Re: [time-nuts] 3.0GHz Channel 3 installation in Agilent 53132A counter

2014-03-19 Thread Randy D. Hunt

On 3/19/2014 8:26 AM, James Robbins wrote:

Thanks for everyone's help here.  I have done some further sleuthing.

I have a second 53132A which has a 3.0GHz Channel 3.  So, I switched the new 
channel 3 board into the old 53132A and it reads properly.  Then, I switched 
the old 3.0GHz Channel 3 board into the new 53132A and it also read four times 
(4x) the input frequency.  From this I conclude that the issue is with the new 
53132A counter box and not the channel 3 board.

Then I powered up each counter and measured at the pins on the ribbon cable connector at 
the Channel 3 board to the main board on both the new and old counters (channel 3 boards 
plugged in and powered).  They each read the same voltages and/or grounds (pin1 = 2v; 
2,5,7,9,10 = ground; 3,8 = +12v; 4 = -12v with numbering based on 1 shown on 
channel board).

Someone suggested that maybe the new main board had been set up for the 12.5 or 
6GHz channel 3 but was sold without that channel.  The idea was that such a 
main board would cause a 4x reading.  To my mind this is opposite to what I 
would think in that the division ratio for 12.5 or 6GHz would be higher than 
the ratio for the 3.0GHz board and would result in a fraction of the frequency 
rather than 4x frequency.

Of course, it could be a FW mod (rather than a HW mod) which has been applied 
also.

Any other thoughts are much appreciated.

73,
Jim Robbins
N1JR
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Jim, I think the problem is with the Ch3 board as the issue migrates to 
the other counter with the board. Also the know good board works 
properly with the first counter that has issues with that board.  I 
don't think that it would be a firmware issue as the board from the 
working counter works properly in the other counter with the non-working 
board.  Hope this helps


Randy, KI6WAS
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] 3.0GHz Channel 3 installation in Agilent 53132A counter

2014-03-19 Thread Dave Martindale
No, the suggestion makes perfect sense.  A 12 GHz input board likely
has an internal prescaler with a division ratio that is 4 times that
of a 3 GHz input board, in order to have the main counter running at
the same internal frequency.  For example, a 3 GHz input might divide
by 32 while a 12 GHz input might divide by 128, so that both feed a
signal in the 0 to 100 MHz range to the main counter circuits.  (And
the 5 GHz input might divide by 64...)

When the firmware goes to display what the main counter measured, it
has to multiply by the prescaler ratio to obtain the original input
frequency.  If this multiply happens on the main board (e.g. in a FPGA
or CPU), it needs to be told which prescaler division ratio was used.
If the firmware thinks the 12 GHz input is installed instead of the 3
GHz input, it will multiply by 128 instead of 32, and the displayed
frequency will be 4 times the actual frequency (since the prescaler
actually divided by 32).

(The prescaler division ratios may not be 32 and 128; the actual
values don't matter.  As long as the 12 GHz option uses 4 times the
prescaler divisor of the 3 GHz option, the display error will be a
factor of 4).

- Dave

On Wed, Mar 19, 2014 at 11:26 AM, James Robbins jsrobb...@earthlink.net wrote:

 Someone suggested that maybe the new main board had been set up for the 12.5 
 or 6GHz channel 3 but was sold without that channel.  The idea was that such 
 a main board would cause a 4x reading.  To my mind this is opposite to what I 
 would think in that the division ratio for 12.5 or 6GHz would be higher than 
 the ratio for the 3.0GHz board and would result in a fraction of the 
 frequency rather than 4x frequency.
 and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] 3.0GHz Channel 3 installation in Agilent 53132A Counter

2014-03-19 Thread Brian Inglis

On 2014-03-19 16:00, James Robbins wrote:

I think I may need to clarify my prior posting.

Both old and new Channel 3 boards work fine (and read correctly) in my older 
53132A counter.

Both old and new Channel 3 boards fail to work properly in (and read 4x actual) 
the newer 53132A counter.

So I think the problem is in the newer main 53132A box.

As far as my understanding of the division ratios, I understand my error now.  
Thank you Dave (and some others).

I have done some further testing and discovered one of HP/Agilent's little 
tricks.  The main counter board channel 3 ribbon cable socket pin numbering 
does not conform to the channel 3 board ribbon cable socket numbering (they are 
crisscrossed).  (I can supply the shifted numbers if anyone is interested.)

However, my search for suspected unwanted zero ohm resistors left in the new 
box (to set the counter up for a 5 or 12.5GHz channel 3) has resulted in no joy 
and only a large headache and sore eyes:(

Thanks everyone for trying to help here.  It was supposed to be a plug and 
play addition to my counter.  So much for the plan.


Presumably you looked for obvious changes between the
main boards?

Have you checked the FW ROM markings or could you dump
and compare the ROM contents?

Thought of trying a straight (uncrossed) ribbon cable
between the channel 3 board and the newer counter?
That's the kind of change that might be done to reduce
costs on newer model upgrades requiring a factory mod.

--
Take care. Thanks, Brian Inglis
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] 3.0GHz Channel 3 installation in Agilent 53132A Counter

2014-03-19 Thread Brooke Clarke

Hi Jim:

I've seen a similar problem when one end of a ribbon cable connector is 
soldered to the wrong side of a PCB.
Maybe you can unsolder a connector and install it on the other side?

Have Fun,

Brooke Clarke
http://www.PRC68.com
http://www.end2partygovernment.com/2012Issues.html

James Robbins wrote:

I think I may need to clarify my prior posting.

Both old and new Channel 3 boards work fine (and read correctly) in my older 
53132A counter.

Both old and new Channel 3 boards fail to work properly in (and read 4x actual) 
the newer 53132A counter.

So I think the problem is in the newer main 53132A box.

As far as my understanding of the division ratios, I understand my error now.  
Thank you Dave (and some others).

I have done some further testing and discovered one of HP/Agilent's little 
tricks.  The main counter board channel 3 ribbon cable socket pin numbering 
does not conform to the channel 3 board ribbon cable socket numbering (they are 
crisscrossed).  (I can supply the shifted numbers if anyone is interested.)

However, my search for suspected unwanted zero ohm resistors left in the new 
box (to set the counter up for a 5 or 12.5GHz channel 3) has resulted in no joy 
and only a large headache and sore eyes:(

Thanks everyone for trying to help here.  It was supposed to be a plug and 
play addition to my counter.  So much for the plan.

Jim Robbins
N1JR
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] 3.0GHz Channel 3 installation in Agilent 53132A Counter

2014-03-19 Thread Tom Knox
Jim the 5 and 12.8GHz modules are the same part and the software/firmware set 
the upper freq limit. I assume at least the 12.8 GHz option must be 
unlocked/accessed most likely through the key pad. The 5GHz may be plug and 
play.

Thomas Knox



 From: jsrobb...@earthlink.net
 Date: Wed, 19 Mar 2014 18:00:09 -0400
 To: time-nuts@febo.com
 Subject: [time-nuts] 3.0GHz Channel 3 installation in Agilent 53132A Counter
 
 I think I may need to clarify my prior posting.
 
 Both old and new Channel 3 boards work fine (and read correctly) in my older 
 53132A counter.
 
 Both old and new Channel 3 boards fail to work properly in (and read 4x 
 actual) the newer 53132A counter.
 
 So I think the problem is in the newer main 53132A box.
 
 As far as my understanding of the division ratios, I understand my error now. 
  Thank you Dave (and some others).
 
 I have done some further testing and discovered one of HP/Agilent's little 
 tricks.  The main counter board channel 3 ribbon cable socket pin numbering 
 does not conform to the channel 3 board ribbon cable socket numbering (they 
 are crisscrossed).  (I can supply the shifted numbers if anyone is 
 interested.)
 
 However, my search for suspected unwanted zero ohm resistors left in the new 
 box (to set the counter up for a 5 or 12.5GHz channel 3) has resulted in no 
 joy and only a large headache and sore eyes:(  
 
 Thanks everyone for trying to help here.  It was supposed to be a plug and 
 play addition to my counter.  So much for the plan.
 
 Jim Robbins
 N1JR
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
  
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] 3.0GHz Channel 3 installation in Agilent 53132A Counter

2014-03-19 Thread John Allen
FYI:  Here is the link to the 53132 CLIP.

http://www.home.agilent.com/agilent/redirector.jspx?action=refcname=AGILENT
_EDITORIALckey=1119025lc=engcc=USnfr=-33788.536880944.00

It wasn't there last I checked.

John K1AE


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] 3.0GHz Channel 3 installation in Agilent 53132A counter

2014-03-18 Thread Brian Inglis

On 2014-03-18 12:19, James Robbins wrote:

I hope this is not too OT.
I have acquired a 3.0GHz original Channel 3 board for my Agilent 53132A
counter.  When installed, it reads four times (4x) the actual frequency.
I'm wondering if anyone has any experience with such an installation and can
point me to some setup item or zero ohm resistor I'm supposed to install.
All that the manual says is to have it done by the factory.  That will cost
me more than the board itself.  Many thanks.
Jim Robbins
N1JR


According to:
http://cp.literature.agilent.com/litweb/pdf/53132A-04A.pdf
early serial numbers may need updated firmware.

--
Take care. Thanks, Brian Inglis
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] 3.0GHz Channel 3 installation in Agilent 53132A counter

2014-03-18 Thread GandalfG8
Hi Jim
 
According to the schematic in the 53132A service manual and the datasheet  
for the MB510 prescaler, the option 030 board is hard wired for a division 
ratio  of 1/128, and that's the lowest division ratio shown for that  
prescaler.
 
I've not fitted one of the original HP boards but have used similar  boards 
from a third party supplier and they just dropped straight in.
 
As a quick first check I would suggest monitoring the output from the  
MB510, either on the 030 board or at the ribbon cable, just to confirm that the 
 
board is dividing as expected.
 
There is an HP Service Note 53132A-04A that discusses the firmware  
requirements for updating which also might be worth checking just in case 
that's  
an issue.
 
Regards
 
Nigel
GM8PZR
 
 
In a message dated 18/03/2014 18:20:09 GMT Standard Time,  
jsrobb...@earthlink.net writes:

I hope  this is not too OT.



I have acquired a 3.0GHz original Channel 3  board for my Agilent 53132A
counter.  When installed, it reads four  times (4x) the actual frequency.
I'm wondering if anyone has any experience  with such an installation and 
can
point me to some setup item or zero ohm  resistor I'm supposed to install.
All that the manual says is to have it  done by the factory.  That will cost
me more than the board  itself.  Many thanks.



Jim  Robbins

N1JR

___
time-nuts  mailing list -- time-nuts@febo.com
To unsubscribe, go to  https://www.febo.com/cgi-bin/mailman/listi
nfo/time-nuts
and follow the  instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] 3.0GHz Channel 3 installation in Agilent 53132A counter

2014-03-18 Thread Tom Knox
I think the solution will be in the firmware. It must be factory set so default 
is the 5GHz C channel option. The 12GHz option is the same part as the 5GHz 
sampler with a different firmware lock.
Best Wishes;
Thomas Knox



 From: jsrobb...@earthlink.net
 To: time-nuts@febo.com
 Date: Tue, 18 Mar 2014 14:19:14 -0400
 Subject: [time-nuts] 3.0GHz Channel 3 installation in Agilent 53132A counter
 
 I hope this is not too OT.
 
  
 
 I have acquired a 3.0GHz original Channel 3 board for my Agilent 53132A
 counter.  When installed, it reads four times (4x) the actual frequency.
 I'm wondering if anyone has any experience with such an installation and can
 point me to some setup item or zero ohm resistor I'm supposed to install.
 All that the manual says is to have it done by the factory.  That will cost
 me more than the board itself.  Many thanks.
 
  
 
 Jim Robbins
 
 N1JR
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
  
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] 3.0GHz Channel 3 installation in Agilent 53132A counter

2014-03-18 Thread GandalfG8
Further checking the service manual indicates that the main board  treats 
pins 9 and 10 of the ribbon cable as CH3code1 and CH3 code2 respectively  so 
perhaps it's using these as identifiers and it might be worth checking that  
those signals are as expected. The prescaler board schematic shows 9 as 
being  connected to +12V and 10 to ground.
If you don't have access to the component level service manual let me know  
and I'll send you a copy.
 
Regards
 
Nigel
GM8PZR
 
 
 
 
In a message dated 18/03/2014 18:20:09 GMT Standard Time,  
jsrobb...@earthlink.net writes:

I hope  this is not too OT.



I have acquired a 3.0GHz original Channel 3  board for my Agilent 53132A
counter.  When installed, it reads four  times (4x) the actual frequency.
I'm wondering if anyone has any experience  with such an installation and 
can
point me to some setup item or zero ohm  resistor I'm supposed to install.
All that the manual says is to have it  done by the factory.  That will cost
me more than the board  itself.  Many thanks.



Jim  Robbins

N1JR

___
time-nuts  mailing list -- time-nuts@febo.com
To unsubscribe, go to  
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the  instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] 3.0GHz Channel 3 installation in Agilent 53132A counter

2014-03-18 Thread J. Forster
Jim,

You were looking for a 0 Ohm resistor.

Sometimes HP did not, in fact, insert a 0 Ohm resistor component in later
revs of some PCBs. They just added the default 0 Ohmers in etch, between
two vias. Then, if someone wanted to open the link, the etch was cut. I
guess it saved a couple of pennies.

-John

===


 Further checking the service manual indicates that the main board  treats
 pins 9 and 10 of the ribbon cable as CH3code1 and CH3 code2 respectively
 so
 perhaps it's using these as identifiers and it might be worth checking
 that
 those signals are as expected. The prescaler board schematic shows 9 as
 being  connected to +12V and 10 to ground.
 If you don't have access to the component level service manual let me know
 and I'll send you a copy.

 Regards

 Nigel
 GM8PZR




 In a message dated 18/03/2014 18:20:09 GMT Standard Time,
 jsrobb...@earthlink.net writes:

 I hope  this is not too OT.



 I have acquired a 3.0GHz original Channel 3  board for my Agilent 53132A
 counter.  When installed, it reads four  times (4x) the actual frequency.
 I'm wondering if anyone has any experience  with such an installation and
 can
 point me to some setup item or zero ohm  resistor I'm supposed to install.
 All that the manual says is to have it  done by the factory.  That will
 cost
 me more than the board  itself.  Many thanks.



 Jim  Robbins

 N1JR

 ___
 time-nuts  mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the  instructions there.

 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.