Also, if it's of any use, the HP-85 controller does an unlisten, my-talk-addr, listen-address-x for each read. There's no mention of the HP-85 doing an automatic group-execute-trigger, and none of the examples issue an explicit trigger.
There is one sentence in the 5334 manual that says a g-e-t will trigger a new measurement, which is what I would expect. Bill Ezell ---------- They said 'Windows or better' so I used Linux. wje wrote: > That does the trick, thanks. Interestingly, the HP manual doesn't > discuss group-execute-trigger at all. It's possible that the HP-85 > programmable controllers would do a trigger before a read; all of the > examples in the 5334A manual are written for that controller. > Bill Ezell > ---------- > They said 'Windows or better' > so I used Linux. > > Prologix wrote: > > You could try explicitly triggering with ++trg followed by ++read 10. > > > -----Original Message----- > From: [EMAIL PROTECTED] [[2]mailto:[EMAIL PROTECTED] On > Behalf Of wje > Sent: Wednesday, August 27, 2008 4:14 PM > To: Discussion of precise time and frequency measurement > Subject: Re: [time-nuts] Prologix GPIB-NET, HP 5370A, HP5334A problem? > > YI don't use read-after-write, and I explicitly set it off before > reading. However, I did try it once just for fun, and it does work as > expected. > > The 5334 doesn't seem to assert eoi, so I do a '++read 10', which works > just fine except for the WA1 strangeness. > > Bill Ezell > ---------- > They said 'Windows or better' > so I used Linux. > > > > Prologix wrote: > > Bill, > > How are you reading from 5334A? Is the read-after-write mode OFF? > I recommend the following sequence: > > ++auto 0 -- turn off read-after-write > ++read eoi -- read until EOI is asserted by the instrument > ++read eoi > .... > > (If your instrument does not assert EOI, specify a character to terminate > ++read command. See manual for details.) > > Yes, ++read will address the instrument to talk, read any output, and then > address it to listen. > > We are testing an update that resets the controller if it detects a hung > GPIB bus. I can send that to you offline. > > Regards, > Abdul > > > > -----Original Message----- > From: [EMAIL PROTECTED] [[4]mailto:[EMAIL PROTECTED] On > Behalf Of wje > Sent: Wednesday, August 27, 2008 3:12 PM > To: Discussion of precise time and frequency measurement > Subject: [time-nuts] Prologix GPIB-NET, HP 5370A, HP5334A problem? > > I just got one of the new networked GPIB controllers, and I've been > having some issues. I'm not sure if it's the Prologix or the > instruments, or both. > > With the 5370A, I can get samples for some period of time ranging from > about 2 to 4 hours before the GPIB controller stops responding. I > suspect this one might be the 5370A -488 controller bug I've heard of. > However, there is no way to recover without power-cycling the Prologix. > Even if this is the 5370A, the Prologix should really have an > asynchronous reset capability. > > With the 5334A, I can get samples for as long as I want, no problem. > However, if I set the 5334A to 'WA1' mode, which is supposed to wait > until the counter is addressed before taking each sample, what happens > is that I can read one value. Every successive read returns the same > value; the 5334A never triggers for a new sample. Either I don't > understand the behavior of WA1, although HP's sample program indicates > that what I've described is proper, or there is some incompatibility > between the Prologix and the 5334A. > > Unfortunately, the documentation for the Prologix is scanty at best. > There is no description of the actual IEEE-488 handshake process that's > going on. For example, does the Prologix do an unlisten after a read? > The HP docs seem to indicate that in WA1 mode, a sample starts when the > 5334A is addressed. So, if it's never unlistened, perhaps it never takes > a new sample? > > Any thoughts or suggestions would be appreciated. > > > > _______________________________________________ > time-nuts mailing list -- [EMAIL PROTECTED] > To unsubscribe, go to > [6]https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- [EMAIL PROTECTED] > To unsubscribe, go to > [8]https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > > References > > 1. mailto:[EMAIL PROTECTED] > 2. mailto:[EMAIL PROTECTED] > 3. mailto:[EMAIL PROTECTED] > 4. mailto:[EMAIL PROTECTED] > 5. mailto:[email protected] > 6. https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > 7. mailto:[email protected] > 8. https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > _______________________________________________ > time-nuts mailing list -- [email protected] > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > > > _______________________________________________ time-nuts mailing list -- [email protected] To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.