Bob Camp wrote:
Hi

If I randomly pick up a FE 5680A data sheet, I find that it's short term 
stability is 1.4/sqrt(Tau) x 10-11.  Since I never doubt anything I see on a 
data sheet, this immediately tells me I should get 1.4x10-12 at 100 seconds, 
and I only have to wait for 10,000 seconds to get to 1.4x10-13.

Since the temperature performance is at the 1x10-12 / C level, I would need a 
room that's stable to *much* better than 0.1 C over a 3 hour period to get 
there. I suspect that 0.01C might not be good enough ...

So here's the question:

Has anybody run any of the cheap rubidiums (FE or Efratom)  in a *very* stable 
temperature environment to see how close they get / what the floor is?  I've 
run through a lot of data on the web, but I haven't really found what I'm 
looking for.

Thanks!

Bob

Figure 7 on the FE5680 page (also on the data sheet) indicates that you may need somewhat less than 3hours to achieve ADEV ~1E-13. 0.01C stability should be adequate.however its not necessary to control the room temperature to this stability if the FE5680 is in an enclosure with a sufficiently high time constant whilst having a sufficiently low thermal resistance so as to avoid overheating the FE5680.