Florian,

Looked at our MSE conversions using MassWolf and we don't have any
lockspray data. The conversion ignores function 3 (lockspray) from the
RAW data. The regular pattern still looks suspicially like lockspray
though, maybe MassWolf not recongising the function as lockspray? I
guess that if the spectra look like calibration compound that's what
it is, otherwise I have no idea. What instrument / MassLynx version is
it?

On 5 Apr, 17:50, Dave Trudgian <[email protected]> wrote:
> Florian,
>
> Is this due to lockspray scans? If you look at the spectra of the
> "extra" low energy scans, are they all very similar, i.e. showing the
> calibration compound? The scan numbering looks a lot like it could be. I
> don't have any masswolf conversions of our MSE data to hand, but I will
> check tomorrow and let you know.
>
> DT
>
> On 05/04/2010 15:21, fmg85 wrote:
>
> > Dear developers, users and especially Rune :-)
>
> > Masswolf seems to have a problem with the low energy MSE data in MSE
> > files. I should have 291 scans in both, the low and high energy scans
> > (I checked this in Masslynx and also parsing the via Water's
> > Databridge converted netCDF files).  The high energy data has 291
> > scans, the low however changes to 365.
> > If I look at the acquisitionNumbers then I can see that every other
> > (5.) scan the low energy function "steels" a scan from the high energy
> > function as you can clearly see here:
> > low:  1 2 4 6 8 10 11 13 15 17 19 20 22 24 26
> > high: 3 5 7 9 12 14 16 18 21 23 25 27
> > You can also see that the file looks funny in the Insilicos viewer
> > (random scan every other value)
> > If I use waters Databridge to convert my MSE into two datafiles
> > (netCDF), then convert them back into two waters .raw functions (high/
> > low energy file) and use masswolf to individually convert them into a
> > mzXML function the scans are correct (both have 291 each) and look
> > normal in the insilicos viewer. Centroiding does not have any
> > influence on this.
> > Does anyone have an idea what this could be caused? I verified this on
> > different MSE files and all software is up2date.
>
> > Best wishes,
>
> > Florian

-- 
You received this message because you are subscribed to the Google Groups 
"spctools-discuss" group.
To post to this group, send email to [email protected].
To unsubscribe from this group, send email to 
[email protected].
For more options, visit this group at 
http://groups.google.com/group/spctools-discuss?hl=en.

Reply via email to