Hi,

I recently purchased an HP 58503A from eBay. When I power it on, with or 
without an antenna, the power LED illuminates on the front panel, followed by 
each LED one at a time. Then, after a moment, the unit alternates between all 
front panel LEDs being illuminated and only the power LED at a rate of about 
once every 3 seconds. I opened the unit and checked the rail voltages. I have 
5.22V, +15.01V, -15.00V. The oven is warming. There are some LEDs on the board 
near the antenna jack and I have two red illuminated and a single green. I 
can’t get anything on the serial port.

Has anyone seen this before? I have a 30 day ROR but I like this model and 
would love to fix it if anyone has any advice.

Thanks,
Mike

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A question (EFC: Err)

2018-01-05 Thread Tom Van Baak
Mitch,

Thanks for the screen shot. That means 99% of your 58503 is working; very good 
news.

If you look at the last line if your screen shot (attached), the self-test is 
failing due to EFC (Electronic Frequency Control), the voltage that pulls the 
quartz oscillator low or high to keep it locked at 10 MHz via GPS. The manual 
suggests this means the oscillator has drifted out of range, but there may be 
additional causes like the DAC or its bipolar power supply or the oscillator 
oven or oven power supply, etc.

I've not seen this error before on my 58503's, but it narrows down your problem 
significantly. Rather than guessing what the root cause is I'll cc time-nuts 
[1] to see if anyone else has seen this error, or knows what you can do about 
it.

There's also a useful thread at: 
www.eevblog.com/forum/testgear/hp-58503a-efc-error/

/tvb

[1] I see your original email went to time-nuts-owner@, which ends up going 
privately to the couple of us who run time-nuts. But if you post to 
time-nuts@febo.com, then it goes out to the 1800 member list.

If you want to join see http://www.leapsecond.com/time-nuts.htm for details 
about the list.

If you don't want to bother joining you can read the live feed for this month 
at: www.febo.com/pipermail/time-nuts/2018-January/date.html

If you want to read anything every written about the 58503A, on time-nuts 
search google using-- site:febo.com 58503A



- Original Message - 
From: Mitch Van Ochten 
To: 'Tom Van Baak' 
Sent: Thursday, January 04, 2018 7:21 PM
Subject: RE: HP 58503A question


Hi Tom,
 
This unit has the display also.  After applying power, the LED's on the front 
do a scan from left to right, the display says "HP 58503A" then the "STATUS" 
LED starts blinking once per second.  Thirty seconds later a message appears 
saying "SLFTST ERR".  The same instant the leftmost RED LED at the left rear of 
the circuit board comes on.
 
There is a GREEN LED (rightmost) in the set of six LED's, which continuously 
blinks once per second.  The 10 MHz output is present and the RS-232 is working.
 
After clearing the alarm, the "OCXO WARMUP" message appears, then a few minutes 
later a message saying "GPS FAILED" occurs. After waiting a while another 
message says "0 SAT AQUIRD". Regardless of how long you wait, it always says "0 
SAT AQUIRD".
 
I have the unit attached to a known good active antenna.  Bought a replacement 
GPS receiver for it and installed it but the same error still appears. It seems 
as though it is not recognizing the signal from the GPS receiver, and it may be 
related to the self-test error.
 
Here is what SatStat shows:
 
 
 
 
Best regards,
 
mitch
 
From: Tom Van Baak [mailto:t...@leapsecond.com] 
Sent: Thursday, January 04, 2018 9:21 PM
To: mitch
Subject: Re: HP 58503A question
 
I opened one up here and it also has no chip in that socket. I never noticed 
that before. If you want me to dig further I will.
What test failed? Or what are the other symptoms? Do you get 10 MHz or 1PPS or 
GPS lock?
Have you tried querying it with the RS232/ SCPI interface?
 
/tvb
 
- Original Message - 
From: Mitch Van Ochten 
Sent: Thursday, January 04, 2018 5:02 AM
Subject: HP 58503A question
 
Hi John,
 
I recently acquired an older HP 58503A.  After power-up the "status" LED blinks 
for about 30 seconds, then a message comes up saying it failed self-test.  
Inside it is missing a large IC (see attached photo).  Do you have one of these 
units?  If so, is yours also missing the IC?
 
 
Best regards,
 
Mitch Van Ochten

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

[time-nuts] HP 58503A GPS time and frequency receiver - How do I determine why the ALARM LED is on?

2017-01-25 Thread Dr. David Kirkby (Kirkby Microwave Ltd)
I've got an HP 58503A purchased from China a couple of years ago. It has a
bit of a hard life, having been in water at some time. But it seems to work
ok, as far as I can tell. Today there was a power failure for 5 minutes or
so. The red ALARM LED is on. It is not clear to me why, although of course
I'm aware there was a power outage, which almost certainly caused this,
although I can't be 100% sure.

Could it be that the log is full? I had a program running trying to get the
position multiple times in a loop, to get some sort of average, so I can
well believe I filled the log up. But this was a few weeks ago. The log is
clearly full .

scpi > diag:log:read:all?
Log status: Full

Log 001:19970504.00:00:00:  Log cleared
Log 002:19970504.00:00:00:  System preset
Log 003:19970504.00:00:00:  Power on

Log 219:20170107.05:30:17:  Position hold mode started
Log 220:20170107.05:30:57:  Holdover started, not tracking GPS
Log 221:20170107.05:31:03:  Survey mode started
Log 222:20170107.05:33:04:  GPS lock started

Nothing has been logged since 7th January. I know for a fact there was a
power outage on the 19th and today (25th), so I suspect the log being full
has NOT caused the ALARM light to come on. A fault, that is not logged, has
caused it.

scpi > SYSTEM:STATUS?
--- Receiver Status
---

SYNCHRONIZATION . [ Outputs
Valid ]
SmartClock Mode ___   Reference Outputs
___
>> Locked to GPS  TFOM 3
FFOM 0
   Recovery   1PPS TI +9.5 ns relative to
GPS
   Holdover   HOLD THR 1.000 us
   Power-up   Holdover Uncertainty

  Predict  432.0 us/initial 24
hrs

ACQUISITION  [ GPS 1PPS
Valid ]
Tracking: 6    Not Tracking: 1    Time

PRN  El  Az   SS   PRN  El  AzUTC  15:03:50 25 Jan
2017
  5  23 186   7324  21 252GPS 1PPS Synchronized to UTC
 13  84 110   81  ANT DLY  0 ns
 15  59 289   94  Position

 20  45 245   96  MODE Hold
 28  51  76  104
 30  27  68   83  LAT  N  51:39:04.155
  LON  E   0:46:36.381
ELEV MASK 20 deg  HGT   +45.55 m
(MSL)
HEALTH MONITOR .. [
ERROR ]
Self Test: Err   Int Pwr: OK   Oven Pwr: OK   OCXO: OK   EFC: OK   GPS Rcv:
OK
scpi >


The self-test shows an error, but no indication what the error was, as the
log was full.

I have a fairly good idea what may have happened. Sometimes when the unit
is powered on, the voltages on the PSU appear to go out of spec. I'm not
convinced this is really happening, as a multimeter on peak hold could not
catch any glitch, but I have not tried a fast scope.

I'm just wondering if there's anything else I can do to see what caused the
LED to come on. I suspect if I execute the self-test, it will pass and the
light go out, but I'd rather find the reason before doing that.

Dave
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] HP 58503A chassis part needed

2016-12-12 Thread Stan
Hello,

 

I'm in the process of restoring my 58503A. In the process of disassembling
it, the rear plastic frame that is held to the back panel with the two T15
Torx screws had become brittle with age and it cracked into several pieces
as I was removing it. It will be difficult to get all of the small pieces
glued back into place, so I'd rather just replace it with an intact part.

 

I have a spare rear frame from another HP bench instrument in my spare parts
bin, but much to my surprise I found that the center-to-center hole spacing
of the screws on the 58503A frame is about 7.5", whereas on all of the other
HP bench instruments that I've checked it's only about 7.375", so my spare
frame does not fit. Does anyone have, or know where to get, the proper rear
frame for my 58503A?

 

Thanks,

Stan 

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] HP 58503A failure

2016-11-04 Thread Stan
I have had an HP 58503A GPSDO with the Option 001 display running
continuously for over ten years without a problem-that is, until today. When
I looked at it this morning, the display was showing a valid time, but it
was frozen and not advancing. Pushing on the buttons had no effect. I
unplugged it, let it rest for a minute or so, and plugged it in again, and
this time the display lit up with a string of several identical "garbage"
characters that extended only part way across the display. The rest of the
display was blank. Pushing on the front panel buttons had no effect, the
display at one point changed to a different string of meaningless characters
and, needless to say, the clock never resumed running. 

 

Unplugging it and plugging it in again resulted in the same behavior,
although different "garbage" characters were displayed, including some of
the annunciators in the VFD that are just 53131A and 53132A counter
functions (+WID, -WID, GATE, etc.). I opened it up and checked the power
supplies, and they were spot on (+5 VDC, +15 VDC, and -15 VDC) with no
measurable ripple. I also checked to see if I could communicate with it
through the serial port using GPSCom (which has worked fine in the past),
but no luck. I just got "Receiver Timeout" error messages. The power LED
glows steady green, but no other LEDs light, even when the buttons are
pushed.

 

I'm not sure if a service manual or CLIP even exists, but I certainly
couldn't locate any service info whatsoever. Any ideas?

 

Thanks,

Stan

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A Com 4 conflict

2015-01-05 Thread Didier Juges
To add a small dose of additional complexity, Visual Basic 6.0 (and I suspect 
other dev tools of that generation) only support COM ports up to 16. Not as bad 
as 4, but still a problem on occasion.

Didier KO4BB


On January 4, 2015 12:03:05 PM CST, Bob Camp kb...@n1k.org wrote:
Hi

Ok, off into serial land …. 

The GPSDO does not really know or care what port it’s on. It will be
just as happy on port 119 as on port 4.  It (by design) does care about
the baud rate / data bits / parity / stop bits. This is as true with a
USB to serial as with a direct serial card. That’s the good news. Now
for the rest of the news.

If your computer operating system is old enough, you may only *have* 4
com ports. Bad news there = upgrade the computer. Typically this is a
Windows 95 and earlier issue. 

If the software running on the computer is old enough it may only
*know* how to use 4 com ports. Bad news is same = upgrade the software.
Typically this is a Windows 95 era (mid 90’s) issue. 

If the drivers on the serial to usb converter are old enough, they may
only talk as 4 com ports (I’ve never seen this on a modern OS). Same
news = upgrade the driver.

All this relates to Windows. The cutoff times and versions for other
OS’s are a bit more obscure. The same basic details apply. 



If you are running on an XP or newer machine, you should have far more
than 4 com ports available. If you are running any of the newer ( =
past 2000) software to talk to the box, it should be quite happy to
deal with more than 4 com ports. 

My *guess* is that you need a different piece of software. Exactly what
you are doing will determine which one you need. ( = more info
required).

-

 On Jan 4, 2015, at 10:50 AM, James Robbins jsrobb...@earthlink.net
wrote:
 
 My new old HP58503A wants to connect to my PC on Com 1-4.  Other PC
devices are already using those 4 Com ports.  
 
 Is there any way to connect it to a Com port other than one of those
1 to 4?
 
 I currently use an Edgeport USB-to-Serial Converter which works very
well to communicate with the PC and to assign Com ports to my various
GPS units.  But, as far as I know, the chosen Edgeport Com port must
fall within the range of Com ports which for which the device is
designed.
 
 If this has been answered, please point me to the discussion.  Many
thanks.  And, Happy New Year to all, with wishes for a more peaceful
(and timely - couldn't resist) year.

I have this terrible urge to start a calendar thread revolving around
“new” years … so far I’ve resisted the urge. It’s *very* difficult
….arhhh….

Bob

 
 Jim Robbins
 N1JR  
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

-- 
Sent from my Motorola Droid Razr HD 4G LTE wireless tracker while I do other 
things.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A Com 4 conflict

2015-01-05 Thread James Robbins
Thanks Bob, Didier and Rex for the consult on the Com 4 Conflict and the 
teaching.

Update:  First, sorry for not posting adequate information in my question to 
start with.  The result of some emails was a clarification that it was not the 
58503A which was causing the issue with the Com ports, but SatStat which only 
allowed Com 1-4 (a very old program).  I was able to switch to using GPSCon 
instead of SatStat and it provided many unused Com port choices.  So, the 
58503A and my ThinkPad are both happy campers.

Jim Robbins
N1JR
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A Com 4 conflict

2015-01-05 Thread Rex
You left out a lot of details, like what OS your PC is running, Assuming 
it is Windows XP or greater, you may be able to juggle the assigned com 
port.


If you are using a USB/serial adapter, be sure it is plugged in so it 
shows on the devices list.


You need to open Device Manager. One way is Start/settings/Control 
Panel. If it is Win 7+ Device Manager should be in the list. If XP, 
select System, the Hardware tab at the top, then the Device Manager 
button. -- I don't have a Win 8 machine; I assume it is the same as Win 7.


Find your serial device under 'Ports (Com  LPT)'. If you are not sure 
which one and it is a USB adapter, you can unplug it and see which one 
goes away.
Double-click on the one you want to change. In the window that opens, 
select the Port Settings tab at the top.  You'll get a window with the 
baud rate, etc. Select the Advanced button. Then in the lower left of 
the window is the assigned Com Port. Click the pull-down triangle button 
and you'll get a list where you can select a different Com Port number. 
If all the ones you want are flagged with (In use), you may have to back 
out and try to find a listing for the device that is using a low Com 
port and select it to change its number higher to free the one you want 
to use. I'm not sure what happens if you try to change to a port that is 
(In use). Maybe it tries to work it out for you. I haven't tried it.


I'd suggest Com 4 or Com 2 as your best choices.

So I don't have an exact answer. It depends on what is using the lower 
numbers now, Hopefully you can rearrange them using this method to find 
something that works.




On 1/4/2015 7:50 AM, James Robbins wrote:

My new old HP58503A wants to connect to my PC on Com 1-4.  Other PC devices are 
already using those 4 Com ports.

Is there any way to connect it to a Com port other than one of those 1 to 4?

I currently use an Edgeport USB-to-Serial Converter which works very well to 
communicate with the PC and to assign Com ports to my various GPS units.  But, 
as far as I know, the chosen Edgeport Com port must fall within the range of 
Com ports which for which the device is designed.

If this has been answered, please point me to the discussion.  Many thanks.  
And, Happy New Year to all, with wishes for a more peaceful (and timely - 
couldn't resist) year.

Jim Robbins
N1JR
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] HP 58503A Com 4 conflict

2015-01-04 Thread James Robbins
My new old HP58503A wants to connect to my PC on Com 1-4.  Other PC devices are 
already using those 4 Com ports.  

Is there any way to connect it to a Com port other than one of those 1 to 4?

I currently use an Edgeport USB-to-Serial Converter which works very well to 
communicate with the PC and to assign Com ports to my various GPS units.  But, 
as far as I know, the chosen Edgeport Com port must fall within the range of 
Com ports which for which the device is designed.

If this has been answered, please point me to the discussion.  Many thanks.  
And, Happy New Year to all, with wishes for a more peaceful (and timely - 
couldn't resist) year.

Jim Robbins
N1JR  
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A Com 4 conflict

2015-01-04 Thread Bob Camp
Hi

Ok, off into serial land …. 

The GPSDO does not really know or care what port it’s on. It will be just as 
happy on port 119 as on port 4.  It (by design) does care about the baud rate / 
data bits / parity / stop bits. This is as true with a USB to serial as with a 
direct serial card. That’s the good news. Now for the rest of the news.

If your computer operating system is old enough, you may only *have* 4 com 
ports. Bad news there = upgrade the computer. Typically this is a Windows 95 
and earlier issue. 

If the software running on the computer is old enough it may only *know* how to 
use 4 com ports. Bad news is same = upgrade the software. Typically this is a 
Windows 95 era (mid 90’s) issue. 

If the drivers on the serial to usb converter are old enough, they may only 
talk as 4 com ports (I’ve never seen this on a modern OS). Same news = upgrade 
the driver.

All this relates to Windows. The cutoff times and versions for other OS’s are a 
bit more obscure. The same basic details apply. 



If you are running on an XP or newer machine, you should have far more than 4 
com ports available. If you are running any of the newer ( = past 2000) 
software to talk to the box, it should be quite happy to deal with more than 4 
com ports. 

My *guess* is that you need a different piece of software. Exactly what you are 
doing will determine which one you need. ( = more info required).

-

 On Jan 4, 2015, at 10:50 AM, James Robbins jsrobb...@earthlink.net wrote:
 
 My new old HP58503A wants to connect to my PC on Com 1-4.  Other PC devices 
 are already using those 4 Com ports.  
 
 Is there any way to connect it to a Com port other than one of those 1 to 4?
 
 I currently use an Edgeport USB-to-Serial Converter which works very well to 
 communicate with the PC and to assign Com ports to my various GPS units.  
 But, as far as I know, the chosen Edgeport Com port must fall within the 
 range of Com ports which for which the device is designed.
 
 If this has been answered, please point me to the discussion.  Many thanks.  
 And, Happy New Year to all, with wishes for a more peaceful (and timely - 
 couldn't resist) year.

I have this terrible urge to start a calendar thread revolving around “new” 
years … so far I’ve resisted the urge. It’s *very* difficult ….arhhh….

Bob

 
 Jim Robbins
 N1JR  
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP-58503A

2015-01-02 Thread EB4APL

Hi,

The past week I had to replace more than 30 capacitors in various of my 
home electronics, some were bulged and others not so.
There were 7 in a computer motherboard, 10 in other, 8 in a TV set power 
supply, 2 in an external USB disk power supply and the rest in other 
things, I know the grand total because I kept it together to test a 
cheap EMS meter against a LCR bridge.
In my experience, all bulged capacitors are very ill, high ESR and very 
low capacitance, causing a lot of symptoms to appear (many of them 
intermittent).  And when a capacitor bulges normally some or all of its 
neighbors are bad also, even without any external sign. I check them 
with a bridge and this is confirmed most of the times, so I suggest to 
replace the bulged capacitor and when you are there check or replace the 
others, they will bulge or burst in short time, high ripple currents and 
heat makes electrolytics the less reliable electronic components today.


Best regards,
Ignacio EB4APL


On 02/01/2015 a las 14:34, Dr. David Kirkby (Kirkby Microwave Ltd) wrote:

On 1 January 2015 at 17:03, Andy Bardagjy andybarda...@gmail.com wrote:

Sounds like the GPS receiver is hosed. I think there are two different 
receivers used in the 58503a, unfortunately I'm away from my lab, otherwise I 
could check mine. It is a standard part, and may be available on the surplus 
market.

Before replacing, I'd check the usual suspects, power supply health (look for 
failed electrolytics) and re-seat the gps board to board connector.

Happy to measure things on my 58503a.

The fact it originally failed with errors indicting the GPS receiver
was not ok (nt Power- OK, OCXO- OK, EFC-OK GPS RCV-err. I, but later
he can't communicate with the 58503A over RS-232, to me indicates the
problem is not likely to be the just (if at all) the GPS receiver.

As you say, power supply is a possible problem.

I have a 58503A here that has a problem. Sometimes when power is first
applied, the Alarm light stays on, and the log show power supply
voltage errors. Yesterday I must have switched the thing on/off about
30 times before I managed to get the Alarm light to stay on. At the
time I had a handheld DVM connected to the +15 V rail with the peak
hold mode enabled. At least according to the handheld DVM, the +15 V
rail was normal, so either the transient is too short for my handheld
DVM to see, or the 85050A is reporting data voltage data incorrectly.
Both are fairly like I suspect.

I noticed a *very* slight bulge at the top of on a 100 uF, 400 V
capacitor on the switch mode power supply.  For various reasons, I am
not going to change that cap now, but obviously a failed cap could
cause this sort of problem.

Dave



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP-58503A

2015-01-02 Thread Andy Bardagjy
Sounds like the GPS receiver is hosed. I think there are two different 
receivers used in the 58503a, unfortunately I'm away from my lab, otherwise I 
could check mine. It is a standard part, and may be available on the surplus 
market. 

Before replacing, I'd check the usual suspects, power supply health (look for 
failed electrolytics) and re-seat the gps board to board connector.  

Happy to measure things on my 58503a. 

Andy ◉ Bardagjy.com ◉ +1-404-964-1641

 On Dec 30, 2014, at 12:35 PM, Richard Thorpe kisso...@gmail.com wrote:
 
 This list was recommended to me by the elecraft list.  I have an HP-58503A 
 sat disciplined “clock” that puts out a steady 10MHz to my K3 radio.  Its 
 been working 24/7 for years, I monitor it with David Anderson's Mac GPS 
 Control X software on my iMac. Several days ago error messages showed up 
 i.e.: Health- err, Self Test-err, Int Power- OK, OCXO- OK, EFC-OK GPS 
 RCV-err. Its not tracking any sats and now I cannot event communicate with it 
 at all even with a simple com program.  Is there anyone out therE who can fix 
 these things?  Any suggestions would be appreciated.  Thank you.
 
 Richard Thorpe K6CG
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP-58503A

2015-01-02 Thread Dr. David Kirkby (Kirkby Microwave Ltd)
On 1 January 2015 at 17:03, Andy Bardagjy andybarda...@gmail.com wrote:
 Sounds like the GPS receiver is hosed. I think there are two different 
 receivers used in the 58503a, unfortunately I'm away from my lab, otherwise I 
 could check mine. It is a standard part, and may be available on the surplus 
 market.

 Before replacing, I'd check the usual suspects, power supply health (look for 
 failed electrolytics) and re-seat the gps board to board connector.

 Happy to measure things on my 58503a.

The fact it originally failed with errors indicting the GPS receiver
was not ok (nt Power- OK, OCXO- OK, EFC-OK GPS RCV-err. I, but later
he can't communicate with the 58503A over RS-232, to me indicates the
problem is not likely to be the just (if at all) the GPS receiver.

As you say, power supply is a possible problem.

I have a 58503A here that has a problem. Sometimes when power is first
applied, the Alarm light stays on, and the log show power supply
voltage errors. Yesterday I must have switched the thing on/off about
30 times before I managed to get the Alarm light to stay on. At the
time I had a handheld DVM connected to the +15 V rail with the peak
hold mode enabled. At least according to the handheld DVM, the +15 V
rail was normal, so either the transient is too short for my handheld
DVM to see, or the 85050A is reporting data voltage data incorrectly.
Both are fairly like I suspect.

I noticed a *very* slight bulge at the top of on a 100 uF, 400 V
capacitor on the switch mode power supply.  For various reasons, I am
not going to change that cap now, but obviously a failed cap could
cause this sort of problem.

Dave
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] HP-58503A

2014-12-30 Thread Richard Thorpe
This list was recommended to me by the elecraft list.  I have an HP-58503A sat 
disciplined “clock” that puts out a steady 10MHz to my K3 radio.  Its been 
working 24/7 for years, I monitor it with David Anderson's Mac GPS Control X 
software on my iMac. Several days ago error messages showed up i.e.: Health- 
err, Self Test-err, Int Power- OK, OCXO- OK, EFC-OK GPS RCV-err. Its not 
tracking any sats and now I cannot event communicate with it at all even with a 
simple com program.  Is there anyone out therE who can fix these things?  Any 
suggestions would be appreciated.  Thank you.

Richard Thorpe K6CG
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A service manual with schematics

2014-06-01 Thread Mark C. Stephens
John,


Most time nuts have searched for this particular manual at one stage.

I am afraid it is unobtainium.


marks




-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf 
Of gianni
Sent: Thursday, 29 May 2014 7:09 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] HP 58503A service manual with schematics

hi John
I am new in the list but not in the art
here a lot of manuals
http://exodus.poly.edu/~kurt/manuals/manuals/HP%20Agilent/

Gianni

In 28 maggio 2014, starb...@uplink.net scritto:
 Hello to the time experts,
 I own a nice HP-58503A and I would be happy to find a pdf version of 
 the complete service manual for it. Is somebody in the position to 
 help me or to give me a hint where I could find it?
 Thank you very much,
 regards,
 
 John
 
 ___
 time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
___
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A service manual with schematics

2014-05-28 Thread gianni
hi John
I am new in the list but not in the art
here a lot of manuals
http://exodus.poly.edu/~kurt/manuals/manuals/HP%20Agilent/

Gianni

In 28 maggio 2014, starb...@uplink.net scritto:
 Hello to the time experts,
 I own a nice HP-58503A and I would be happy
 to find a pdf version of the complete service manual
 for it. Is somebody in the position to help me
 or to give me a hint where I could find it?
 Thank you very much,
 regards,
 
 John
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] HP 58503A service manual with schematics

2014-05-27 Thread starbook

Hello to the time experts,
I own a nice HP-58503A and I would be happy
to find a pdf version of the complete service manual
for it. Is somebody in the position to help me
or to give me a hint where I could find it?
Thank you very much,
regards,

John

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] HP 58503A Service Manual with schematics

2014-05-03 Thread starbook
Does anyone have a service manual for the HP 58503A GPS Time and 
Frequency Receiver?

Thanks,
John
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A acurate UTC time

2014-03-12 Thread Timestep
Thanks guys

It is now reading UTC, but only after a power off routine.  That confused me !

I can't afford the time to let it re-survey, and used the Query command with a 
0 instead of a ?.  Re-querying it with a question mark it replied that 
there would be no survey on power up.  The iPhone photo of the screen is here 
www.soundhifi.com/FOURUMIMAGES/gps.jpg

So the question is, will this really stop the annoying survey every time ?

And finally, does anyone have the circuit of the display so I can put a switch 
in the reduce the brightness please ?

Thanks

Dave
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A acurate UTC time

2014-03-12 Thread Hal Murray
 So the question is, will this really stop the annoying survey every time ?

I'd browse the manual looking for fine print.  The idea you are looking for 
is that after a survey it will save the location.  If it has a saved location 
on power up it will use it rather than do a new survey.

The other question is why are you turning it on/off often enough for a survey 
to be annoying?  That type of box works a lot better if left on all the time.

-- 
These are my opinions.  I hate spam.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A acurate UTC time

2014-03-12 Thread Timestep
HP 58503A 


Hi Hal

Good point, I don't want the display lighting my office up at night, hence I 
often turn it off.

What I really need is the circuit of the front panel, then I could dim the 
display and leave in on 24/7

Thanks

Dave

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] HP 58503A acurate UTC time

2014-03-11 Thread Timestep
From:  Dave Cawley
Timestep Dartmouth
Tel:   +44 (0) 1803 833366
Fax:  +44 (0) 1803 839498

HP 58503A accurate UTC time

I have the version with the display panel and also the SatStat50 software all 
working fine.

The time on both the display and PC appears to be 16 seconds fast.  I queried 
the leap second and it says correctly 16 seconds.

The display reads :  GPS  hh:mm:ss 

  a.. So how do I set it to read real time ? 
  b.. And is there a way of dimming or tuning off the display ? 
  c.. Finally, how do I stop it doing a survey every time ?
I know this must be obvious, but right now to me it isn't !
Thanks

Dave
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A acurate UTC time

2014-03-11 Thread Tom Van Baak
  a.. So how do I set it to read real time ? 


Try :DIAG:GPS:UTC 1

  b.. And is there a way of dimming or tuning off the display ? 

No software command, but you try adding a resistor or switch to the VFD power.

  c.. Finally, how do I stop it doing a survey every time ?

Try :GPS:POS: ...

/tvb

- Original Message - 
From: Timestep informat...@time-step.com
To: time-nuts@febo.com
Sent: Tuesday, March 11, 2014 11:33 AM
Subject: [time-nuts] HP 58503A acurate UTC time


 From:  Dave Cawley
 Timestep Dartmouth
 Tel:   +44 (0) 1803 833366
 Fax:  +44 (0) 1803 839498
 
 HP 58503A accurate UTC time
 
 I have the version with the display panel and also the SatStat50 software all 
 working fine.
 
 The time on both the display and PC appears to be 16 seconds fast.  I queried 
 the leap second and it says correctly 16 seconds.
 
 The display reads :  GPS  hh:mm:ss 
 
  a.. So how do I set it to read real time ? 
  b.. And is there a way of dimming or tuning off the display ? 
  c.. Finally, how do I stop it doing a survey every time ?
 I know this must be obvious, but right now to me it isn't !
 Thanks
 
 Dave
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A acurate UTC time

2014-03-11 Thread Hal Murray
  a.. So how do I set it to read real time ? 

My notes say that for a Z3801A:
:diag:gps:utc 1
It may need a reboot.


A google search for HP 58503A manual gets lots of hits that look 
interesting.  This is the top of the list:

58503B GPS Time and Frequency Reference ... - LeapSecond.com
www.leapsecond.com/museum/hp58503a/097-58503-13-iss-1.pdf



-- 
These are my opinions.  I hate spam.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] Hp 58503A

2012-10-16 Thread Mark Danon
Hey guys, new to the group here and wanted to throw a question out there. I 
just picked up an HP 58503A and hooked it all up. I have it connected to my 
computer and I can see the system status via satstat and all looks to be 
working correctly. However, I notice that there is distortion on the 10 meg 
sine wave on the 10 meg out on the back. This distortion consists of a slightly 
   mis-shapen negative peak. The positive peak it normal. The unit has now 
been on for 15 hours and no change. It is locked to gps. The 10 meg out 
directly from the back of the XTAL is a perfectly normal sine wave. The 
question: is this a problem with my unit? Is this normal? And, how can I fix 
it. Thanks fellas, Mark.

Sent from my iPad
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Hp 58503A

2012-10-16 Thread Said Jackson

Mark,

Are you loading the 10Mhz output with 50 Ohms when you measure it?

Said

Sent from my iPad

On Oct 16, 2012, at 6:52, Mark Danon mdan...@gmail.com wrote:

 Hey guys, new to the group here and wanted to throw a question out there. I 
 just picked up an HP 58503A and hooked it all up. I have it connected to my 
 computer and I can see the system status via satstat and all looks to be 
 working correctly. However, I notice that there is distortion on the 10 meg 
 sine wave on the 10 meg out on the back. This distortion consists of a 
 slightlymis-shapen negative peak. The positive peak it normal. The 
 unit has now been on for 15 hours and no change. It is locked to gps. The 10 
 meg out directly from the back of the XTAL is a perfectly normal sine wave. 
 The question: is this a problem with my unit? Is this normal? And, how can I 
 fix it. Thanks fellas, Mark.
 
 Sent from my iPad
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Hp 58503A

2012-10-16 Thread Mark Danon
I guess this it what they would call a rookie move? WOW, i feel like an
idiot. I just loaded it down with a 50 ohm load and looks perfect. Thanks
for enlightening me.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Hp 58503A

2012-10-16 Thread Tom Knox

I overlooked it also. Glad the problem is resolved.

Thomas Knox



 Date: Tue, 16 Oct 2012 12:31:13 -0400
 From: mdan...@gmail.com
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] Hp 58503A
 
 I guess this it what they would call a rookie move? WOW, i feel like an
 idiot. I just loaded it down with a 50 ohm load and looks perfect. Thanks
 for enlightening me.
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
  
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Hp 58503A

2012-10-16 Thread SAIDJACK
No problem, happens to me too some times,
 
usually the 10MHz Sine should be loaded with 50 Ohms, and the CMOS/TTL 1PPS 
 pulses should be run open-ended (1M or higher)..
 
I say usually.. because there are cases where proper termination is  
critical such as LVDS 1PPS outputs.. and where the 10MHz does not necessarily  
need to be loaded with 50 Ohms, resulting in a higher sine wave voltage - such  
as on our JLT GPSDO products..
 
bye,
Said
 
 
In a message dated 10/16/2012 09:36:25 Pacific Daylight Time,  
mdan...@gmail.com writes:

I guess  this it what they would call a rookie move? WOW, i feel like an
idiot. I  just loaded it down with a 50 ohm load and looks perfect. Thanks
for  enlightening  me.
___
time-nuts mailing  list -- time-nuts@febo.com
To unsubscribe, go to  
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the  instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Hp 58503A

2012-10-16 Thread Charles P. Steinmetz



I  just loaded it down with a 50 ohm load and looks perfect.


One of my design pet peeves is not anticipating what users may 
do.  For example, not anticipating that someone might terminate an 
output in a high impedance rather than in the rated impedance, or 
leave unused outputs unterminated.  I have seen several multichannel 
isolation amplifiers where any output not terminated in 50 ohms is in 
hard clipping -- even with halfway decent isolation, this can put 
garbage on the other outputs.  The very definition of an isolation 
amplifier is that each output should be independent of how the other 
outputs are terminated -- short, open, inductor, capacitor, other 
signal, whatever.  I understand it is not trivial to provide 1 Vrms 
(+13 dBm) outputs with 6 dB overhead from a 5 V (or 3.3 V) supply, 
but that's why they pay designers the big bucks.



Best regards,

Charles





___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A

2012-08-03 Thread Christoph Kopetzky
Peter Bell bell.peter@... writes:

 
 If it is, then you should see a pulse every 2 seconds, lined up with the
 even seconds in GPS time.  This might seem a rather strange signal to
 provide, but it's what the IS-95 derived CDMA systems use to trigger the
 start of another 100 block signalling frame.  It also tends to suggest that
 the unit you have was originally being used as part of a CDMA system.
 
 Regards,
 
 Pete
 

Hi Pete,

I snapped on the osci probe and see no 1 PPS or 2 PPS signal.
It seems like a noise signal with several positive and negative spikes with an 
peak to peak value of 50 to 60 mV...

disapointing...

Next is to analyze the rest with a logicprobe but the signal level seems to be
too small for beeing a logic signal...

TTL should be at 15 V, CMOS 5 V... but 50 mV ? 

Chris



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A

2012-08-03 Thread Doug Reed
I ran into a wiki description of GPS using WGS84 a couple days ago. It 
included a mention of ESEC and it was something like: Earth Static, 
Earth Centric. I think I was following a link about the Z3801A.


It referred to the fact that lat-lon is referenced to a static grid on 
the Earth and doesn't change as the Earth rotates on its axis or around 
the sun. Earth static and Earth centric.


73, Doug Reed, N0NAS.

Christoph Kopetzky wrote:

But now I am cleaning my devices and I am reading the label
ESEC again on the back.
So the question comes up again, what is ESEC?
Google knows nothing about the signal or protocol...


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A

2012-08-03 Thread Jim Lux

On 8/2/12 6:24 PM, Doug Reed wrote:

I ran into a wiki description of GPS using WGS84 a couple days ago. It
included a mention of ESEC and it was something like: Earth Static,
Earth Centric. I think I was following a link about the Z3801A.

It referred to the fact that lat-lon is referenced to a static grid on
the Earth and doesn't change as the Earth rotates on its axis or around
the sun. Earth static and Earth centric.



Earth Centered Inertial - ECI   centered on the earth, doesn't rotate 
with earth - very popular with earth orbiting satellites, because their 
center of rotation is at the center of the earth but their orbital 
plane is fixed in celestial terms.  like RA/Decl


Earth Centered Earth Fixed - ECEF (or ECF)  centered on earth, rotates 
with earth.  More like conventional lat/lon.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A

2012-08-03 Thread bg
Doug,

 I ran into a wiki description of GPS using WGS84 a couple days ago. It
 included a mention of ESEC and it was something like: Earth Static,
 Earth Centric. I think I was following a link about the Z3801A.

 It referred to the fact that lat-lon is referenced to a static grid on
 the Earth and doesn't change as the Earth rotates on its axis or around
 the sun. Earth static and Earth centric.

That reference frame is called Earth Centered Earth Fixed  (ECEF). And
that is what GPS is using internally. However it is not lat/lon/h, but
cartesian with origo in the earth center, xy in the equatorial plane and z
pointing towards the north pole.

http://en.wikipedia.org/wiki/ECEF

--

   Björn



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A

2012-08-03 Thread Azelio Boriani
What 'scope are you using (brand and model)? TLL is 5V, CMOS 4000 series
can be 15V but runs starting at 3V. Set the scope vertical (channel 1) at
1V/div (probe at 10:1), use NORMAL for the trigger and set the trigger (for
the channel 1) *slowly* from 0V to +200mV (assuming that the trigger isn't
aware of the 10:1) trying to follow the trigger indicator or the trace
appearing on the screen.

On Fri, Aug 3, 2012 at 3:31 PM, b...@lysator.liu.se wrote:

 Doug,

  I ran into a wiki description of GPS using WGS84 a couple days ago. It
  included a mention of ESEC and it was something like: Earth Static,
  Earth Centric. I think I was following a link about the Z3801A.
 
  It referred to the fact that lat-lon is referenced to a static grid on
  the Earth and doesn't change as the Earth rotates on its axis or around
  the sun. Earth static and Earth centric.

 That reference frame is called Earth Centered Earth Fixed  (ECEF). And
 that is what GPS is using internally. However it is not lat/lon/h, but
 cartesian with origo in the earth center, xy in the equatorial plane and z
 pointing towards the north pole.

 http://en.wikipedia.org/wiki/ECEF

 --

Björn



 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A

2012-08-03 Thread Christoph Kopetzky
Azelio Boriani azelio.boriani@... writes:

 
 What 'scope are you using (brand and model)? TLL is 5V, CMOS 4000 series
 can be 15V but runs starting at 3V. Set the scope vertical (channel 1) at
 1V/div (probe at 10:1), use NORMAL for the trigger and set the trigger (for
 the channel 1) *slowly* from 0V to +200mV (assuming that the trigger isn't
 aware of the 10:1) trying to follow the trigger indicator or the trace
 appearing on the screen.
 
 On Fri, Aug 3, 2012 at 3:31 PM, bg@... wrote:
 

Azelio,

I tried my two HAMEG Scopes for this job, a digital variant HMO3524 and
my analog HM2005-2 scope...
I am quite familiar with scope measurements :-)
so there are definitely no PPS signals on the output...

Chris


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A

2012-08-03 Thread Christoph Kopetzky
Hal Murray hmurray@... writes:


 We discussed this area a week or three ago.
 
 You don't need a digital scope to determine if there is a pulse.  A digital 
 scope may help to see the pulse and figure out what it looks like.
 
 With an analog scope, you can either look at the blinking light that tells 
 you it's triggering, or you can reduce the sweep speed until you can easily 
 see the (flat) line from the beam each time it triggers.
 
 I can see a 10 microsecond pulse with my old Tek 465.  It blinks and I 
 roughly remember what the picture looks like.  If I want to know a detail, I 
 have to look at the right spot and wait for the next pulse.
 


Hi Hal,

with my analog scope (Hameg HM2005-2) I can trigger a low level voltage
spike signal with  a p-p voltage of 30 mV and a frequency of 52,6 kHz.
Could that be real?
The signal level is very low, 30 mV...

Chris


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A

2012-08-03 Thread Christoph Kopetzky
Doug Reed n0nas@... writes:

 
 I ran into a wiki description of GPS using WGS84 a couple days ago. It 
 included a mention of ESEC and it was something like: Earth Static, 
 Earth Centric. I think I was following a link about the Z3801A.
 
 It referred to the fact that lat-lon is referenced to a static grid on 
 the Earth and doesn't change as the Earth rotates on its axis or around 
 the sun. Earth static and Earth centric.
 
 73, Doug Reed, N0NAS.
 


Hi Doug,

do you have more infos about the signal type of that Earth static, Earth 
centric?
is it a digital or analog signal?

Thanks

Chris




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A

2012-08-03 Thread Azelio Boriani
Here:
www.leap*second*.com/museum/*hp58503*a/097-58503-12-iss-1.pdf
you can find that there is the option 002 for the even second but they call
it PP2S...
Try to discover if your 58503 has this option 002.

On Fri, Aug 3, 2012 at 5:09 PM, Christoph Kopetzky dekag...@gmail.comwrote:

 Hal Murray hmurray@... writes:


  We discussed this area a week or three ago.
 
  You don't need a digital scope to determine if there is a pulse.  A
 digital
  scope may help to see the pulse and figure out what it looks like.
 
  With an analog scope, you can either look at the blinking light that
 tells
  you it's triggering, or you can reduce the sweep speed until you can
 easily
  see the (flat) line from the beam each time it triggers.
 
  I can see a 10 microsecond pulse with my old Tek 465.  It blinks and I
  roughly remember what the picture looks like.  If I want to know a
 detail, I
  have to look at the right spot and wait for the next pulse.
 


 Hi Hal,

 with my analog scope (Hameg HM2005-2) I can trigger a low level voltage
 spike signal with  a p-p voltage of 30 mV and a frequency of 52,6 kHz.
 Could that be real?
 The signal level is very low, 30 mV...

 Chris


 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A

2012-08-03 Thread Doug Reed

OK, that's what I get for working from memory. :-)
I got the idea more or less right but it has nothing to do with the 
abbreviation under discussion

Thanks for the correction.
73, Doug.

b...@lysator.liu.se wrote:

Doug,


I ran into a wiki description of GPS using WGS84 a couple days ago. It
included a mention of ESEC and it was something like: Earth Static,
Earth Centric. I think I was following a link about the Z3801A.

It referred to the fact that lat-lon is referenced to a static grid on
the Earth and doesn't change as the Earth rotates on its axis or around
the sun. Earth static and Earth centric.


That reference frame is called Earth Centered Earth Fixed  (ECEF). And
that is what GPS is using internally. However it is not lat/lon/h, but
cartesian with origo in the earth center, xy in the equatorial plane and z
pointing towards the north pole.

 http://en.wikipedia.org/wiki/ECEF

--

Björn




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A

2012-08-03 Thread Christoph Kopetzky
Azelio Boriani azelio.boriani@... writes:

 
 Here:
 www.leap*second*.com/museum/*hp58503*a/097-58503-12-iss-1.pdf
 you can find that there is the option 002 for the even second but they call
 it PP2S...
 Try to discover if your 58503 has this option 002.
 

Azelio,

thanks for the link but I already know this manual.
And for remembering I was talking about my 58503A Receiver. This receiver
do not have a PP2S Output. Only the TTL Alarm and the ESEC (?)

And as I wrote in my posting above, I only can see with my analog scope a 
little 
periodic signal with a level of approximately 28-30 mV and a triggered frequency
of 56,3 kHz at a timebase setting of 5µs/div.
If I am changing my timebase to 10µs/div and measuring the frequency I am 
getting
65,4 kHz.
This is a little bit confusing and seems to be reasoned from the little signal 
level...

So now I am more curious what signal the ESEC seems to be


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] HP 58503A

2012-08-03 Thread Christoph Kopetzky
Hello followers,

I can state that my measuring of the signal (30 mV, 65,2 kHz) was caused by
the serial line of the com port of the receiver... :-)

After looking for sources of the noisy signal I disconnected my first the 
GPS antenna signal from the 58512A GPS amplifier. - Signal intact
Then I disconnect my frequency meter from the 10 MHz output - signal intact
Now I disconnected the serial connection to my notebook - signal gone!

So it was an interspersed signal from the serial connector ...

But at the end the question what the ESEC output may be is still mysterious...

Chris


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A

2012-08-03 Thread Hal Murray

dekag...@gmail.com said:
 with my analog scope (Hameg HM2005-2) I can trigger a low level voltage
 spike signal with  a p-p voltage of 30 mV and a frequency of 52,6 kHz. Could
 that be real? The signal level is very low, 30 mV... 

That sounds like junk pickup.  My guess would be from a switching power 
supply.

Any real signal that I've seen is at least several volts.



-- 
These are my opinions.  I hate spam.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A

2012-08-03 Thread Christoph Kopetzky
Hal Murray hmurray@... writes:

 
 
 dekagoon@... said:
  with my analog scope (Hameg HM2005-2) I can trigger a low level voltage
  spike signal with  a p-p voltage of 30 mV and a frequency of 52,6 kHz. Could
  that be real? The signal level is very low, 30 mV... 
 
 That sounds like junk pickup.  My guess would be from a switching power 
 supply.
 
 Any real signal that I've seen is at least several volts.
 

Hal, you are very right,

in the meantime I found that the junk came over the serial connection from my 
notebook to the 58503A...

Regards

Chris



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A

2012-08-02 Thread Christoph Kopetzky
Azelio Boriani azelio.boriani@... writes:

 
 Why are you assuming this is a protocol or signal label? Maybe a
 maintenance or repair-by-exchange indication too.
 


Because the label is above the bnc connector on the back side of the 58503A 
beside the bnc connector with an alarm label.
The 58503A has the single option H14. May be this is a special output protocol 
on this connector like the IRIG protocol...

Any ideas?

Chris



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A

2012-08-02 Thread Azelio Boriani
OK, it is in the place of the PPS output... can you hook an oscilloscope
and see if there is any output?

On Thu, Aug 2, 2012 at 10:25 AM, Christoph Kopetzky dekag...@gmail.comwrote:

 Azelio Boriani azelio.boriani@... writes:

 
  Why are you assuming this is a protocol or signal label? Maybe a
  maintenance or repair-by-exchange indication too.
 


 Because the label is above the bnc connector on the back side of the 58503A
 beside the bnc connector with an alarm label.
 The 58503A has the single option H14. May be this is a special output
 protocol
 on this connector like the IRIG protocol...

 Any ideas?

 Chris



 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A

2012-08-02 Thread Peter Bell
My guess is that it's a PP2S / Even Second output - that would also match
with the label...

On Thu, Aug 2, 2012 at 5:02 PM, Azelio Boriani azelio.bori...@screen.itwrote:

 OK, it is in the place of the PPS output... can you hook an oscilloscope
 and see if there is any output?

 On Thu, Aug 2, 2012 at 10:25 AM, Christoph Kopetzky dekag...@gmail.com
 wrote:

  Azelio Boriani azelio.boriani@... writes:
 
  
   Why are you assuming this is a protocol or signal label? Maybe a
   maintenance or repair-by-exchange indication too.
  
 
 
  Because the label is above the bnc connector on the back side of the
 58503A
  beside the bnc connector with an alarm label.
  The 58503A has the single option H14. May be this is a special output
  protocol
  on this connector like the IRIG protocol...
 
  Any ideas?
 
  Chris
 
 
 
  ___
  time-nuts mailing list -- time-nuts@febo.com
  To unsubscribe, go to
  https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
  and follow the instructions there.
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A

2012-08-02 Thread Azelio Boriani
OK, so a digital 'scope is needed or, at least, with memory.

On Thu, Aug 2, 2012 at 1:12 PM, Peter Bell bell.pe...@gmail.com wrote:

 My guess is that it's a PP2S / Even Second output - that would also match
 with the label...

 On Thu, Aug 2, 2012 at 5:02 PM, Azelio Boriani azelio.bori...@screen.it
 wrote:

  OK, it is in the place of the PPS output... can you hook an oscilloscope
  and see if there is any output?
 
  On Thu, Aug 2, 2012 at 10:25 AM, Christoph Kopetzky dekag...@gmail.com
  wrote:
 
   Azelio Boriani azelio.boriani@... writes:
  
   
Why are you assuming this is a protocol or signal label? Maybe a
maintenance or repair-by-exchange indication too.
   
  
  
   Because the label is above the bnc connector on the back side of the
  58503A
   beside the bnc connector with an alarm label.
   The 58503A has the single option H14. May be this is a special output
   protocol
   on this connector like the IRIG protocol...
  
   Any ideas?
  
   Chris
  
  
  
   ___
   time-nuts mailing list -- time-nuts@febo.com
   To unsubscribe, go to
   https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
   and follow the instructions there.
  
  ___
  time-nuts mailing list -- time-nuts@febo.com
  To unsubscribe, go to
  https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
  and follow the instructions there.
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A

2012-08-02 Thread Christoph Kopetzky
Azelio Boriani azelio.boriani@... writes:

 
 OK, so a digital 'scope is needed or, at least, with memory.
 
 On Thu, Aug 2, 2012 at 1:12 PM, Peter Bell bell.peter@... wrote:
 
  My guess is that it's a PP2S / Even Second output - that would also match
  with the label...
 

OK Azelio, good idea, I will check that if I am back in the lab...

EverySECond... makes sense because the ESEC label really is over the
original PPS labeling..

I will keep you up-to-date!

Thanks for the answers!

Chris



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A

2012-08-02 Thread Peter Bell
If it is, then you should see a pulse every 2 seconds, lined up with the
even seconds in GPS time.  This might seem a rather strange signal to
provide, but it's what the IS-95 derived CDMA systems use to trigger the
start of another 100 block signalling frame.  It also tends to suggest that
the unit you have was originally being used as part of a CDMA system.

Regards,

Pete


On Thu, Aug 2, 2012 at 8:00 PM, Christoph Kopetzky dekag...@gmail.comwrote:

 Azelio Boriani azelio.boriani@... writes:

 
  OK, so a digital 'scope is needed or, at least, with memory.
 
  On Thu, Aug 2, 2012 at 1:12 PM, Peter Bell bell.peter@... wrote:
 
   My guess is that it's a PP2S / Even Second output - that would also
 match
   with the label...
  

 OK Azelio, good idea, I will check that if I am back in the lab...

 EverySECond... makes sense because the ESEC label really is over the
 original PPS labeling..

 I will keep you up-to-date!

 Thanks for the answers!

 Chris



 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A

2012-08-02 Thread Hal Murray

 My guess is that it's a PP2S / Even Second output - that would
 also match with the label...

azelio.bori...@screen.it said:
 OK, so a digital 'scope is needed or, at least, with memory. 

We discussed this area a week or three ago.

You don't need a digital scope to determine if there is a pulse.  A digital 
scope may help to see the pulse and figure out what it looks like.

With an analog scope, you can either look at the blinking light that tells 
you it's triggering, or you can reduce the sweep speed until you can easily 
see the (flat) line from the beam each time it triggers.

I can see a 10 microsecond pulse with my old Tek 465.  It blinks and I 
roughly remember what the picture looks like.  If I want to know a detail, I 
have to look at the right spot and wait for the next pulse.


-- 
These are my opinions.  I hate spam.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] HP 58503A

2012-08-01 Thread Christoph Kopetzky
Hi all,

I have a question on my 58503A:

no problems aside, it is running for several years without
any problems.
But now I am cleaning my devices and I am reading the label
ESEC again on the back.
So the question comes up again, what is ESEC?
Google knows nothing about the signal or protocol...

I am now curious again and finally want to know what is ESEC?

Hope someone can illuminate me with this question.

Best regards and thanks in advance ;-)

Chris


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A

2012-08-01 Thread Azelio Boriani
Why are you assuming this is a protocol or signal label? Maybe a
maintenance or repair-by-exchange indication too.

On Wed, Aug 1, 2012 at 11:40 PM, Christoph Kopetzky dekag...@gmail.comwrote:

 Hi all,

 I have a question on my 58503A:

 no problems aside, it is running for several years without
 any problems.
 But now I am cleaning my devices and I am reading the label
 ESEC again on the back.
 So the question comes up again, what is ESEC?
 Google knows nothing about the signal or protocol...

 I am now curious again and finally want to know what is ESEC?

 Hope someone can illuminate me with this question.

 Best regards and thanks in advance ;-)

 Chris


 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A Dim Display Replacement

2012-07-25 Thread Stan
Hi Jeff,

The VFD module is identical to the ones used in the 53131A and 53132A
counters. It is a part of the A2 input assembly in the counters (p/n
53131-60002). The good news is that Agilent still has 7 new and tested units
in stock. The bad news is that the list price is $756.00. Even though all
you really need is the actual glass VFD module, it's not available from
Agilent as a separate part.

Regards,
Stan

2. HP 58503A Dim Display Replacement (jeffh...@comcast.net)

Message: 2
Date: Wed, 25 Jul 2012 00:29:02 + (UTC)
From: jeffh...@comcast.net
To: time-nuts@febo.com
Subject: [time-nuts] HP 58503A Dim Display Replacement
Message-ID:

1924087406.102486.1343176142270.javamail.r...@sz0079a.emeryville.ca.mail.co
mcast.net

Content-Type: text/plain; charset=utf-8

Hi, 

I have an older HP 58 503A with option 001 display but it is very dim from
age. 

Does anyone know of a replacement display IC for this? 

Thanks 
Jeff 




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 58503A Dim Display Replacement

2012-07-25 Thread Tom Knox

Hi;
If you can find a part number for the VFD only you may be able to purchase 
them, but you would need a part number. In the recent past have purchased 3458A 
VFD displays alone with the part number for about $100ea , but I believe if you 
ask parts ID they will tell you that 3458A VFD are not available separately. 
Agilent is really good compared to some companies but in some cases it really 
depends on who you talk to. 
Sometimes you can go directly to the display manufacturer. Let us know if you 
find a source for the VFD.
Best Wishes;
Thomas 

 From: swp...@earthlink.net
 To: time-nuts@febo.com
 Date: Wed, 25 Jul 2012 17:37:52 -0700
 Subject: Re: [time-nuts] HP 58503A Dim Display Replacement
 
 Hi Jeff,
 
 The VFD module is identical to the ones used in the 53131A and 53132A
 counters. It is a part of the A2 input assembly in the counters (p/n
 53131-60002). The good news is that Agilent still has 7 new and tested units
 in stock. The bad news is that the list price is $756.00. Even though all
 you really need is the actual glass VFD module, it's not available from
 Agilent as a separate part.
 
 Regards,
 Stan
 
 2. HP 58503A Dim Display Replacement (jeffh...@comcast.net)
 
 Message: 2
 Date: Wed, 25 Jul 2012 00:29:02 + (UTC)
 From: jeffh...@comcast.net
 To: time-nuts@febo.com
 Subject: [time-nuts] HP 58503A Dim Display Replacement
 Message-ID:
   
 1924087406.102486.1343176142270.javamail.r...@sz0079a.emeryville.ca.mail.co
 mcast.net
   
 Content-Type: text/plain; charset=utf-8
 
 Hi, 
 
 I have an older HP 58 503A with option 001 display but it is very dim from
 age. 
 
 Does anyone know of a replacement display IC for this? 
 
 Thanks 
 Jeff 
 
 
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
  
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] HP 58503A Dim Display Replacement

2012-07-24 Thread jeffhook



Hi, 



I have an older HP 58 503A with option 001 display but it is very dim from age. 

Does anyone know of a replacement display IC for this? 