Did you try the LF injection test?

On Sun, Oct 5, 2008 at 8:48 AM, Jeffrey Okamitsu <[EMAIL PROTECTED]> wrote:
> I've spent a good deal of time this weekend working with the 5061B I recently 
> acquired.
>
> The unit appears to be making stable 10 MHz.  I've compared it the the 
> Rubidium source I have and the two compare as expected.
>
> There are two problems:
>
> 1) Beam current is low - even after going carefully through the alignment 
> procedure outline in the service manual, the best I can get (Loop Open, Mod 
> Off) is a bit over 10 nA.
> 2) The "Continuous Operation" LED will not come on after pressing Logic 
> Reset.  I used the trouble shooting guide to diagnose the LED and it's 
> pointing to the circuit on the A14 board.  This is as far as I have gotten so 
> far. Note that the trouble shooting guide is suggesting that it might not be 
> a problem with beam current, but an electronics problem (the LED driver).
>
> Anyone have any insight?
>
> Thanks for reading.
>
> Jeff
>
> Jeffrey K. Okamitsu, PhD, MBA
> +1-609-638-5402
> _______________________________________________
> time-nuts mailing list -- [email protected]
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>