On 06/20/2010 11:53 PM, jimlux wrote:
Robert Benward wrote:
Bob
Boy, you guys are really making me read a lot. I'm digesting Wiki
right now.

I see tau, but does identifying a tau of 1E-14 allow you to say you
are locked to 10fs? The smallest tau I've seen in my E1938 collection
is 1E-1.

Bob


tau is the time over which the measurement is made, typically 1 second
or greater.

loosely speaking, the 1e-14 is the average fractional deviation of
frequency over that time period.

It is a RMS type (much like statisticians standard deviation) of frequency stability over the "observation interval" of tau (little greek letter looking similar but not quite like a little t, which is the real reason for using it). Since it is a RMS type of measure, it indicates the effective power of noise, but not what the actual deviation in frequency will be, it's just a statistical measure. You may form a confidence interval such as that for 99,7 % or something which forms a scale-factor, quite similar to the use of the error function for the Gaussian distribution.

An Allan deviation measure of 1E-14 is however not quite the same as 10 fs. Besides the units being wrong (Allan deviation is a relative and unit-less measure, essentially Hz/Hz) the Allan variance (and hence the Allan deviation) is a frequency stability measure, indicating the stability of normalized frequency rather than stability of normalized phase. The time deviation represents the stability of phase over some observation time. Assuming the nominal frequency and linear effects removed, then this would indicate the time error noise of the phase, here use of seconds could be used, but it would be to stretch things a bit.

The time and frequency world has it's own qualities of noise...

Cheers,
Magnus