2w sequence where the distance between CSF intensity and
 grey matter intensity is much greater, some of the internal settings to the
 -T2pial flag will likely not be optimal.  I don't know if recon-all allows you
 to modify those with expert options or not.  Bruce, is it possible for the
 user to modify the ­nsigma_above and -nsigma_below flags for
 mris_make_surfaces when using -T2pial?
 
 Peace,
 
 Matt.
 
 From:  Michael Waskom mwas...@stanford.edu
 Date:  Friday, September 27, 2013 2:20 PM
 To:  Matt Glasser m...@ma-tea.com
 Cc:  Bruce Fischl fis...@nmr.mgh.harvard.edu,
 Freesurfer@nmr.mgh.harvard.edu Freesurfer@nmr.mgh.harvard.edu
 Subject:  Re: [Freesurfer] Major problems with T2-based pial removal
 
 Hi Matt,
 
 It's this sequence:
 http://cni.stanford.edu/wiki/MR_Protocols#T2_weighted_scans (the 3D T2)
 
 You can click the console screenshots for a closer view of the parameters --
 let me know if you need me to track down more info.
 
 mw
 
 
 On Fri, Sep 27, 2013 at 11:38 AM, Matt Glasser m...@ma-tea.com wrote:
 Someone sent me a different kind of T2w image and it didn't work so well,
 so we need to know what kind of image this is.
 
 Peace,
 
 Matt.
 
 On 9/27/13 1:33 PM, Bruce Fischl fis...@nmr.mgh.harvard.edu wrote:
 
 nope, pretty much only with T2-space with and without a flair pulse
 On Fri,
 27 Sep 2013, Matt Glasser wrote:
 
  Also, what kind of T2w sequence is this?  I think we've mostly tested it
  with the T2-SPACE (unless Bruce has tested with other things).
 
  Peace,
 
  Matt.
 
  On 9/27/13 12:58 PM, Bruce Fischl fis...@nmr.mgh.harvard.edu wrote:
 
  Hi Michael
 
  that looks pretty bad! What is the resolution of the T2? The T1 that
 you
  overlay on also looks somewhat blurry - what was it?
 
  Bruce
 
 
  On Fri, 27 Sep
  2013, Michael Waskom wrote:
 
  Hi, just to make the problem a bit more transparent, I took a second
  screenshot of the slices with the pial surface from the -T2pial run
in
  red
  and the pial surface from the normal run in
  cyan: 
 https://dl.dropbox.com/s/jv8rcxnq50tdac6/bad_slice_both_pial.png
 
  On Fri, Sep 27, 2013 at 10:37 AM, Michael Waskom
 mwas...@stanford.edu
  wrote:
Hi,
  I'm using FS 5.3 and trying to use hires T2-weighted scans to refine
  the pial mesh.
 
  However, I've found that using the T2 is causing rather dramatic
  problems with the pial surface.
 
  I processed the same subject's data with and without the -T2pial
  option. Here are links to a) pial surface image b) slices through
the
  volume and c) the recon-all.log:
 
  With -T2pial
  
  https://dl.dropbox.com/s/rtjgplajfptr4xb/bad_surf_withT2.png
  https://dl.dropbox.com/s/d7so5qtujgg8j80/bad_slice_withT2.png
  https://dl.dropbox.com/s/7o00jy014hihrgu/recon_log_withT2.txt
 
  Without -T2pial
  
  https://dl.dropbox.com/s/51y15kogecltfbi/bad_surf_noT2.png
  https://dl.dropbox.com/s/3gps45kp0ngkuli/bad_slice_noT2.png
  https://dl.dropbox.com/s/6jih6f8fh4nwh0k/recon_log_noT2.txt
 
  The release notes for 5.3 claim that the T2 is registered to the T1
  before refining the surface, but I actually can't find any evidence
of
  that sequence in the log file. For this subject, the T1 and T2 came
  from different scan sessions, although I have other data where they
  were acquired in the same session that have similar, if less
 dramatic