>Check the frequency multiplier chain alignment.

>Bruce

Thanks Bruce,

I used a HP5335A counter to measure the ref out at 10.000,000,3MHz and the 
200MHz test point-1
on the multiplier board as 200.000,007 MHz
The HP5335 is not calibrated but it looks like the multiplier is working O.K. I 
have no way to measure the harmonics as I do not have a RF spectrum analyser.
The test point has a nice 200MHz sine wave as tested with my eye and a 150MHz 
CRO.
I also adjusted the DAC voltages on A18 although they were close enough.
BTW I think the "set test switch" settings for the DAC module A18 may have a 
error.
I think the OA cal switch settings should be the same as the OO cal settings.

Could the harmonics from the multiplier be causing the problem?
Thanks & Regards Scott