[time-nuts] HP 5345A Noise Floor

2020-10-30 Thread Giorgio Barinetti
Hi Bob,



The current setup is:



Counter locked ( 10 mhz in ) to the Cesium.



Same 10 mhz signal from the Cesium feeds both the inputs and counter is running 
in “phase difference” so time A-B.



That means the resolution will be hiher, I guess, since the counter is counting 
the “difference” between the two

Signals, not the whole 10 mhz.



I’m a complete dumdum in maths but I’m sure that somebody here can explain in a 
simple way.





Regards,

Giorgio.









Hi



According to page 1-3 in https://bama.edebris.com/download/hp/5345a/5345a_6.pdf 
<https://bama.edebris.com/download/hp/5345a/5345a_6.pdf>



The 5345 is a 2 ns counter. Roughly speaking, that normally gets you around  
2x10^-9

at 1 second.



That would suggest there is something a bit off in your setup.



Bob



> On Oct 29, 2020, at 9:26 AM, Giorgio Barinetti  barinetti.it<http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com>>
>  wrote:

>

> Hi All.

>

> While I’m waiting for a 5370B to come ( find it cheap and beautiful from 
> France ) I’ve put my old 5345A on a trial, connecting both the

> Inputs to the 10 mhz out from an Oscilloquartz 3200 Cesium Oscillator.

>

> Those are the results. Please advise 

>

> Regards,

> Giorgio.

> [cid:image001.gif at 
> 01D6ADF5.5EDF8CF0<http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com>]

> ___

> time-nuts mailing list -- time-nuts at 
> lists.febo.com<http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com>

> To unsubscribe, go to 
> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com

> and follow the instructions there.

___
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


[time-nuts] HP 5345A Noise Floor

2020-10-29 Thread Giorgio Barinetti
Hi All.

While I’m waiting for a 5370B to come ( find it cheap and beautiful from France 
) I’ve put my old 5345A on a trial, connecting both the
Inputs to the 10 mhz out from an Oscilloquartz 3200 Cesium Oscillator.

Those are the results. Please advise 

Regards,
Giorgio.
[cid:image001.gif@01D6ADF5.5EDF8CF0]
___
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


[time-nuts] Frequency Counter Choice

2020-10-23 Thread Giorgio Barinetti
Hi Attila,



Thanks for the very nice job. I’ll sharpen out my math and try to read it 



Regards,

Giorgio.



On Thu, 22 Oct 2020 13:29:07 -0400

Bob kb8tq http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com>> wrote:



> There is a paper running around behind the IEEE paywall by Collins that 
> details

> one approach to designing multi-stage limiters. There are other ways to do it.

> The key is to not go to crazy all in one stage.





You can find a modern, more accurate version of Collins result

in [1]. While most formulas in there give the noise up to some

proportionality factor, all the information to calculate the noise

level is in there. The big difference to Collins here is that this

result is exact if you use all harmonics up to infinity (which is

easy) or can break up with a known error if you don't want to go

that far. Collins uses a trapezoid approximation throughout and thus

over-estimates the noise contribution of low-gain stages, which

are the first stages thus contribute to a relatively large error

in the calculation.







   Attila Kinali



[1] "A Physical Sine-to-Square Converter Noise Model"

by yours truly, IFCS 2018

http://people.mpi-inf.mpg.de/~adogan/pubs/IFCS2018_comparator_noise.pdf



--

   The bad part of Zurich is where the degenerates

throw DARK chocolate at you.

___
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


[time-nuts] Frequency Counter Choice

2020-10-22 Thread Giorgio Barinetti
Hi Bob,



Nice explanation. I'll see what I can do to find a 53132.



Maybe, in the meantime, if I'll be lucky to find a 5370, I'll play on it.



Regarding the mixer based setup, I'm already in progress for setting up a 
suitable test jig.

Please point to some information, if you have it handy.



Regards,

Giorgio.





Hi



> On Oct 22, 2020, at 7:50 AM, Giorgio Barinetti  barinetti.it<http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com>>
>  wrote:

>

> Hi,

>

> As a newbie in the field, I've collected, by chance, some frequency standards.

>

> Now is the time to measure them, and see how they perform.

>

> I've inherited a 5371, but something tells me that is not the right 
> instrument - and maybe is even faulty.

>

> So, I'm in the need to buy a counter to use togheter with TimeLab.

>

> Choices are many, but I'll try to avoid the "older" machines lile 5370 or 
> 5335. The 531xx series seems nice ( money apart )

> But again : which one between the 3 ? 53131, 53132 or 53181 ?



53181 = single channel 400 ps counter

53131 = dual channel 400 ps counter

53132 = dual channel 100 ps counter



5370 (when working) = dual channel 20 ps counter



The gotcha is that none of them are good enough to properly measure 1

second ADEV on a good ( but still could be cheap ) OCXO. A mixer based

setup is a cheap way to get things done, even with a 5335.



The 531xx counters all share a couple of issues:



1) The power supply was made by who knows who and their quality is not

the best. Good news is that the power supply out of a (cheap) 53181 will

swap over to revive an expensive 53132.



2) They have a fan, it plugs with dust. When that happens it gets hot inside

the cabinet. Things (like the power supply) die as a result.



3) They have a VFD display. It's a nice one, but they do wear out. Replacing

them is problematic.



4) They have a fancy setup to add digits to a frequency measurement. It

can lead you astray. It also tends to go deaf right at 10 MHz. There are app

notes out there that explain the details.



5) With *any* counter, a good external reference is the way to go for "Time Nut"

grade measurements.



None of that is to say they are a bad counter, far from it. I have a number

of them and have been using them here and at work for decades. The 532xx

counters are the latest and greatest. They are on eBay, but not cheap.



Bob







>

> Can somebody shed some light, and maybe help even to found a baseline for us 
> beginners ?

>

> Regards,

> IZ2JGB

> Giorgio

>

> My NTP servers: (ntpd on FreeBSD - PPS in via Serial)

>

> HP/Symmetricom 55300A  https://www.ntppool.org/scores/93.41.196.243

>

> Efratom Rb/Xc GPSDO https://www.ntppool.org/scores/95.255.136.126

>

>

>

> ___

> time-nuts mailing list -- time-nuts at 
> lists.febo.com<http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com>

> To unsubscribe, go to 
> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com

> and follow the instructions there.

___
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


[time-nuts] Frequency Counter Choice

2020-10-22 Thread Giorgio Barinetti
Hi,

As a newbie in the field, I've collected, by chance, some frequency standards.

Now is the time to measure them, and see how they perform.

I've inherited a 5371, but something tells me that is not the right instrument 
- and maybe is even faulty.

So, I'm in the need to buy a counter to use togheter with TimeLab.

Choices are many, but I'll try to avoid the "older" machines lile 5370 or 5335. 
The 531xx series seems nice ( money apart )
But again : which one between the 3 ? 53131, 53132 or 53181 ?

Can somebody shed some light, and maybe help even to found a baseline for us 
beginners ?

Regards,
IZ2JGB
Giorgio

My NTP servers: (ntpd on FreeBSD - PPS in via Serial)

HP/Symmetricom 55300A  https://www.ntppool.org/scores/93.41.196.243

Efratom Rb/Xc GPSDO https://www.ntppool.org/scores/95.255.136.126



___
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


[time-nuts] HP 10811-60164 for sale

2019-11-27 Thread Giorgio Barinetti
Hello,

As per subject, i'm selling some quartz oscillator, togheter with the HP 55429B 
card, salvaged from a telco clocking system.

6pc avail.

For inquiries, my email or +393402123322 (english / italian speaking )

Asking 110€ + shipping


Thanks !


Regards
IZ2JGB
Giorgio
___
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


[time-nuts] HP 5371A - Basic Howtos

2019-10-23 Thread Giorgio Barinetti
Hi all.

I've been given a 5371A  that i'll like to use to mesaure Allen Variation and 
plot results with TimeLab.

Can somebody be so kind to give me some hints on how to do that ? (instrument 
setup etc)

Regards,
Giorgio.
___
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


[time-nuts] EFRATOM MFST rollover way before 6/4 23:59:59

2019-04-06 Thread Giorgio Barinetti
Gents,


i realized that my EFRATOM MFST mainframe Rb + OXCO  with GPS has rolled over 
many months before expeted.

This summer, in August 21, the display has rolled over to 21/08/1999, and 
stayed locked then.

I wasn't aware, since the unit is located in a rack away from my working desk, 
but since i've a logging PC
Connected to the serial port, saving the logs from the GPS,tonight  i've found 
the rollover event.

Did someone notice something similar ?


Btw, happy rollover night...
Giorgio






___
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


[time-nuts] Oscilloquartz 3210 Cesium Standard Schematics

2018-07-27 Thread Giorgio Barinetti
Good News.

I finally decided to leave it running for a few days, after a bit of inspection 
(changed some caps here and there, removing '77 vingage NiCs batteries, etc).

Everything is fine, no alarms, readings are ok.

But, still no Ion Current. At least at the meters, both internal and external.

Since HV generation is off a sealed module, and not directly, I'll run a 
measurement on the bare wire going to the tube.

https://www.ebay.it/itm/High-Voltage-PS-from-OSCILLOQUARTZ-3102-CESIUM-FREQUENCY-STANDARD-ATOMIC-CLOCK/253730161665?hash=item3b137f0401:g:djEAAMXQ01tRWP7

Voltage is quite High, around 5Kv, that probably means no load from the tube.


Any advice ?

Ciao,
Giorgio.



From: Giorgio Barinetti
Sent: giovedì 12 luglio 2018 21:49
To: time-nuts@lists.febo.com
Subject: Oscilloquartz 3210 Cesium Standard Schematics


Hi Tom,





Ok and stable were referred to the general behavior of the unit (PSU voltages, 
current draw).



After a 2 h warmup I've run the following test:



Frequency has then been measured with the aid of an HP 5345 clocked to a 
Efratom MFST GPS+Rb and is stable.



Phase noise is just around -135 @ 100 hz, output power is 10 dBm.



I did not run a full test since I have to be sure that the clock is "safe" in 
term of leaky capacitors

And so on. Leaving a 44 Year instruments running unattended in the lab after 20 
years of sleep

Does not sound very safe to me :)



During the weekend I'll try to keep it running for 48H so I'll see what will 
happen.



In the meantime, thanks for your help!





Ciao,

Giorgio.





> i have been lucky to found a working Oscilloquartz 3210 Cesium ( warehouse of 
> a telco operator ) in a good shape.



Nice find. The OSA 32xx manual that I scanned is at:



http://leapsecond.com/museum/osa3200/



That's 33 pages of "Operating and Instruction Manual". If your unit is working 
within spec, that's sufficient. I'll check if additional manual pages are 
around here somewhere, in case you need to repair it.



> seems ok and stable



Words like "ok" or "stable" are vague; try to give numerical measurements for 
accuracy and stability. That will help determine if your vintage Cs clock is 
actually working to spec or not.



Also log values over time. The ion pump should not have been zero if the unit 
was sitting in a warehouse for 20 years. But after some hours or days, yes, it 
could be near zero and this might be a good sign.



For an old, surplus, vintage Cs standard, its best to carefully monitor 
everything immediately after power up and continuing for several hours or days. 
The trend lines will tell you a great deal about the condition and prognosis of 
the instrument.



/tvb





- Original Message -

From: "Giorgio Barinetti" http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com>>

To: http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com>>

Sent: Wednesday, July 11, 2018 3:17 PM

Subject: [time-nuts] Oscilloquartz 3210 Cesium Standard Schematics





> Hi Guys,

>

> i have been lucky to found a working Oscilloquartz 3210 Cesium ( warehouse of 
> a telco operator ) in a good shape.

>

> Powers up, seems ok and stable ( capacitors on the unit has 10/1974 date, so 
> very old unit, serial 83), but the signal

> From the preamplifier is a little low and the Ion Pump current is 0. Like the 
> 3 Kv supply is low ?

>

> Like to investigate but it will be nice to have a schematic.

>

> I've found a 2014 message where Tom was willing to scan the remaining part of 
> the manual, so I'm kindly asking :)

>

>

> Thanks,

> Giorgio Barinetti

___
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


[time-nuts] Oscilloquartz 3210 Cesium Standard Schematics

2018-07-12 Thread Giorgio Barinetti
Hi Tom,





Ok and stable were referred to the general behavior of the unit (PSU voltages, 
current draw).



After a 2 h warmup I've run the following test:



Frequency has then been measured with the aid of an HP 5345 clocked to a 
Efratom MFST GPS+Rb and is stable.



Phase noise is just around -135 @ 100 hz, output power is 10 dBm.



I did not run a full test since I have to be sure that the clock is "safe" in 
term of leaky capacitors

And so on. Leaving a 44 Year instruments running unattended in the lab after 20 
years of sleep

Does not sound very safe to me :)



During the weekend I'll try to keep it running for 48H so I'll see what will 
happen.



In the meantime, thanks for your help!





Ciao,

Giorgio.





> i have been lucky to found a working Oscilloquartz 3210 Cesium ( warehouse of 
> a telco operator ) in a good shape.



Nice find. The OSA 32xx manual that I scanned is at:



http://leapsecond.com/museum/osa3200/



That's 33 pages of "Operating and Instruction Manual". If your unit is working 
within spec, that's sufficient. I'll check if additional manual pages are 
around here somewhere, in case you need to repair it.



> seems ok and stable



Words like "ok" or "stable" are vague; try to give numerical measurements for 
accuracy and stability. That will help determine if your vintage Cs clock is 
actually working to spec or not.



Also log values over time. The ion pump should not have been zero if the unit 
was sitting in a warehouse for 20 years. But after some hours or days, yes, it 
could be near zero and this might be a good sign.



For an old, surplus, vintage Cs standard, its best to carefully monitor 
everything immediately after power up and continuing for several hours or days. 
The trend lines will tell you a great deal about the condition and prognosis of 
the instrument.



/tvb





- Original Message -

From: "Giorgio Barinetti" http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com>>

To: http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com>>

Sent: Wednesday, July 11, 2018 3:17 PM

Subject: [time-nuts] Oscilloquartz 3210 Cesium Standard Schematics





> Hi Guys,

>

> i have been lucky to found a working Oscilloquartz 3210 Cesium ( warehouse of 
> a telco operator ) in a good shape.

>

> Powers up, seems ok and stable ( capacitors on the unit has 10/1974 date, so 
> very old unit, serial 83), but the signal

> From the preamplifier is a little low and the Ion Pump current is 0. Like the 
> 3 Kv supply is low ?

>

> Like to investigate but it will be nice to have a schematic.

>

> I've found a 2014 message where Tom was willing to scan the remaining part of 
> the manual, so I'm kindly asking :)

>

>

> Thanks,

> Giorgio Barinetti

___
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


[time-nuts] Oscilloquartz 3210 Cesium Standard Schematics

2018-07-11 Thread Giorgio Barinetti
Hi Guys,

i have been lucky to found a working Oscilloquartz 3210 Cesium ( warehouse of a 
telco operator ) in a good shape.

Powers up, seems ok and stable ( capacitors on the unit has 10/1974 date, so 
very old unit, serial 83), but the signal
>From the preamplifier is a little low and the Ion Pump current is 0. Like the 
>3 Kv supply is low ?

Like to investigate but it will be nice to have a schematic.

I've found a 2014 message where Tom was willing to scan the remaining part of 
the manual, so I'm kindly asking :)


Thanks,
Giorgio Barinetti




___
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.