Stanley Reynolds wrote: > From: Richard Moore <[EMAIL PROTECTED]> > To: [email protected] > Sent: Monday, December 1, 2008 4:53:23 PM > Subject: Re: [time-nuts] time-nuts Digest, Vol 53, Issue 8 > > > Yes my suggestion was to reduce the range not increase it. I was not sure if > you were using the 8 volt reference inside the OCXO or the on board 5V now or > as you say an external 8 volts. In any case using a lower reference voltage > does decrease the range and step size. With no way to adjust the 10Mhz > voltage point you could raise the lower limit by modifying the op amp to > offset above ground. If using Bruce's suggestion then switch between two > different positive regulated voltages and not regulated and ground with the > analog switch. > > I have several 10811's just don't know how good they are yet ... > > If you have 3 or more then you could use the 3 (or N) cornered hat technique to identify the best of them.
However this requires a significant amount of phase comparison equipment and adequate isolation between all the OCXOs. Its all too easy to have one or more of the oscillators injection lock if the OCXO buffer reverse isolation is inadequate. Bruce _______________________________________________ time-nuts mailing list -- [email protected] To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.