Hi Rick:

I agree.

It's just that my FTS4060 IS drifting.
The Excel equation and fit for a month of data as of this morning are:
y = -1.2954x2 + 243.1x - 10631
R2 = 0.9607
Where y is in ns, x is the Day of Year and R2 is the quality of the fit (very good).


The first derivative (after dividing by 86400) gives:
Offset = -3E-14 * X + 243
i.e. minus 3E-14 per day.

I'll keep watching and post from time to time, maybe this will stabilize???

Have Fun,

Brooke Clarke, N6GCE

--
w/Java http://www.PRC68.com
w/o Java http://www.pacificsites.com/~brooke/PRC68COM.shtml
http://www.precisionclock.com



Richard (Rick) Karlquist (N6RK) wrote:

My understanding from talking to the experts is
that no aging, or environmental effects are observable
on the 5071A, and measurements have been made into
the 10^-15 range.  Also, no systematic frequency offset
has been observed down to at least 1x10^-14.  IE,
frequency offset between different 5071A's is randomly
distributed with a mean within 10-14 of laboratory
standards, such as Cs fountains.  Knowing the way we
overdesigned the 5071A, I am not surprised that it
is this good.

Rick Karlquist
(member of 5071A design team)