Thank you very much! This makes sense, and the converted ECG signal I
obtain seems meaningful!

Thanks again,

raffaele

2012/4/2 Benjamin Kuris <[email protected]>

> A quick treatment is:
> ADC = 3.0V fullscale.
> 3.0/4095 = volts per bit.
> Or you can also use a ratio:
> ADC (bits) / 4095 (bits) = ADCvolts (V) /3.0 (V)
>
> Then you look at the ECG datasheet and above by gain to figure out
> volts on the electrode input.
> For more detail read section 9 of the manual which covers some sources of
> error.
>
> -Ben
>
>
> See section 9 of the user manual which describes how to get "volts"
> from the ADC bit value.
>
> On Mon, Apr 2, 2012 at 12:37 PM, Raffaele Gravina
> <[email protected]> wrote:
> > Dear all,
> >
> > this might be a silly question to some of you, but I wonder how I can
> > convert (from RAW ADC values to Volt) the two ECG signals I get from the
> > Shimmer2R equipped with the 4-leads ECG sensor-board.
> >
> > Any help would be greatly appreciated!
> >
> > Best,
> >
> > Raffaele
> >