Martyn Smith wrote:
> Hi,
>
> I need some help on using the HP3048A Phase Noise Test Set (11848A, 3561A, 
> 3585A, 3325A).
>
> My company makes a ultra low noise frequency standard with ten outputs. 
> Apart from very low phase noise (-112 dBc/Hz @ 1 Hz with a -170 dBc/Hz noise 
> floor) I also claim all spurious to be < -120 dBc.  This includes the line 
> frequency spurious.
>
> I'm happy my test set is making accurate phase noise results as I agree with 
> other test results on sample oscillators to < 2 dB.
>
> But I seem to have an error on measuring spurs.
>
> As you all know, when the 3048A detects a big peak, it marks it in red and 
> shows the actual value, rather than normalizing it to a 1 Hz bandwidth.
>
> To test the 3048A accuracy I inject a spur at a known level using a 
> directional coupler (similar to the single-sided spur calibration method).
>
> I confirm the spur level on the spectrum analyzer.  However the 3048A always 
> says the spur is 6 dB lower than it actually is (6 dB plus/minus 0.5 dB).
>
> Can anyone tell me why this is so.
>
> I wondered if it had something to do with phase noise and amplitude noise 
> not being the same thing.
>
> Any response would be appreciated.
>
> Best Regards
>
> Martyn
>
> This Email is from:
>
> Martyn Smith
> Precision Test Systems LTD
> Tel: +44 (0) 1245 329608
> Email: [EMAIL PROTECTED]
> Web: www.ptsyst.com
>
>   
3dB since AM and PM components are equal plus another 3dB to correct
measurement from actual DSB to SSB??

Bruce