[EMAIL PROTECTED] wrote:
> Hi David,
>  
> that HP puck runs it's crystal at over 100 Degrees C, and the enclosure  
> temperature of a typical OCXO gets to above 60C in still air.
>  
> They also have Tantalum caps, and anyone who has seen a fiery Tantalum cap  
> explosion knows these things can be very dangerous.
>  
> You may wake up dead, or castrated.
>  
> I would not put that thing near my groin if I were you!
>  
> bye,
> Said
>   
Said

The E1938A oven has exceptionally high thermal gain so variations in the
OCXO frequency due to ambient temperature changes may well be insignificant.
Before proposing drastic/dangerous measures David should attempt to
measure the E1938A frequency variations as a function if temperature to
see if the frequency change is actually significant.
Making such a measurement is likely to be an extremely
difficult/challenging  task. Even so measurement beats speculation
everytime.

Bruce