 I guess it depends on signal to noise ratio. With reciprocal counters, you
 only need one period to measure as acurately as you need, but to have good
 acuracy, you need very good S/N, as there is no filtering possible. 
 
 For example, the HP 5370 can measure a single period of a signal with a
 resolution of 20pS (excluding noise and trigger imperfections), so excluding
 these errors, the HP 5370 could measure a single period of a ~3.5 MHz signal
 with 7 x10-5 precision (if I have not goofed the calculations) More
 periods improve the resolution proportionately to the quare root. Accuracy
 is another matter.
 
 Didier KO4BB

The jitter on a single period is likely very, very high, especially
if it comes over the air. That's why one usually measures over
a duration of thousands or even millions of periods (effectively
called the gate time).

The HP 53132A makes something like 200,000 measurements
per second. As a result, for a certain range of frequencies, it
claims 12 digits/sec of resolution (vs. HP 5370 ~11 digits/sec).

/tvb


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FMT on October 13

2007-09-26 Thread Hal Murray
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

 Did mounting it on a block of foam help?

 That is a rather bad solution. You want much softer material to react
 to quicker things, such as silicon rubber. Also, that would only be a
 12 dB/Oct solution. You would really like a few more poles there. The
 trick is to add weight to the calculation. So you want a very soft
 material, holding a thick block (lead) and from this base suspend the
 oscillator through a soft material again. Now you have a 24 dB/Oct
 solution. The trouble you now will have is that the wires will be
 another shock/vibration transport mechanism. They would need to be
 connected to the middle-frame such that outer forces hit the middle
 weigth and not directly on the sensitive part. They would need to be
 soft and arranged is such a way that they do not push or pull the
 inner end, but is allowed to flex alot. 

Yes, but inserting a chunk of foam is a lot easier than finding a block of 
lead.  It's likely to be good enough.  (Make that good enough for most 
application.  This is time-nuts.  Nothing is good-enough that somebody won't 
suggest something better/nuttier.  :)

Packing bubbles might work too.  I'm thinking of the sheets of bubble wrap 
that are fun to snap rather than the foam peanuts that get all over the place 
and are really nasty if you have a slight static charge.

Does anybody have any data on the sensitivity of a crystal oscillator vs 
frequency of mechanical shock/vibration?  Does it scale with amplitude or 
acceleration or ???





-- 
These are my opinions, not necessarily my employer's.  I hate spam.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FMT on October 13

2007-09-26 Thread Magnus Danielson
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

From: Tom Van Baak [EMAIL PROTECTED]
Subject: Re: [time-nuts] FMT on October 13
Date: Tue, 25 Sep 2007 23:45:12 -0700
Message-ID: [EMAIL PROTECTED]

 ); SAEximRunCond expanded to false
 Errors-To: [EMAIL PROTECTED] RETRY
 
  I guess it depends on signal to noise ratio. With reciprocal counters, you
  only need one period to measure as acurately as you need, but to have good
  acuracy, you need very good S/N, as there is no filtering possible. 
  
  For example, the HP 5370 can measure a single period of a signal with a
  resolution of 20pS (excluding noise and trigger imperfections), so excluding
  these errors, the HP 5370 could measure a single period of a ~3.5 MHz signal
  with 7 x10-5 precision (if I have not goofed the calculations) More
  periods improve the resolution proportionately to the quare root. Accuracy
  is another matter.
  
  Didier KO4BB
 
 The jitter on a single period is likely very, very high, especially
 if it comes over the air. That's why one usually measures over
 a duration of thousands or even millions of periods (effectively
 called the gate time).
 
 The HP 53132A makes something like 200,000 measurements
 per second. As a result, for a certain range of frequencies, it
 claims 12 digits/sec of resolution (vs. HP 5370 ~11 digits/sec).

As was discussed recently, didn't they do averaging such that they updated
value every second but the raw singel-shot resolution doess not give you the
12 digits/sec. There was a nice explanation in an article on how this was
not improving say ADEV measurements in the end.

Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FMT on October 13

2007-09-26 Thread Didier Juges
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

I just wanted to point out that with reciprocal counters, you can get
resolution much better than the 1Hz/s you would get with conventional
frequency counters, even though the actual accuracy of the measurement may
be way off.

The original question seemed to imply that with a short transmission time,
you could not guarantee a frequency accuracy of 1e-6 Hz, which you probably
can't anyhow, but the limit is not the resolution of the instrument or the
measurement method. 

I do not know how far off calibration my HP 5370s are, but the 20pS
resolution is at best only usable under some circumstances that I have not
isolated yet, due to jitter.

When measuring a 3.5 MHz signal (@1dBm) from my HP 8657B through 1 meter of
good coax cable (with counter and generator phase locked to the Thunderbolt
GPSDO) in Frequency mode with a 1s gate time, the resolution is 1e-5Hz, with
about 1e-3Hz p-p variation. When measuring over 1 period with 10,000 periods
sample size, the resolution is only 1e-1Hz with a standard deviation of ~400
Hz (or about 0.1%). Of course, over the air, it will be much worse due to
noise, let alone propagation, fading and multipath.

Didier KO4BB

 -Original Message-
 From: [EMAIL PROTECTED] 
 [mailto:[EMAIL PROTECTED] On Behalf Of Tom Van Baak
 Sent: Wednesday, September 26, 2007 1:45 AM
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] FMT on October 13
 
  I guess it depends on signal to noise ratio. With 
 reciprocal counters, 
  you only need one period to measure as acurately as you 
 need, but to 
  have good acuracy, you need very good S/N, as there is no 
 filtering possible.
  
  For example, the HP 5370 can measure a single period of a 
 signal with 
  a resolution of 20pS (excluding noise and trigger 
 imperfections), so 
  excluding these errors, the HP 5370 could measure a single 
 period of a 
  ~3.5 MHz signal with 7 x10-5 precision (if I have not goofed the 
  calculations) More periods improve the resolution 
 proportionately 
  to the quare root. Accuracy is another matter.
  
  Didier KO4BB
 
 The jitter on a single period is likely very, very high, 
 especially if it comes over the air. That's why one usually 
 measures over a duration of thousands or even millions of 
 periods (effectively called the gate time).
 
 The HP 53132A makes something like 200,000 measurements per 
 second. As a result, for a certain range of frequencies, it 
 claims 12 digits/sec of resolution (vs. HP 5370 ~11 digits/sec).
 
 /tvb
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com To unsubscribe, 
 go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FMT on October 13

2007-09-26 Thread Magnus Danielson
From: Didier Juges [EMAIL PROTECTED]
Subject: Re: [time-nuts] FMT on October 13
Date: Wed, 26 Sep 2007 05:57:14 -0500
Message-ID: [EMAIL PROTECTED]

Didier,

 I just wanted to point out that with reciprocal counters, you can get
 resolution much better than the 1Hz/s you would get with conventional
 frequency counters, even though the actual accuracy of the measurement may
 be way off.

These days conventional counteras are reciprocal counters. It is only the
old-school counters which is not reciprocal. Nothing wrong with old-school,
but a conventional counter of the shelf today is probably a reciprocal jobbie.

 The original question seemed to imply that with a short transmission time,
 you could not guarantee a frequency accuracy of 1e-6 Hz, which you probably
 can't anyhow, but the limit is not the resolution of the instrument or the
 measurement method. 
 
 I do not know how far off calibration my HP 5370s are, but the 20pS
 resolution is at best only usable under some circumstances that I have not
 isolated yet, due to jitter.
 
 When measuring a 3.5 MHz signal (@1dBm) from my HP 8657B through 1 meter of
 good coax cable (with counter and generator phase locked to the Thunderbolt
 GPSDO) in Frequency mode with a 1s gate time, the resolution is 1e-5Hz, with
 about 1e-3Hz p-p variation. When measuring over 1 period with 10,000 periods
 sample size, the resolution is only 1e-1Hz with a standard deviation of ~400
 Hz (or about 0.1%). Of course, over the air, it will be much worse due to
 noise, let alone propagation, fading and multipath.

When measuring over a longer period you see a different spot on the ADEV/MDEV
curve. Chances are that you are more unstable there for an OCXO. Both linear
and noise products will make things harder. It can be a challenge to separate
the drift rate due to signal path shifts and that of the OCXO.

Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FMT on October 13

2007-09-25 Thread John Ackermann N8UR
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

Tom Van Baak said the following on 09/24/2007 08:57 PM:

 I would think this is especially true for non-local frequencies,
 such as one received over the air. I'll leave it to you FMT
 guys to comment on the magnitude of degradation due to
 transmission and reception noise.

Absolutely.  Probably the best real-world performance you can get with a
skywave signal is on the order of 0.01 Hz.  Propagation effects play
havoc, but the longer the averaging period, the more short-term effects
will average away.  One of the reasons for the fairly long transmission
periods is to both allow longer averaging, but also provide the
opportunity to observe the atmospheric conditions at work.

 While were at it, in the case mentioned above I'm a curious
 about their FMT frequency standard -- if it's really accurate
 to parts in 10^12, as they imply, over 10 minutes. I could
 believe this if it were an Rb or Cs-based GPSDO.

We're using an Austron 1250A OXCO that's been measured as better than
9x10e-13 for averaging times of 1 second out to 1000 seconds; over a
broader range, it's better than 3x10e-12 from 0.1 seconds to 40,000 seconds.

Now, an important point -- we're not trying to trim the Austron to be
precisely on frequency.  We're going to let it run at whatever offset it
happens to be.  That will help make sure that the signal doesn't have
lots of zero's at the end, even though the resolution of the
synthesizers driving the transmitters is limited to 0.1 Hz.

We'll be comparing the Austron against a Z3801a (via my TSC-5120A
analyzer) and logging the frequency difference for at least several
hours prior to the test until several hours following.  The TSC gives 16
digits over 1000 seconds; depending on how much jitter we see, we'll
probably throw away the last two or three.  Even though the Z3801A may
be wandering around a bit, with successive 1000 second measurements we
should have confidence in the actual frequency over 1000 second periods
to at least parts in the 12s, ultimately limited by the Austron's
stability.  But since that's known to be in the 13s over the averaging
period of interest, we think we're safe in claiming accuracy and
stability of parts in the 12s.

Tom, if I'm missing something in this analysis, I'm seriously open to
education...

By the way -- the synthesizers used to drive the transmitter amplifiers
will be PTS 250 SX-51 low noise units, so hopefully the transmitted
signals will have a better-than-the-average-ham-rig phase noise.  The
synthesizers will directly feed the driver and final amplifier stages of
some vintage Kenwood TS-520 ham transceivers with no other mixing --
it'll purely be the synthesizer and a transistor buffer amp driving two
vacuum tube stages to get up to about 75 watts (the rigs can run 100
watts, but we're derating -- and adding fans -- to support the long
key-down times).

John


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FMT on October 13

2007-09-25 Thread John Ackermann N8UR
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

Didier Juges said the following on 09/24/2007 09:40 PM:
 ); SAEximRunCond expanded to false
 Errors-To: [EMAIL PROTECTED] RETRY
 
 I guess it depends on signal to noise ratio. With reciprocal counters, you
 only need one period to measure as acurately as you need, but to have good
 acuracy, you need very good S/N, as there is no filtering possible. 
 
 For example, the HP 5370 can measure a single period of a signal with a
 resolution of 20pS (excluding noise and trigger imperfections), so excluding
 these errors, the HP 5370 could measure a single period of a ~3.5 MHz signal
 with 7 x10-5 precision (if I have not goofed the calculations) More
 periods improve the resolution proportionately to the quare root. Accuracy
 is another matter.

I did some measurements on the frequency counter capability of my 5370B
some time ago, and found that the performance wasn't as good as in time
interval mode.

But it's still not bad -- the internal noise floor was 4x10e-11 for 1
second (using the 1 second gate time).  See
http://www.febo.com/time-freq/hardware/5370B/index.html

John


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FMT on October 13

2007-09-25 Thread Tom Van Baak
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

 While were at it, in the case mentioned above I'm a curious
 about their FMT frequency standard -- if it's really accurate
 to parts in 10^12, as they imply, over 10 minutes. I could
 believe this if it were an Rb or Cs-based GPSDO.
 
 We're using an Austron 1250A OXCO that's been measured as better than
 9x10e-13 for averaging times of 1 second out to 1000 seconds; over a
 broader range, it's better than 3x10e-12 from 0.1 seconds to 40,000 seconds.

Ah, if they is you, then I have no more worries. Yes, using
that free-running 1250A is the perfect solution; much better
than using the output of a GPSDO.


 Now, an important point -- we're not trying to trim the Austron to be
 precisely on frequency.  We're going to let it run at whatever offset it
 happens to be.  That will help make sure that the signal doesn't have
 lots of zero's at the end, even though the resolution of the
 synthesizers driving the transmitters is limited to 0.1 Hz.

Clever.


 We'll be comparing the Austron against a Z3801a (via my TSC-5120A
 analyzer) and logging the frequency difference for at least several
 hours prior to the test until several hours following.  The TSC gives 16
 digits over 1000 seconds; depending on how much jitter we see, we'll
 probably throw away the last two or three.  Even though the Z3801A may
 be wandering around a bit, with successive 1000 second measurements we
 should have confidence in the actual frequency over 1000 second periods
 to at least parts in the 12s, ultimately limited by the Austron's
 stability.  But since that's known to be in the 13s over the averaging
 period of interest, we think we're safe in claiming accuracy and
 stability of parts in the 12s.

Yes, running the measurement for hours before and after is
the right thing to do. All sounds good.

Make sure not to get near any of the equipment. Free-running
oscillators are sensitive to vibration or shock. You've probably
heard the story of my best Sulzer oscillator making small phase
or jumps which I eventually correlated to when the kids flushed
the toilet down the hall.


 Tom, if I'm missing something in this analysis, I'm seriously open to
 education...

Nothing missing; you nailed it.

/tvb


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FMT on October 13

2007-09-25 Thread Hal Murray
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY


 Make sure not to get near any of the equipment. Free-running
 oscillators are sensitive to vibration or shock. You've probably heard
 the story of my best Sulzer oscillator making small phase or jumps
 which I eventually correlated to when the kids flushed the toilet down
 the hall. 

Did mounting it on a block of foam help?


-- 
These are my opinions, not necessarily my employer's.  I hate spam.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] FMT on October 13

2007-09-24 Thread John Ackermann N8UR
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

The Midwest VHF/UHF Society (located in Southwest Ohio) is pleased to
announce that the first annual MVUS Frequency Measuring Test will be
held on Saturday, October 13, 2007.  There will be two transmission
periods: the first at 14:30 EDT (1830 UTC), and the second at 21:30 EDT
(0130 UTC Sunday).  Transmissions will be on the 80M, 40M, and 30M
amateur bands from Dayton, Ohio under the callsign W8KSE.

Frequencies will be approximately: 3555 kHz, 7055 kHz, and 10115 kHz,
but be prepared to tune as we will adjust to minimize QRM.  The
transmitters will be running about 75 watts output into wire antennas
for each band.  All the transmitters will be driven from a common
frequency standard.

We will transmit on all three bands simultaneously.  Plans are to
transmit two 10 minute test periods, and a third if the transmitters
aren't melting by that point.  The frequency will be changed by a small
amount (less than 200 Hz) between transmission periods.  So, a complete
submission will include two or three separate measurements for each band.

Our goal is to transmit a signal known in frequency to parts in 10e-12
(i.e., less than 0.0001 Hz error at 10 MHz) and stable to a similar
level during the course of the transmission.  Frequencies will be
measured at the transmitter site with a system capable of microHertz
resolution referenced to a GPS disciplined oscillator, and will also be
monitored by another station in groundwave range that can measure the
frequencies with similar accuracy.

The MVUS Frequency Measuring Test is intended to supplement, not
replace, the ARRL FMT.

Further information, including approximate transmission frequencies,
will be posted at http://www.febo.com/time-freq/FMT.  You can also send
email with questions or comments (or, after the test, your results!) to
[EMAIL PROTECTED].

For discussion about off-air frequency measurement, we suggest you check
out the FMT-nuts mailing list, sponsored by Connie Marshall, K5CM.  For
details, go to http://tech.groups.yahoo.com/group/FMT-nuts/.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FMT on October 13

2007-09-24 Thread Hal Murray

 Plans are to transmit two 10 minute test periods, and a third if the
 transmitters aren't melting by that point. 

 Our goal is to transmit a signal known in frequency to parts in 10e-12
 (i.e., less than 0.0001 Hz error at 10 MHz) and stable to a similar
 level during the course of the transmission.  Frequencies will be
 measured at the transmitter site with a system capable of microHertz
 resolution referenced to a GPS disciplined oscillator, and will also
 be monitored by another station in groundwave range that can measure
 the frequencies with similar accuracy. 

Suppose I have a pile of good lab gear, and it gets N seconds of signal.

How accurately can it measure the frequency?


-- 
These are my opinions, not necessarily my employer's.  I hate spam.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FMT on October 13

2007-09-24 Thread Tom Van Baak
 Plans are to transmit two 10 minute test periods, and a third if the
 transmitters aren't melting by that point. 
 
 Our goal is to transmit a signal known in frequency to parts in 10e-12
 (i.e., less than 0.0001 Hz error at 10 MHz) and stable to a similar
 level during the course of the transmission.  Frequencies will be
 measured at the transmitter site with a system capable of microHertz
 resolution referenced to a GPS disciplined oscillator, and will also
 be monitored by another station in groundwave range that can measure
 the frequencies with similar accuracy. 
 
 Suppose I have a pile of good lab gear, and it gets N seconds of signal.
 
 How accurately can it measure the frequency?

Hi Hal,

If you have a low noise CW signal, a cheap, legacy 1 ns
resolution counter will give you 9 digits of resolution per
second. So to measure to parts in 10^12 requires gate
times on the order of a thousand seconds.

A fancier, modern counter like a HP 53132A is almost ten
times better than that so 100 s gate times are all you need
for 12 digits. Further, if it's an oddball frequency (i.e., not
a nice multiple or fraction of 10 MHz) even 10 second gate
times are sufficient with this counter (it does clever CW
oversampling inside).

For extreme counters like HP 5370 or SR 620 with resolution
well under 50 ps you can measure any frequency to 12 digits
in a matter of tens of seconds.

The main problems at this level are often that neither your
frequency reference nor the frequency you are measuring
are stable to parts in 10^12th. So the measurements you
get will contain the sum of noise in both sources and the
counter itself. And this noise is often well above parts in
10^12th. It takes time, statistics, or other tests to determine
the noise contribution of each.

I would think this is especially true for non-local frequencies,
such as one received over the air. I'll leave it to you FMT
guys to comment on the magnitude of degradation due to
transmission and reception noise.

While were at it, in the case mentioned above I'm a curious
about their FMT frequency standard -- if it's really accurate
to parts in 10^12, as they imply, over 10 minutes. I could
believe this if it were an Rb or Cs-based GPSDO.

Usually the accuracy of GPS disciplined oscillators are spec'd
for averaging times over a day. And at one day, parts in 10^12
is very easy (many are down in the low 13's or 14's). But over
a short span like 10 minutes most quartz-based GPSDO wander
in frequency by many parts in 10^11. See, for example, these
two nice quartz GPSDO over 10 minutes and note the scale is
1e-11 per *division*; which is almost 1e-10 full-scale.

http://www.leapsecond.com/pages/fury/#6

/tvb


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FMT on October 13

2007-09-24 Thread Didier Juges
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

I guess it depends on signal to noise ratio. With reciprocal counters, you
only need one period to measure as acurately as you need, but to have good
acuracy, you need very good S/N, as there is no filtering possible. 

For example, the HP 5370 can measure a single period of a signal with a
resolution of 20pS (excluding noise and trigger imperfections), so excluding
these errors, the HP 5370 could measure a single period of a ~3.5 MHz signal
with 7 x10-5 precision (if I have not goofed the calculations) More
periods improve the resolution proportionately to the quare root. Accuracy
is another matter.

Didier KO4BB


 -Original Message-
 From: [EMAIL PROTECTED] 
 [mailto:[EMAIL PROTECTED] On Behalf Of Hal Murray
 Sent: Monday, September 24, 2007 5:26 PM
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] FMT on October 13
 
 
  Plans are to transmit two 10 minute test periods, and a 
 third if the 
  transmitters aren't melting by that point.
 
  Our goal is to transmit a signal known in frequency to 
 parts in 10e-12 
  (i.e., less than 0.0001 Hz error at 10 MHz) and stable to a similar 
  level during the course of the transmission.  Frequencies will be 
  measured at the transmitter site with a system capable of 
 microHertz 
  resolution referenced to a GPS disciplined oscillator, and 
 will also 
  be monitored by another station in groundwave range that 
 can measure 
  the frequencies with similar accuracy.
 
 Suppose I have a pile of good lab gear, and it gets N seconds 
 of signal.
 
 How accurately can it measure the frequency?
 
 
 --
 These are my opinions, not necessarily my employer's.  I hate spam.
 
 
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FMT on October 13

2007-09-24 Thread Bruce Griffiths
Didier Juges wrote:
 I guess it depends on signal to noise ratio. With reciprocal counters, you
 only need one period to measure as acurately as you need, but to have good
 acuracy, you need very good S/N, as there is no filtering possible. 

 For example, the HP 5370 can measure a single period of a signal with a
 resolution of 20pS (excluding noise and trigger imperfections), so excluding
 these errors, the HP 5370 could measure a single period of a ~3.5 MHz signal
 with 7 x10-5 precision (if I have not goofed the calculations) More
 periods improve the resolution proportionately to the quare root. Accuracy
 is another matter.

 Didier KO4BB
   
Didier

With band limited gaussian noise and an SNR of 40dB the rms error in
measuring the period of a single cycle is about 0.3%

Bruce

___