Hi Brooke,

Yes, I am aware of appendix B procedure. I am using that exact setup with the 
exception of how I feed A-Ch. Not sure why, but I get less jitter of 5-7ps when 
I use the "Source" Ref feature to feed the external reference derived 1KHz to 
the A-Ch. When I use the BNC "T" method, I see about 10-12ps jitter. For all 
practical purposes, I am setup like appendix B and I use the ARMing mode as 
described.

All that said. The problem is when I try to do back to back GPIB MEAS? reads of 
the counter, the request and reply datum and subsequent time stamp is not real 
predictable. The end goal here is not to adjust the UUT to an accurate 
frequency as covered in appendix B but rather to build a list of time 
deviations with a consistent sampling rate to feed to the Plotter program. If I 
use an ARM sample of 1000 (1.0 sec), I see pretty good stability, less at 100 
samples (0.1 sec) and really bad at 10 or 1 sample. This limits my Tau-0 to 0.1 
sec for a reasonable ADEV analysis.

Regards...
Don 
  ----- Original Message ----- 
  From: Brooke Clarke 
  To: Discussion of precise time and frequency measurement 
  Sent: Saturday, November 01, 2008 4:48 PM
  Subject: Re: [time-nuts] ADEV and Tau-0 guestion


  Hi Don:

  If you want to average many (100, 1000) reads in one second you need to use 
the 
  ARM function as described in appendix B of the PRS10 manual.
  http://www.prc68.com/I/PRS10.shtml

  Have Fun,

  Brooke Clarke
  http://www.prc68.com/P/Prod.html  Products I make and sell
  http://www.prc68.com/Alpha.shtml  All my web pages listed based on html name
  http://www.PRC68.com
  http://www.precisionclock.com
  http://www.prc68.com/I/WebCam2.shtml 24/7 Sky-Weather-Astronomy Web Cam

  Don @ True-Cal wrote:
  > Fellow Time-Nuts,
  > 
  > I am having great fun with Ulrich's EZGPIB and Plotter programs to automate 
my ADEV and TI measurements. Wow, what a nice set of programs, thanks Ulrich!
  > 
  > I use the SR620 TIC with a Fury board as an external reference. The Fury 
disciplines an 10811-60168 external oscillator. I can go unlocked to improve 
the range around Tau 100s if and when necessary. For a series of tests, I used 
an LPRO-101 10Mhz signal to drive B-Ch (Stop) of the SR620; the A-Ch (Start) 
was set to Ref. for a Zero-Crossing TIME measurement on the TIC. I streamlined 
the EZGPIB SR620 query program and experimented with counter settings to 
minimize the inevitable and inherent latencies of the computer layers, network, 
GPIB-Enet/100 bridge and the counter (counter being the worst). With the 
counter set to 100 samples and the 1KHz "Ref" being used as the START, I was 
expecting a new, 100 sample TI average, every 0.1 seconds. My first evidence of 
something not being ideal was embedded in the details of the EZGPIB output 
console and accompanying file. Sometimes there were 7, 8 or 9 samples per 
second of time and never 10. Also, the total time span of a large coll
  ection of samples was always slightly longer than the product of the sample 
rate and count. I used Excel to scan 18000, 0.1s TI samples to determine what 
the actual statistics might be:
  > 
  > Average = 0.122302796 sec
  > Min = 0.188015099 sec
  > Max = 0.108984648 sec
  > 
  > Since the ADEV function as well as Ulrich's Plotter program requires a 
constant Tau-0, I experimented with the nominal 0.1s and the real "average" of 
0.1223s Tau-0 setting and attached a graph that illustrates the variance across 
Tau. My question is; what is "acceptable" practice for defining Tau-0 when the 
likelihood of having a stable sampling interval is difficult. It was rather 
simple to specify a more accurate time sample interval once determined by the 
extra step of spreadsheet analysis and the effect on the results is obvious. 
But that is still, only an average. What about the effect of the deviation 
about the average value? It would seem that would be a much more complex issue 
to deal with.
  > 
  > See attached export or Plotter graphic.
  > 
  > Regards...
  > Don
  > 
  > 
  > ------------------------------------------------------------------------
  > 
  > _______________________________________________
  > time-nuts mailing list -- [email protected]
  > To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
  > and follow the instructions there.

  _______________________________________________
  time-nuts mailing list -- [email protected]
  To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
  and follow the instructions there.
_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to