Thanks to everyone for their replies, and to Bob Raker for a scan from the 1990 
Vectron catalog.  

The unit is spec'd to run from 24V, +/-5%, with 12-28 V tolerable at varied 
output power..  The CO-22 indicates a higher output level and frequency. A 
standard unit should deliver +13 dBm into 50 Ohms at +24 V.  An L2 unit should 
deliver greater than +10 dBm into 50 ohms at 24 V.  This particular unit does 
not have the external voltage tune option.  Stability is spec'd at +/- 3 e10 ^ 
-8 over +15C to +35C (Temp Range A).  The "3" indicates an aging of 3 e10 ^ -8 
/ day, 5 e10 ^ -6 / yr.

The L2 suffix indicates an ultra low phase noise option, which is 10 dB better 
than the standard option.  Specs for standard and L2 options:

Offset  Standard        L2
kHz     dBc/Hz          dBc/Hz
0.1     -120            -130
1       -135            -145
10      -140            -150
50      -145            -155


The unit has never had wires soldered to it.  The output power measures +12 dBm 
at 100 MHz on my 8560 spectrum analyzer.  (An indicative but non-authoritative 
power measurement)  Dropping the input voltage to 15 V drops the output power 
to +10.5 dBm.  The startup current at 24 V was about 240 mA, dropping to about 
118 mA after warm-up.

I will have to look up the specs on the company Agilent E5500 phase noise test 
set; it is quite possible that this oscillator is spec'ed beneath its noise 
floor.

Regards,

John

John P. Wendler, PE
 
Sr. Principal Engineer
Harris Corporation
1011 Pawtucket Blvd
Lowell, MA 01853
 
(O) 978-442-4822
(F) 978-442-4239
(M) 978-729-3040 (New)

This e-mail, including any attachments, may be confidential, privileged or 
otherwise legally protected. It is intended only for the addressee. If you 
received this e-mail in error or from someone who was not authorized to send it 
to you, do not disseminate, copy or otherwise use this e-mail or its 
attachments. Please notify the sender immediately by reply e-mail and delete 
the e-mail from your system.


_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to