Hello Martyn,
Martyn Smith [EMAIL PROTECTED] wrote:
Despite the AGC we still achieve good phase noise (-135 dBc/Hz @ 1 Hz
with
a -168 dBc noise floor) together with 130 dB isolation and 10 ps/C phase
stability.
I have problems to understand, how your amplifier can achieve phase
noise of -135 dBc
I suggest not touching any of the adjustment points.
Check the rb light. If it gets too reddish, then probably the lamp gets too
hot. There is a resistor in the heater circuitry that may be faulty. I'm away
at the moment, and don't have my notes on hand, but the resistor is on the
bottom side
Sometimes trying to fix a faulty unit is simply a challenge.
Some more tips.
The LPRO with cover removed cannot work in the presence of artificial light
coming from AC powered sources!! It locks randomly to the light ripples.
My definition of the normal color of the rubidium light would be
You heard the Russian station RWM (.-. .-- --) which transmits on 4996,
9996, 14996 and maybe 19996. It is the easiest to receive station in Europe,
quite strong signal.
73,
Antonio I8IOV
Hi,
I listened today the WWW station and I've found it on 9.996 Mhz instead of
10.000Mhz and my
Maybe this could be of interest.
I have an LPRO continuously running since ending last January. Lamp volt was
7.92V.
As of October 22 (today, 9 months later) the voltage is 7.7V.
The voltage dropped by 0.22V.
The VCXO Volt was 6.7V, and now is 6.8V.
It is powered at 24V, and the temperature at
Hi Antonio,
precise pendulum clocks also suffer rate jumps, the process
of rate jumping seems to be common to most time counting
systems.
Thanks, I was missing this info.
I have an LPRO rubidium oscillator. I mounted it on a heat
sink with fins and placed it in an insulated box. A small
Hi Rick
I'm not quite sure what the question is here
The question is:
may anybody tell me (date and time) when a crystal jumped?
(a sample response could be: my crystal jumped on January 22, 2006 12:25 UT).
I would like to map in time as many jumps as possible.
I hope someone in the
Are you sure it is the tuning fork and not the reference
for your counter? -
73 - Mike
Hi Mike,
the tuning fork jumps were in the order of e-7, while the references were a HP
10811 and a Racal Opt 04E both in the range of e-10.
Thanks and 73,
ANtonio
Neville,
Hi Antonio,
precise pendulum clocks also suffer rate jumps, the process
of rate jumping seems to be common to most time counting
systems.
I have an interest in this. May you point me to any references?
73,
Antonio I8IOV
___
Hi all,
Thanks for your comments, info, ideas and references.
I've found the plots that Tim Shoppa pointed me to, and they
are a good response to my original question.
It has been interesting hearing about the jumps of pendulums
too, this was new to me.
Also I argue from the discussion that,
Thanks, Said.
What seldom happens to my tuning forks is similar to your
brown line, of course.
Do you have date and time of this beatiful jump?
Have you observed, by chance, a negative (restoring) jump at any time after
this jump?
(the sense of the latter question is that, if you observed a
On Oct 30 SAIDJACK wrote:
Hi guys,
there is a very nice Crystal jump visible online today:
_http://resco.ucol.mx/Fury/gpsstat.htm_
(http://resco.ucol.mx/Fury/gpsstat.htm)
Has anyone else seen a similar effect today?
This jump will be visible online for a couple more hours
On Fri, 31 Oct 2008 13:59:06 EDT SAIDJACK wrote:
Hi Antonio,
wow, that's amazing. Your prediction was pretty accurate!!
So I wonder if we have one of the most sensitive gravitational sensors out
there :)
You know that the issue is reproducibility.
How did you correlate the jump to
Yuri,
as a rule of thumb, read my message posted here last month:
http://www.febo.com/pipermail/time-nuts/2008-October/033926.html
I have 10 EFRATOM LPRO, and only one came with a faulty lamp.
I think it lost vacuum, and the glass looks clear but with
small white pigments. I inspected some of
Hi Neville,
inspecting the lamp is a quite easy job.
Yes, it is in the smaller housing.
Unlock the check nut just enough, and unscrew the lamp (along
with the check nut) using a point shaped tool (a toothpick).
It is easy.
Don't touch it with fingers.
Before doing this, take note of how much
Hi all,
I'm planning to use a HP 53131A for phase measurements.
In a limit case, just to understand, suppose that I'm
reading, say, 100 degrees, and the next reading gives again
100 degrees, but meanwhile the signal under test jumped 360
degrees. Is there any way to detect such a case?
Thanks Magnus and Tom, you clarified more than enough.
73,
Antonio I8IOV
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Time-nuts,
I have to admit that I have an interest in gravitational
anomalies.
If anybody out there has nothing better to do, here is the
stuff to spend some minutes.
In early 90's a Chinese researcher (Zhou) opened a debate
claiming to have observed anomalous behaviour on atomic
clocks
I got some private answers.
Sorry, I've been concise not to bias your thoughts.
Only, I would precise that for unlocked I mean a clock in
which the crystal is free running not locked to Rb or Cs, and not to lock
between clocks,
Antonio I8IOV
Time-nuts,
I have to admit that I have an interest
Time-nuts,
I was wrong with my interpretation.
The clocks were all working ok.
Thanks.
Antonio I8IOV
I got some private answers.
Sorry, I've been concise not to bias your thoughts.
Only, I would precise that for unlocked I mean a clock in
which the crystal is free running not locked to Rb or
What is the reference used to measure stability and accuracy
of H Masers?
Do Caesium clocks have something like a C field adjustment
(as Rb clocks have), and in the case, what it acts upon?
Thanks in advance,
Antonio I8IOV
___
time-nuts mailing
In order for cesium beam standards to even work, one must apply
a slight magnetic field, the so-called C-field, which rather strongly
distorts the shape of the resonance peak.
Tom (and all),
do you think that a Cs clock, having a not-well-adjusted
C field, might have a shape of the resonant
Rick Harold wrote:
All,
I'm planning doing some experiments in distance measurement. They don't
deal with atomic time directly but with extreme short periods of time.
I need to determine the position of a instrument with a 1mm accuracy or
less.
The instrument is not connected to a
John
You might also look at http://www.qsl.net/zl1bpu/PRECISION/Reference1.htm
Peter
ZL2AYX
On Wed, 18 Jan 2006 11:27:31 -0500, Neon John [EMAIL PROTECTED] wrote:
On Wed, 18 Jan 2006 00:03:37 -0800, Christopher Hoover
[EMAIL PROTECTED] wrote:
John-
If I divided this down to 1 PPS, would
Dear colleagues
When I lately visited the research facility where I had worked in the
80ies*) I was told that the local timescale is still generated by the two
BVA oscillators I had installed back then. One is free running and the other
one is daily corrected against GPS time-scale. (The place is
to: John,KE5FX
Hi John
I read your question on the THUNDERBOLT look-alike sold by 'accelium'. Since
I have bought this unit (for 500$!!!), I can answer some of it:
The hp58533a has a similar form factor as the Thunderbolt but it's an
entirely different design. I made the following notes when
Agilent have a Product Overview PDF for the 10811E/D on their website:
http://cp.literature.agilent.com/litweb/pdf/5091-1639E.pdf
which has a pinout. Hopefully they will have stuck to the same pinout.
Peter
Hello,
I purchased one of the HP 10811-60155 standards on ebay one month
Further to my previous reply, the HP Archive website has an Operating
Service Manual for the 10811A/B at:
http://www.hparchive.com/Manuals/HP-10811AB-Manual.pdf
In section 2.20 on the page numbered 8 (actually the sixteenth page in the
PDF), a 15-pin Cinch connector is described.
Certainly we use UTC in Britain, although the Government are keen to try
and keep the Greenwich link, so the term GMT is erroniously used more often
than not (especially by the BBC!).
Yes, there are people who want to stop this irritating leap
second business until well after they are dead,
Hi all,
does anybody out there have any experience in replacing the lamp on an LPRO
rubidium oscillator?
I have two LPROs, one working and another which reads lamp volt = zero. I
feel it is not a matter of lamp, as the lamp housing doesn't heat (it should).
I would try swapping the lamps
Hi all,
I've just joined this list today, and have found that the archive is full of
interesting stuff.
So I would ask, is there any possibility to search the
archive by keywords?
Thanks in advance.
Antonio I8IOV
___
time-nuts mailing list --
Hi Frank,
in fact Symmetricom states that LPRO is not field-repairable, and I suspect
that disassembly would be destructive of the delicate mechanical/thermal
coupling between the lamp assembly and the outer case. Anyway, as my unit
doesn't work, I will try to repair it, and even in the case of
Hi all,
I have some Qualcomm RF/SYNTH W/GPS boards which are programmable
synthesizers for the 750-1000 MHz band. They bear a 10 MHz TCXO as a reference.
Based on the above mentioned lettering on the PCB, I guess the oscillator could
be GPS disciplined supplying some GPS derived signal somewhere
. Can anyone help me ?
Thanks a lot...
Best regards
[EMAIL PROTECTED]
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
. The power
consumtion is about 3W and the unit is gettuing warm on the outside. I guess
this is okay ?
[EMAIL PROTECTED]
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow
Bernd
DK1AG
Original-Nachricht
Betreff: [time-nuts] HP10811 problem - wrong frequency
Datum: Sun, 13 Jan 2008 19:37:48 +0100
Von: [EMAIL PROTECTED] [EMAIL PROTECTED]
Antwort an: Discussion of precise time and frequency measurement
time-nuts@febo.com
Just speaking of 10811 clones, I have one made by Piezo Systems, model #
2810007-1. It fits electrically and mechanically HP boards, and looks like a
10811. But I miss the specs. Does anybody have any info?
Thanks,
Antonio
Geoff Blake wrote:
Hi,
I know this has been asked before - but I
Tom,
for your archive of tuning fork oscillators pictures, look at my Bryans
Aeroquipment (later a Negretti Zambra division) 50 Hz fork at
http://xoomer.alice.it/iovane
and click on fork.htm -
This appeared to be quite stable.
Antonio
___
time-nuts
Whether or not the rate of atomic clocks may be affected by a solar eclipse is
a controversial matter.
NATURE magazine published the article Chronometry: Effect of the 1999 solar
eclipse on atomic clocks,
see:
http://www.nature.com/nature/journal/v402/n6763/abs/402749a0.html
but the full text
It should be 58540A, not 58450A. Look at:
http://www.leapsecond.com/museum/hp58540a/ds-58540a.pdf
Antonio I8IOV
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow
I purchased on ebay four rubidium LPRO and I'm quite satisfied with them.
The lamp-Volt parameter (which is an indication of the lamp's health) was
respectively 8.4, 7.95, 6.8 and 5.1 Volt.
When an LPRO leaves the factory its lamp-V is around 12V. A drop of up to 3 V
would be expected in the
Antonio I8IOV ([EMAIL PROTECTED]) wrote:
The unconvenience with the 02M version is that it uses a military
GPIB language (MATE/CIIL), and I had to get familiar with it.
The Racal units that I have (with no button problems) have a
jumper for military or standard GPIB. They do have
Dan Rae wrote:
In my original copy of the 1992 manual, there is only one jumper
indicated on the GPIB board. It is at the front of the board. Pages
7-34 and 7-35 show it. On the first page it says : Make certain the
jumper plug is installed in the upper two pins of SK4 as shown.
On
Ulrich Bangert, DF6JB wrote:
So a better phrase would be that the lamps voltage is an indicator of
the lamp's health IF and ONLY IF everything else is sane. Don't blame
the lamp before you have checked the temperature and the exciting
field's frequency.
Agreed completely. I'm not yet
I have 2 different Racal Dana 1992 frequency counters:
The first one (kept for spares because the display is broken) has a
GPIB interface of type 19-1146:
this interface has NOT the SK4 jumper (see
RACAL-DANA-1991-1992-ServiceManualSch.pdf pages 16/17 in my archive).
The other one, in
Alternatively since you only need 25mA @ +12V, 25mA @ -12V and 300mA
@+5V (if I'm reading the right datasheet) you can use a much lower
voltage output transformer suited to the 5V regulator and use a pair of
voltage doublers or triplers to feed your 12V regulators.
A 7.5- 0 - 7.5V
Does it make any sense GPS disciplining a rubidium oscillator?
In such a case we have a chain made of
GPS - Rubidium - XTAL
as opposed to the simpler case of
GPS - XTAL
(assume that XTALs are of the same quality, and so the control loops).
Does the addition of Rb in the middle of the chain add
Tom Duckworth wrote:
Antonio,
Absolutely! With a good XTAL you have parts in the 9th, short term. With a
XTAL controlled by a Rubidium, in the phase-lock feedback loop, you have
parts in the 12th, short term. With the Rubidium disciplined by the GPS,
with its on-board Rubidium/Cesium
More precisely, if I had two black boxes, one containing a GPS-Rb-XTAL setup
and another containing GPS-XTAL, what measurement would you make from outside
the boxes to distinguish from one another?
Antonio I8IOV
Tom Duckworth wrote:
Antonio,
Absolutely! With a good XTAL you have
Thanks to all for the interesting discussion. I understand that,
having a thunderbolt, I might feel free to delay a bit (or even cancel)
my Rb disciplining project.
A last question to tvb. You said that
6) Turn-over -- if you flip the black box upside down, the Rb
version will show little or no
Rainer,
I presume your Power Supply has a fan which may absorb 2 to 4
Watt. Given the low consumption of Thunderbolt, you could try
disconnecting the fan to save a few Watts.
The Meanwell T-30B has no fan. I' ve not yet measured its
idling consumption (no load), but to answer your question it
Any sources for an MRF160 power mosfet? Or any suggestions for a replacement?
Thanks in advance.
Antonio I8IOV
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow
I want one. Where to send the request?
Antonio I8IOV
2) Those of you newcomers to Time-Nuts should expect
to wait until the first batch has been shipped and
TAPR is ready for you. There should be enough for
everyone; so don't panic.
53 matches