Re: [time-nuts] Measuring Phase diference between different GPSDO

2013-05-03 Thread John Miles


 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of Mark C. Stephens
 Sent: Thursday, May 02, 2013 9:29 PM
 To: li...@lazygranch.com; Discussion of precise time and frequency
 measurement
 Subject: Re: [time-nuts] Measuring Phase diference between different GPSDO
 
 According to http://www.realhamradio.com/z3801a-turning-point.htm they use
 a HP 3575A phase meter to perform the measurement.
 Or perhaps I have misinterpreted the whole thing?

5370Bs were a $20,000 instrument by the time they went out of production, so
that's probably why HP recommended the use of a more economical phase meter.
Given that you have a 5370B, it's definitely the right thing to use.

-- john, KE5FX
Miles Design LLC

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Measuring Phase diference between different GPSDO

2013-05-03 Thread Bob Camp
Hi

If you have a pile of GPIB cables, you may also have a GPIB card on a PC. You 
will have a lot more luck finding free software to read and log the data from 
the 5370 than from the 3575. As mentioned before, the 5370 will be much more 
accurate / higher resolution in this application. 

Bob

On May 3, 2013, at 12:28 AM, Mark C. Stephens ma...@non-stop.com.au wrote:

 According to http://www.realhamradio.com/z3801a-turning-point.htm they use a 
 HP 3575A phase meter to perform the measurement.
 Or perhaps I have misinterpreted the whole thing?
 
 
 I do have a temperature controlled workshop that is always 24 degrees so 
 hopefully thermal drift won't be too much of an issue.
 
 
 mark
 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On 
 Behalf Of li...@lazygranch.com
 Sent: Friday, 3 May 2013 12:12 PM
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] Measuring Phase diference between different GPSDO
 
 Meditating on this a bit, I assume in a strict sense, you can only consider 
 GPSDOs phase locked if they are disciplined from the same GPS.
 
 Or is this being pedantic?
 
 ___
 time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] Looking for HP115BR service manual

2013-05-03 Thread Luciano Paramithiotti
I am looking for the HP115BR service manual.
thank You

-- 
Luciano
Timeok

visit : www.timeok.it
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Measuring Phase diference between different GPSDO

2013-05-03 Thread Mark C. Stephens
Now the trick is:
How to use the 5370B to measure the phase difference.
Anyone care to point me towards an APP note or two?!

mark

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf 
Of John Miles
Sent: Friday, 3 May 2013 5:30 PM
To: 'Discussion of precise time and frequency measurement'; li...@lazygranch.com
Subject: Re: [time-nuts] Measuring Phase diference between different GPSDO



 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] 
 On Behalf Of Mark C. Stephens
 Sent: Thursday, May 02, 2013 9:29 PM
 To: li...@lazygranch.com; Discussion of precise time and frequency 
 measurement
 Subject: Re: [time-nuts] Measuring Phase diference between different 
 GPSDO
 
 According to http://www.realhamradio.com/z3801a-turning-point.htm they 
 use a HP 3575A phase meter to perform the measurement.
 Or perhaps I have misinterpreted the whole thing?

5370Bs were a $20,000 instrument by the time they went out of production, so 
that's probably why HP recommended the use of a more economical phase meter.
Given that you have a 5370B, it's definitely the right thing to use.

-- john, KE5FX
Miles Design LLC

___
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Measuring Phase diference between different GPSDO

2013-05-03 Thread Mark C. Stephens
Hey Bob, I was thinking the exact same thought, and I found a NI PCI-GPIB card 
in a PC with a dead HDD.
These appear to be 5V PCI cards so I can't stick it in a nice rack mount server 
with raid as they all seem to be 3.3v PCI.
So I put a new HDD in the PC and I have stuck WinXP on it.
Now I am awaiting further orders!  ;)

mark

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf 
Of Bob Camp
Sent: Friday, 3 May 2013 9:21 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Measuring Phase diference between different GPSDO

Hi

If you have a pile of GPIB cables, you may also have a GPIB card on a PC. You 
will have a lot more luck finding free software to read and log the data from 
the 5370 than from the 3575. As mentioned before, the 5370 will be much more 
accurate / higher resolution in this application. 

Bob

On May 3, 2013, at 12:28 AM, Mark C. Stephens ma...@non-stop.com.au wrote:

 According to http://www.realhamradio.com/z3801a-turning-point.htm they use a 
 HP 3575A phase meter to perform the measurement.
 Or perhaps I have misinterpreted the whole thing?
 
 
 I do have a temperature controlled workshop that is always 24 degrees so 
 hopefully thermal drift won't be too much of an issue.
 
 
 mark
 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] 
 On Behalf Of li...@lazygranch.com
 Sent: Friday, 3 May 2013 12:12 PM
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] Measuring Phase diference between different 
 GPSDO
 
 Meditating on this a bit, I assume in a strict sense, you can only consider 
 GPSDOs phase locked if they are disciplined from the same GPS.
 
 Or is this being pedantic?
 
 ___
 time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Measuring Phase diference between different GPSDO

2013-05-03 Thread Azelio Boriani
5370, PC, XP, GPIB... you are ready for the TimeLab!


On Fri, May 3, 2013 at 4:12 PM, Azelio Boriani azelio.bori...@screen.itwrote:

 Take a look at the TimeLab software
 http://www.ke5fx.com/timelab/readme.htm that reads directly the 5370.


 On Fri, May 3, 2013 at 4:04 PM, Mark C. Stephens ma...@non-stop.com.auwrote:

 Now the trick is:
 How to use the 5370B to measure the phase difference.
 Anyone care to point me towards an APP note or two?!

 mark

 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of John Miles
 Sent: Friday, 3 May 2013 5:30 PM
 To: 'Discussion of precise time and frequency measurement';
 li...@lazygranch.com
 Subject: Re: [time-nuts] Measuring Phase diference between different GPSDO



  -Original Message-
  From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com]
  On Behalf Of Mark C. Stephens
  Sent: Thursday, May 02, 2013 9:29 PM
  To: li...@lazygranch.com; Discussion of precise time and frequency
  measurement
  Subject: Re: [time-nuts] Measuring Phase diference between different
  GPSDO
 
  According to http://www.realhamradio.com/z3801a-turning-point.htm they
  use a HP 3575A phase meter to perform the measurement.
  Or perhaps I have misinterpreted the whole thing?

 5370Bs were a $20,000 instrument by the time they went out of production,
 so that's probably why HP recommended the use of a more economical phase
 meter.
 Given that you have a 5370B, it's definitely the right thing to use.

 -- john, KE5FX
 Miles Design LLC

 ___
 time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.


 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Measuring Phase diference between different GPSDO

2013-05-03 Thread Azelio Boriani
Take a look at the TimeLab software
http://www.ke5fx.com/timelab/readme.htmthat reads directly the 5370.


On Fri, May 3, 2013 at 4:04 PM, Mark C. Stephens ma...@non-stop.com.auwrote:

 Now the trick is:
 How to use the 5370B to measure the phase difference.
 Anyone care to point me towards an APP note or two?!

 mark

 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of John Miles
 Sent: Friday, 3 May 2013 5:30 PM
 To: 'Discussion of precise time and frequency measurement';
 li...@lazygranch.com
 Subject: Re: [time-nuts] Measuring Phase diference between different GPSDO



  -Original Message-
  From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com]
  On Behalf Of Mark C. Stephens
  Sent: Thursday, May 02, 2013 9:29 PM
  To: li...@lazygranch.com; Discussion of precise time and frequency
  measurement
  Subject: Re: [time-nuts] Measuring Phase diference between different
  GPSDO
 
  According to http://www.realhamradio.com/z3801a-turning-point.htm they
  use a HP 3575A phase meter to perform the measurement.
  Or perhaps I have misinterpreted the whole thing?

 5370Bs were a $20,000 instrument by the time they went out of production,
 so that's probably why HP recommended the use of a more economical phase
 meter.
 Given that you have a 5370B, it's definitely the right thing to use.

 -- john, KE5FX
 Miles Design LLC

 ___
 time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.


 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Measuring Phase difference between different GPSDO

2013-05-03 Thread Bob Camp
Hi

Grab an old garage sale category PC, blow the dust out of it and fire up the
GPIB. Once it's all working next steps:

Check out TimeLab and EZGpib. Both are pretty good at talking to stuff.
TimeLab does support the 5370, not sure about the other programs.

If you have two GPSDO's with pps outputs, feed one into each channel on the
5370. Put it in start / stop mode and you should see the difference between
the two. Since they can be either ahead or behind each other things can get
a little messy. 

Next, go to the software that runs your GPSDO and look for the cable delay
setting. Adjust it on one GPSDO so it's always early (or late) by a couple
hundred nanoseconds. Now fiddle the cables to the counter so it reads a
couple hundred nanoseconds and not almost a full second :).

At this point you are reading frequency to a very high resolution. The
GPSDO's likely are only good to a couple of nanoseconds at 1 second. The
counter should be good to ~0.02 ns.

Run like this for a few minutes and see what you get.

---

Now that you have conquered that approach it's time to tear it all apart and
try something new.

Feed the 1 pps into one input of the counter (just like before). Hit the
other input with 10 MHz. You now will get a reading in the 0 to 100 ns
range. Each time a cycle slips past the 1 pps the reading will skip from
99.xxx to 0.xxx ns. That's called a phase wrap around. Click the appropriate
boxes on which ever software you are running and it should take care of the
wrap.



Either way the software should show you frequency, and frequency stability. 

Bob


-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of Mark C. Stephens
Sent: Friday, May 03, 2013 10:07 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Measuring Phase diference between different GPSDO

Hey Bob, I was thinking the exact same thought, and I found a NI PCI-GPIB
card in a PC with a dead HDD.
These appear to be 5V PCI cards so I can't stick it in a nice rack mount
server with raid as they all seem to be 3.3v PCI.
So I put a new HDD in the PC and I have stuck WinXP on it.
Now I am awaiting further orders!  ;)

mark

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of Bob Camp
Sent: Friday, 3 May 2013 9:21 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Measuring Phase diference between different GPSDO

Hi

If you have a pile of GPIB cables, you may also have a GPIB card on a PC.
You will have a lot more luck finding free software to read and log the data
from the 5370 than from the 3575. As mentioned before, the 5370 will be much
more accurate / higher resolution in this application. 

Bob

On May 3, 2013, at 12:28 AM, Mark C. Stephens ma...@non-stop.com.au wrote:

 According to http://www.realhamradio.com/z3801a-turning-point.htm they use
a HP 3575A phase meter to perform the measurement.
 Or perhaps I have misinterpreted the whole thing?
 
 
 I do have a temperature controlled workshop that is always 24 degrees so
hopefully thermal drift won't be too much of an issue.
 
 
 mark
 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] 
 On Behalf Of li...@lazygranch.com
 Sent: Friday, 3 May 2013 12:12 PM
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] Measuring Phase diference between different 
 GPSDO
 
 Meditating on this a bit, I assume in a strict sense, you can only
consider GPSDOs phase locked if they are disciplined from the same GPS.
 
 Or is this being pedantic?
 
 ___
 time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] hp5065b !!!

2013-05-03 Thread ed breya
I'm interested in these improvements too, and hope some can be 
applied to other models. I don't have a 5065A, nor do I expect to 
ever get one, but I do have three identical old Efratom units.


Whether it's worthwhile to modify any Rb units also depends on 
whether it's possible to rejuvenate their Rb lamps, as discussed here 
a number of times, so that they can be maintained virtually indefinitely.


Ed

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Measuring Phase difference between different GPSDO

2013-05-03 Thread Mark C. Stephens


Am on the way to doing my first 'phase difference' with timelab - takes an hour 
:)

This is good stuff, everything went so smoothly, from installing the OS to 
using Timelab.

Now I have to figure out how to measure and set the xtal turning temperature 
for this bunch of Z3805A's I have here.

I have a Z3816A that is super stable. SYST:STAT? usually reports 0.0us for PU, 
sometimes it will go out as far as 0.2us, but very rarely :p

So I am planning on using that guy as my reference to tune the Z3805A's.

How does that sound?


Many thanks,
mark


-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf 
Of Bob Camp
Sent: Saturday, 4 May 2013 1:09 AM
To: 'Discussion of precise time and frequency measurement'
Subject: Re: [time-nuts] Measuring Phase difference between different GPSDO

Hi

Grab an old garage sale category PC, blow the dust out of it and fire up the 
GPIB. Once it's all working next steps:

Check out TimeLab and EZGpib. Both are pretty good at talking to stuff.
TimeLab does support the 5370, not sure about the other programs.

If you have two GPSDO's with pps outputs, feed one into each channel on the 
5370. Put it in start / stop mode and you should see the difference between the 
two. Since they can be either ahead or behind each other things can get a 
little messy. 

Next, go to the software that runs your GPSDO and look for the cable delay
setting. Adjust it on one GPSDO so it's always early (or late) by a couple 
hundred nanoseconds. Now fiddle the cables to the counter so it reads a couple 
hundred nanoseconds and not almost a full second :).

At this point you are reading frequency to a very high resolution. The GPSDO's 
likely are only good to a couple of nanoseconds at 1 second. The counter should 
be good to ~0.02 ns.

Run like this for a few minutes and see what you get.

---

Now that you have conquered that approach it's time to tear it all apart and 
try something new.

Feed the 1 pps into one input of the counter (just like before). Hit the other 
input with 10 MHz. You now will get a reading in the 0 to 100 ns range. Each 
time a cycle slips past the 1 pps the reading will skip from 99.xxx to 0.xxx 
ns. That's called a phase wrap around. Click the appropriate boxes on which 
ever software you are running and it should take care of the wrap.



Either way the software should show you frequency, and frequency stability. 

Bob


-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf 
Of Mark C. Stephens
Sent: Friday, May 03, 2013 10:07 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Measuring Phase diference between different GPSDO

Hey Bob, I was thinking the exact same thought, and I found a NI PCI-GPIB card 
in a PC with a dead HDD.
These appear to be 5V PCI cards so I can't stick it in a nice rack mount server 
with raid as they all seem to be 3.3v PCI.
So I put a new HDD in the PC and I have stuck WinXP on it.
Now I am awaiting further orders!  ;)

mark

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf 
Of Bob Camp
Sent: Friday, 3 May 2013 9:21 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Measuring Phase diference between different GPSDO

Hi

If you have a pile of GPIB cables, you may also have a GPIB card on a PC.
You will have a lot more luck finding free software to read and log the data 
from the 5370 than from the 3575. As mentioned before, the 5370 will be much 
more accurate / higher resolution in this application. 

Bob

On May 3, 2013, at 12:28 AM, Mark C. Stephens ma...@non-stop.com.au wrote:

 According to http://www.realhamradio.com/z3801a-turning-point.htm they 
 use
a HP 3575A phase meter to perform the measurement.
 Or perhaps I have misinterpreted the whole thing?
 
 
 I do have a temperature controlled workshop that is always 24 degrees 
 so
hopefully thermal drift won't be too much of an issue.
 
 
 mark
 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com]
 On Behalf Of li...@lazygranch.com
 Sent: Friday, 3 May 2013 12:12 PM
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] Measuring Phase diference between different 
 GPSDO
 
 Meditating on this a bit, I assume in a strict sense, you can only
consider GPSDOs phase locked if they are disciplined from the same GPS.
 
 Or is this being pedantic?
 
 ___
 time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.


Re: [time-nuts] Measuring Phase difference between different GPSDO

2013-05-03 Thread Bob Camp
Hi

Sounds like you are up and running. TimeLab can be set for any length run you 
wish. Three to five minutes is a common one. I have a 30 day run going at the 
moment. Even while the run is going, you can switch between displays to see 
what's happening. 

Unless you have some very unusual OCXO's in your 3805's fiddling with the oven 
temperature isn't going to help things much. If they are SC based, there may 
not be a turn over temperature. Often the set point is derived either from 
crystal data or from wide range temp run performance. 

Like any piece of surplus, you can indeed have something broken somewhere. In 
that case, repair is a good idea. Unless you blew the thermistor in the OCXO, 
it's an unlikely repair though.

Bob


On May 3, 2013, at 12:48 PM, Mark C. Stephens ma...@non-stop.com.au wrote:

 
 
 Am on the way to doing my first 'phase difference' with timelab - takes an 
 hour :)
 
 This is good stuff, everything went so smoothly, from installing the OS to 
 using Timelab.
 
 Now I have to figure out how to measure and set the xtal turning temperature 
 for this bunch of Z3805A's I have here.
 
 I have a Z3816A that is super stable. SYST:STAT? usually reports 0.0us for 
 PU, sometimes it will go out as far as 0.2us, but very rarely :p
 
 So I am planning on using that guy as my reference to tune the Z3805A's.
 
 How does that sound?
 
 
 Many thanks,
 mark
 
 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On 
 Behalf Of Bob Camp
 Sent: Saturday, 4 May 2013 1:09 AM
 To: 'Discussion of precise time and frequency measurement'
 Subject: Re: [time-nuts] Measuring Phase difference between different GPSDO
 
 Hi
 
 Grab an old garage sale category PC, blow the dust out of it and fire up the 
 GPIB. Once it's all working next steps:
 
 Check out TimeLab and EZGpib. Both are pretty good at talking to stuff.
 TimeLab does support the 5370, not sure about the other programs.
 
 If you have two GPSDO's with pps outputs, feed one into each channel on the 
 5370. Put it in start / stop mode and you should see the difference between 
 the two. Since they can be either ahead or behind each other things can get a 
 little messy. 
 
 Next, go to the software that runs your GPSDO and look for the cable delay
 setting. Adjust it on one GPSDO so it's always early (or late) by a couple 
 hundred nanoseconds. Now fiddle the cables to the counter so it reads a 
 couple hundred nanoseconds and not almost a full second :).
 
 At this point you are reading frequency to a very high resolution. The 
 GPSDO's likely are only good to a couple of nanoseconds at 1 second. The 
 counter should be good to ~0.02 ns.
 
 Run like this for a few minutes and see what you get.
 
 ---
 
 Now that you have conquered that approach it's time to tear it all apart and 
 try something new.
 
 Feed the 1 pps into one input of the counter (just like before). Hit the 
 other input with 10 MHz. You now will get a reading in the 0 to 100 ns range. 
 Each time a cycle slips past the 1 pps the reading will skip from 99.xxx to 
 0.xxx ns. That's called a phase wrap around. Click the appropriate boxes on 
 which ever software you are running and it should take care of the wrap.
 
 
 
 Either way the software should show you frequency, and frequency stability. 
 
 Bob
 
 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On 
 Behalf Of Mark C. Stephens
 Sent: Friday, May 03, 2013 10:07 AM
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] Measuring Phase diference between different GPSDO
 
 Hey Bob, I was thinking the exact same thought, and I found a NI PCI-GPIB 
 card in a PC with a dead HDD.
 These appear to be 5V PCI cards so I can't stick it in a nice rack mount 
 server with raid as they all seem to be 3.3v PCI.
 So I put a new HDD in the PC and I have stuck WinXP on it.
 Now I am awaiting further orders!  ;)
 
 mark
 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On 
 Behalf Of Bob Camp
 Sent: Friday, 3 May 2013 9:21 PM
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] Measuring Phase diference between different GPSDO
 
 Hi
 
 If you have a pile of GPIB cables, you may also have a GPIB card on a PC.
 You will have a lot more luck finding free software to read and log the data 
 from the 5370 than from the 3575. As mentioned before, the 5370 will be much 
 more accurate / higher resolution in this application. 
 
 Bob
 
 On May 3, 2013, at 12:28 AM, Mark C. Stephens ma...@non-stop.com.au wrote:
 
 According to http://www.realhamradio.com/z3801a-turning-point.htm they 
 use
 a HP 3575A phase meter to perform the measurement.
 Or perhaps I have misinterpreted the whole thing?
 
 
 I do have a temperature controlled workshop that is always 24 degrees 
 so
 hopefully thermal drift won't be too much of an issue.
 

Re: [time-nuts] hp5065b !!!

2013-05-03 Thread Bob Camp
Hi

One of the unique things about the 5065 is it's great big cells. The small cell 
compact Rb's are a very different beast. There is a body of evidence 
suggesting that the cell size is significant for ultra low ADEV.

The basic concepts of improving signal (light intensity) and signal to noise 
(filtering) would apply to other units. Weather the mods give significant 
improvements on small cells is the open question. The compact Rb's may already 
be at the floor for their cell size….

Bob

On May 3, 2013, at 10:40 AM, ed breya e...@telight.com wrote:

 I'm interested in these improvements too, and hope some can be applied to 
 other models. I don't have a 5065A, nor do I expect to ever get one, but I do 
 have three identical old Efratom units.
 
 Whether it's worthwhile to modify any Rb units also depends on whether it's 
 possible to rejuvenate their Rb lamps, as discussed here a number of times, 
 so that they can be maintained virtually indefinitely.
 
 Ed
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] hp5065b !!!

2013-05-03 Thread Poul-Henning Kamp
In message bb988fb5-d829-4c9d-a8bd-f2c8731b8...@rtty.us, Bob Camp writes:

The basic concepts of improving signal (light intensity) and signal to noise
(filtering) would apply to other units.

I was wondering about that.  The normally quoted number for change
in (detected) light intensity is approx 0.1%.

It would be interesting to find out if this changes significantly
with the extra filtering ?

If so, that would likely help smaller cells more than big cells...

-- 
Poul-Henning Kamp   | UNIX since Zilog Zeus 3.20
p...@freebsd.org | TCP/IP since RFC 956
FreeBSD committer   | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] Extended length Phase Noise Measurements with TimeLab

2013-05-03 Thread Tom Knox
Recently I was looking at a co-workers Lab View measurements taken with a 
5120A. She is studying vibration on Phase Noise. I asked how Lab View compared 
to TimeLab and she said in many ways she prefers TimeLab but ran into a 
problem. The research she is currently conducting requires collecting data in 
extremely long measurement periods beyond the TimeLab limits. Is she missing 
something? If not is measurement length an easy fix? One other problem during 
her project was related to the 5120A, occasionally the vibration must hit a 
resonance and the 5120A stops measurement saying measurement  parameters were 
exceeded. It would be nice to be able to adjust or disable that feature. Thanks 
in advance for your ideas.

Thomas Knox


  
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Extended length Phase Noise Measurements with TimeLab

2013-05-03 Thread John Ackermann N8UR

Hi Tom --

Are you talking about capture time limits, or minimum tau?  I've never 
encountered any capture time limits in TimeLab and have done out to 30 days.


However, one thing I have noticed with the 5120A ethernet using my own 
capture software is that the data flow will sometimes stop for several 
seconds and then restart.  You don't actually lose any samples, but the 
client software can suffer a TCP timeout.  I got around this by setting 
the timeout in my software (using Perl's libtelnet module under Linux) 
to something like 30 seconds.  That solved the issue for me.


I've never seen the TCP timeout affect TimeLab; I don't know what kind 
of time limit John has coded but it seems to be long enough.


Re the 5120A itself erroring out, it does complain if either the 
frequency or the amplitude change too much during the measurement. 
Unfortunately, I don't know of a way to disable that.


John

On 5/3/2013 2:21 PM, Tom Knox wrote:

Recently I was looking at a co-workers Lab View measurements taken with a 
5120A. She is studying vibration on Phase Noise. I asked how Lab View compared 
to TimeLab and she said in many ways she prefers TimeLab but ran into a 
problem. The research she is currently conducting requires collecting data in 
extremely long measurement periods beyond the TimeLab limits. Is she missing 
something? If not is measurement length an easy fix? One other problem during 
her project was related to the 5120A, occasionally the vibration must hit a 
resonance and the 5120A stops measurement saying measurement  parameters were 
exceeded. It would be nice to be able to adjust or disable that feature. Thanks 
in advance for your ideas.

Thomas Knox



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Extended length Phase Noise Measurements with TimeLab

2013-05-03 Thread John Miles
As JohnA says there's no hard limit on the duration for phase data logging,
other than available RAM.  She would definitely want to use a 64-bit PC for
this.  The TSC 5120A cranks out phase data at 1000 samples per second by
default, so even a well-endowed 64 bit PC will run out of memory when trying
to record for  1 day.  Two things you can do about that:

- Use the 'set phaserate' command over a Telnet connection to lower the data
rate from the 5120A to 10 or 100 samples per second.  Change the 'Input Data
Rate' field in the TimeLab acquisition dialog accordingly.  This will work
only with newer 5120A firmware.

- Leave the TSC's phase data rate (and the Input Data Rate field) at 1000
samples/second and use the Output Decimation field to reduce the amount of
stored data.  

Both of those options should be used with TimeLab 1.1 or greater.  

For the issue where the 5120A stops the measurement, I'd log the signal with
a counter or a scope, maybe, to make sure that the signal isn't going even
farther out of range than expected.  If so, it would be best to fix that
problem at the source.  Otherwise, consider using a bandpass filter or PLL
whose bandwidth is wide enough to accommodate the desired analysis spectrum
but narrow enough to act as a flywheel during the (hopefully brief)
excursions that are confusing the 5120A. 

-- john, KE5FX
Miles Design LLC


 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of Tom Knox
 Sent: Friday, May 03, 2013 11:22 AM
 To: Time-Nuts
 Subject: [time-nuts] Extended length Phase Noise Measurements with TimeLab
 
 Recently I was looking at a co-workers Lab View measurements taken with a
 5120A. She is studying vibration on Phase Noise. I asked how Lab View
 compared to TimeLab and she said in many ways she prefers TimeLab but ran
 into a problem. The research she is currently conducting requires
collecting
 data in extremely long measurement periods beyond the TimeLab limits. Is
she
 missing something? If not is measurement length an easy fix? One other
 problem during her project was related to the 5120A, occasionally the
vibration
 must hit a resonance and the 5120A stops measurement saying measurement
 parameters were exceeded. It would be nice to be able to adjust or disable
that
 feature. Thanks in advance for your ideas.
 
 Thomas Knox
 

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Extended length Phase Noise Measurements with TimeLab

2013-05-03 Thread Bob Camp
Hi

There are some warning messages in TimeLab about runs with more than 
xx,xxx,xxx,xxx,xxx,xxx samples in them may harm poor fuzzy animals. I've 
routinely exceeded them with no obvious damage to the local environment. I may 
not have the wording of the error message quite right….

If the resonance on the vibration table is enough to push the frequency beyond 
a certain point, all of these boxes get a bit bothered. The 5120 is by no means 
unique in that regard. The trip points vary from model to model. The ones I 
seem to remember were in the 10's of ppm. 

Bob

On May 3, 2013, at 2:21 PM, Tom Knox act...@hotmail.com wrote:

 Recently I was looking at a co-workers Lab View measurements taken with a 
 5120A. She is studying vibration on Phase Noise. I asked how Lab View 
 compared to TimeLab and she said in many ways she prefers TimeLab but ran 
 into a problem. The research she is currently conducting requires collecting 
 data in extremely long measurement periods beyond the TimeLab limits. Is she 
 missing something? If not is measurement length an easy fix? One other 
 problem during her project was related to the 5120A, occasionally the 
 vibration must hit a resonance and the 5120A stops measurement saying 
 measurement  parameters were exceeded. It would be nice to be able to adjust 
 or disable that feature. Thanks in advance for your ideas.
 
 Thomas Knox
 
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Measuring Phase difference between different GPSDO

2013-05-03 Thread Mark C. Stephens
Actually Bob, What put me onto this was one of the Z3805A, the PU decreases 
dramatically when a cooling fan is pointed at it.

The Second one, is the opposite, positioned near the hot air exhaust of a 
server gets the best PU.

I have 3 other Z3805A which I have yet to play with.

Am I onto something or clutching at straws?


mark

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf 
Of Bob Camp
Sent: Saturday, 4 May 2013 3:01 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Measuring Phase difference between different GPSDO

Hi

Sounds like you are up and running. TimeLab can be set for any length run you 
wish. Three to five minutes is a common one. I have a 30 day run going at the 
moment. Even while the run is going, you can switch between displays to see 
what's happening. 

Unless you have some very unusual OCXO's in your 3805's fiddling with the oven 
temperature isn't going to help things much. If they are SC based, there may 
not be a turn over temperature. Often the set point is derived either from 
crystal data or from wide range temp run performance. 

Like any piece of surplus, you can indeed have something broken somewhere. In 
that case, repair is a good idea. Unless you blew the thermistor in the OCXO, 
it's an unlikely repair though.

Bob


On May 3, 2013, at 12:48 PM, Mark C. Stephens ma...@non-stop.com.au wrote:

 
 
 Am on the way to doing my first 'phase difference' with timelab - 
 takes an hour :)
 
 This is good stuff, everything went so smoothly, from installing the OS to 
 using Timelab.
 
 Now I have to figure out how to measure and set the xtal turning temperature 
 for this bunch of Z3805A's I have here.
 
 I have a Z3816A that is super stable. SYST:STAT? usually reports 0.0us 
 for PU, sometimes it will go out as far as 0.2us, but very rarely :p
 
 So I am planning on using that guy as my reference to tune the Z3805A's.
 
 How does that sound?
 
 
 Many thanks,
 mark
 
 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] 
 On Behalf Of Bob Camp
 Sent: Saturday, 4 May 2013 1:09 AM
 To: 'Discussion of precise time and frequency measurement'
 Subject: Re: [time-nuts] Measuring Phase difference between different 
 GPSDO
 
 Hi
 
 Grab an old garage sale category PC, blow the dust out of it and fire up the 
 GPIB. Once it's all working next steps:
 
 Check out TimeLab and EZGpib. Both are pretty good at talking to stuff.
 TimeLab does support the 5370, not sure about the other programs.
 
 If you have two GPSDO's with pps outputs, feed one into each channel on the 
 5370. Put it in start / stop mode and you should see the difference between 
 the two. Since they can be either ahead or behind each other things can get a 
 little messy. 
 
 Next, go to the software that runs your GPSDO and look for the cable delay
 setting. Adjust it on one GPSDO so it's always early (or late) by a couple 
 hundred nanoseconds. Now fiddle the cables to the counter so it reads a 
 couple hundred nanoseconds and not almost a full second :).
 
 At this point you are reading frequency to a very high resolution. The 
 GPSDO's likely are only good to a couple of nanoseconds at 1 second. The 
 counter should be good to ~0.02 ns.
 
 Run like this for a few minutes and see what you get.
 
 ---
 
 Now that you have conquered that approach it's time to tear it all apart and 
 try something new.
 
 Feed the 1 pps into one input of the counter (just like before). Hit the 
 other input with 10 MHz. You now will get a reading in the 0 to 100 ns range. 
 Each time a cycle slips past the 1 pps the reading will skip from 99.xxx to 
 0.xxx ns. That's called a phase wrap around. Click the appropriate boxes on 
 which ever software you are running and it should take care of the wrap.
 
 
 
 Either way the software should show you frequency, and frequency stability. 
 
 Bob
 
 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] 
 On Behalf Of Mark C. Stephens
 Sent: Friday, May 03, 2013 10:07 AM
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] Measuring Phase diference between different 
 GPSDO
 
 Hey Bob, I was thinking the exact same thought, and I found a NI PCI-GPIB 
 card in a PC with a dead HDD.
 These appear to be 5V PCI cards so I can't stick it in a nice rack mount 
 server with raid as they all seem to be 3.3v PCI.
 So I put a new HDD in the PC and I have stuck WinXP on it.
 Now I am awaiting further orders!  ;)
 
 mark
 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] 
 On Behalf Of Bob Camp
 Sent: Friday, 3 May 2013 9:21 PM
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] Measuring Phase diference between different 
 GPSDO
 
 Hi
 
 If you have a pile of GPIB cables, you may also have a GPIB card 

Re: [time-nuts] Extended length Phase Noise Measurements with TimeLab

2013-05-03 Thread John Miles
Yes, those messages are just there to warn users that if they try to record
multi-million point phase records, they may either run out of memory or
generate a gigabyte+ file that will be painfully slow to work with.  The
warning threshold is quite a bit higher in the 64-bit version but it will
still pop up if you get too aggressive with the data rate and/or duration.

-- john, KE5FX
Miles Design LLC

 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of Bob Camp
 Sent: Friday, May 03, 2013 1:46 PM
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] Extended length Phase Noise Measurements with
 TimeLab
 
 Hi
 
 There are some warning messages in TimeLab about runs with more than
 xx,xxx,xxx,xxx,xxx,xxx samples in them may harm poor fuzzy animals. I've
 routinely exceeded them with no obvious damage to the local environment. I
 may not have the wording of the error message quite right..
 
 If the resonance on the vibration table is enough to push the frequency
beyond
 a certain point, all of these boxes get a bit bothered. The 5120 is by no
means
 unique in that regard. The trip points vary from model to model. The ones
I
 seem to remember were in the 10's of ppm.
 
 Bob
 
 On May 3, 2013, at 2:21 PM, Tom Knox act...@hotmail.com wrote:
 
  Recently I was looking at a co-workers Lab View measurements taken with
a
 5120A. She is studying vibration on Phase Noise. I asked how Lab View
 compared to TimeLab and she said in many ways she prefers TimeLab but ran
 into a problem. The research she is currently conducting requires
collecting
 data in extremely long measurement periods beyond the TimeLab limits. Is
she
 missing something? If not is measurement length an easy fix? One other
 problem during her project was related to the 5120A, occasionally the
vibration
 must hit a resonance and the 5120A stops measurement saying measurement
 parameters were exceeded. It would be nice to be able to adjust or disable
that
 feature. Thanks in advance for your ideas.
 
  Thomas Knox
 
 
 
  ___
  time-nuts mailing list -- time-nuts@febo.com
  To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-
 nuts
  and follow the instructions there.
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-
 nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Extended length Phase Noise Measurements with TimeLab

2013-05-03 Thread Tom Van Baak
Tom,

For long measurement runs especially, I almost always do my data collection 
separately from my data analysis and plotting. That way there is no TimeLab CPU 
or RAM limit. You can start/stop/exit TimeLab or Stable32 or Plotter or Excel 
at will without worrying about losing raw data.

The data collection tool I use does a file checkpoint every minute, so if there 
is an app or OS or PC or cable or switch or TSC glitch or TSC crash or lab 
power failure I don't lose hours or days or weeks of data. It also means I can 
use old cheap PC's that run at 1% busy (no fan) to log tens of data streams 
instead of having to have an instance(s) of TimeLab peg my CPU and turn my 
laptop into a hair dryer.

I collect TSC 5110A and 5120A data from serial (1 Hz) and LAN (1 kHz) ports. I 
have never, ever, had problems with serial data. Serial ports may be old 
fashion but they are stateless and so more reliable for data logging than 
stateful USB or LAN connections.

I do have issues with long-term reliability of the TSC 5120 LAN port data 
stream. The TCP timeout bug was introduced in one of the very early firmware 
upgrades and no one has been able to track it down. I've spent hours snooping 
packets. The pre-production TSC5120/BSD firmware used to output 1 kHz phase 
samples evenly about every millisecond over ethernet. At some point before 
production the firmware got smart and buffered packets into chunks as large 
as 1024 or 5120 or 8192 samples each. My hunch is this optimization correlates 
with the rare TCP hangs that users see. I've tried multiple PC's and multiple 
OS's with the same result so I'm pretty sure it's the TSC firmware, and not me 
or my PC's or even TSC hardware to blame.

I can't help you with the measurement  parameters exceeded problem. I know 
the TSC boxes have limits on dynamic frequency variations. To debug, or at 
least to get around this, I would run a high-res frequency counter (like a 
53132A or CNT-90) in parallel with the TSC analyzer, since those counters have 
no dynamic range limits.

/tvb


- Original Message - 
From: Tom Knox act...@hotmail.com
To: Time-Nuts time-nuts@febo.com
Sent: Friday, May 03, 2013 11:21 AM
Subject: [time-nuts] Extended length Phase Noise Measurements with TimeLab


 Recently I was looking at a co-workers Lab View measurements taken with a 
 5120A. She is studying vibration on Phase Noise. I asked how Lab View 
 compared to TimeLab and she said in many ways she prefers TimeLab but ran 
 into a problem. The research she is currently conducting requires collecting 
 data in extremely long measurement periods beyond the TimeLab limits. Is she 
 missing something? If not is measurement length an easy fix? One other 
 problem during her project was related to the 5120A, occasionally the 
 vibration must hit a resonance and the 5120A stops measurement saying 
 measurement  parameters were exceeded. It would be nice to be able to adjust 
 or disable that feature. Thanks in advance for your ideas.
 
 Thomas Knox
 


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] hp5065b !!!

2013-05-03 Thread Tom Van Baak
Great big and small Rb cells:
http://leapsecond.com/corby/rb-lamp.htm
Also:
http://leapsecond.com/corby/5065a-optical/

/tvb

- Original Message - 
From: Bob Camp li...@rtty.us
To: Discussion of precise time and frequency measurement time-nuts@febo.com
Sent: Friday, May 03, 2013 10:07 AM
Subject: Re: [time-nuts] hp5065b !!!


Hi

One of the unique things about the 5065 is it's great big cells. The small cell 
compact Rb's are a very different beast. There is a body of evidence 
suggesting that the cell size is significant for ultra low ADEV.

The basic concepts of improving signal (light intensity) and signal to noise 
(filtering) would apply to other units. Weather the mods give significant 
improvements on small cells is the open question. The compact Rb's may already 
be at the floor for their cell size….

Bob

On May 3, 2013, at 10:40 AM, ed breya e...@telight.com wrote:

 I'm interested in these improvements too, and hope some can be applied to 
 other models. I don't have a 5065A, nor do I expect to ever get one, but I do 
 have three identical old Efratom units.
 
 Whether it's worthwhile to modify any Rb units also depends on whether it's 
 possible to rejuvenate their Rb lamps, as discussed here a number of times, 
 so that they can be maintained virtually indefinitely.
 
 Ed


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] 5V antenna on 3V receiver via dist. amp?

2013-05-03 Thread paul swed
Thats fine
If the 3 v rcvr complains connect a 450 ohm from the antenna to ground that
should make it happy. The resistor really isn't picky


On Thu, May 2, 2013 at 10:38 PM, Paul tic-...@bodosom.net wrote:

 On Thu, May 2, 2013 at 5:51 PM,  time-nuts-requ...@febo.com wrote:
  A distribution amplifier (like the 58516A) should also be OK. If you are
  going to have a GPS receiver that can provide 5V to the antenna and
  amplifier, that would be ideal.

 Yes, my plan was to attach a 5V receiver to power in (port 1 on a
 58516A) and 3V receivers to the DC-blocked ports.
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Measuring Phase diference between different GPSDO

2013-05-03 Thread Al Wolfe
How about just running the two different 10 mhz outputs into a doubly 
balanced mixer and monitoring the DC output with a volt meter from the IF 
port of the mixer? Sure beats $20,000 for a fancy HP. Of course this is a 
very simple method and is boring as heck to sit and watch a voltmeter but it 
could easily be automated.


Suitable doubly balanced mixers can be often found on eBay for under $50. I 
found a nice Mini-Circuit unit with BNC connectors for around $20 that I 
have often used.


I often just use a scope and Lissajous figures for looking at phase 
differences.


Al, K9SI



Now the trick is:
How to use the 5370B to measure the phase difference.
Anyone care to point me towards an APP note or two?!

mark


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Measuring Phase difference between different GPSDO

2013-05-03 Thread Bob Camp
Hi

OCXO's don't really like major temperature changes. They are much happier in 
steady state environments. Transient behavior may or may not correlate with 
steady state behavior, so trying to correct it by fiddling the oven - not a 
good idea. The one exception to that would be if the oven has burned out. In 
that case the EFC voltage should be pegged. 

---

Better way to start:

Fire up the software of your choice on the PC. Let it talk to the Z3805's make 
sure that it's doing what it should (locking up to GPS etc). I'm guessing 
that's already taken care of. 

Let it run for a while and watch the EFC voltage. It'll start out moving fairly 
quickly. After a few days / weeks it should settle out. It should be near 
center scale after settling. 

While the OCXO is settling out, you can watch the PPS output and the 10 MHz. 
The 10 MHz should be pretty close after the first day or two. In this case 
pretty close is  0.1 Hz on the 10 MHz. Pretty close on the pps should be  
20 ns. If they are not, the disciplining part of the unit is broke.

Depending on what you do / don't find there are lots of different directions to 
head. 

Bob

On May 3, 2013, at 5:04 PM, Mark C. Stephens ma...@non-stop.com.au wrote:

 Actually Bob, What put me onto this was one of the Z3805A, the PU decreases 
 dramatically when a cooling fan is pointed at it.
 
 The Second one, is the opposite, positioned near the hot air exhaust of a 
 server gets the best PU.
 
 I have 3 other Z3805A which I have yet to play with.
 
 Am I onto something or clutching at straws?
 
 
 mark
 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On 
 Behalf Of Bob Camp
 Sent: Saturday, 4 May 2013 3:01 AM
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] Measuring Phase difference between different GPSDO
 
 Hi
 
 Sounds like you are up and running. TimeLab can be set for any length run you 
 wish. Three to five minutes is a common one. I have a 30 day run going at the 
 moment. Even while the run is going, you can switch between displays to see 
 what's happening. 
 
 Unless you have some very unusual OCXO's in your 3805's fiddling with the 
 oven temperature isn't going to help things much. If they are SC based, there 
 may not be a turn over temperature. Often the set point is derived either 
 from crystal data or from wide range temp run performance. 
 
 Like any piece of surplus, you can indeed have something broken somewhere. In 
 that case, repair is a good idea. Unless you blew the thermistor in the OCXO, 
 it's an unlikely repair though.
 
 Bob
 
 
 On May 3, 2013, at 12:48 PM, Mark C. Stephens ma...@non-stop.com.au wrote:
 
 
 
 Am on the way to doing my first 'phase difference' with timelab - 
 takes an hour :)
 
 This is good stuff, everything went so smoothly, from installing the OS to 
 using Timelab.
 
 Now I have to figure out how to measure and set the xtal turning temperature 
 for this bunch of Z3805A's I have here.
 
 I have a Z3816A that is super stable. SYST:STAT? usually reports 0.0us 
 for PU, sometimes it will go out as far as 0.2us, but very rarely :p
 
 So I am planning on using that guy as my reference to tune the Z3805A's.
 
 How does that sound?
 
 
 Many thanks,
 mark
 
 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] 
 On Behalf Of Bob Camp
 Sent: Saturday, 4 May 2013 1:09 AM
 To: 'Discussion of precise time and frequency measurement'
 Subject: Re: [time-nuts] Measuring Phase difference between different 
 GPSDO
 
 Hi
 
 Grab an old garage sale category PC, blow the dust out of it and fire up the 
 GPIB. Once it's all working next steps:
 
 Check out TimeLab and EZGpib. Both are pretty good at talking to stuff.
 TimeLab does support the 5370, not sure about the other programs.
 
 If you have two GPSDO's with pps outputs, feed one into each channel on the 
 5370. Put it in start / stop mode and you should see the difference between 
 the two. Since they can be either ahead or behind each other things can get 
 a little messy. 
 
 Next, go to the software that runs your GPSDO and look for the cable delay
 setting. Adjust it on one GPSDO so it's always early (or late) by a couple 
 hundred nanoseconds. Now fiddle the cables to the counter so it reads a 
 couple hundred nanoseconds and not almost a full second :).
 
 At this point you are reading frequency to a very high resolution. The 
 GPSDO's likely are only good to a couple of nanoseconds at 1 second. The 
 counter should be good to ~0.02 ns.
 
 Run like this for a few minutes and see what you get.
 
 ---
 
 Now that you have conquered that approach it's time to tear it all apart and 
 try something new.
 
 Feed the 1 pps into one input of the counter (just like before). Hit the 
 other input with 10 MHz. You now will get a reading in the 0 to 100 ns 
 range. Each time a cycle slips past the 1 pps the reading will