Bob,

For dBm or Volts across 50 Ohms, I normally set it for 10K bridging and use a 50 Ohm terminating connector. 10K terminated by 50 Ohms is close to 49.75 Ohms, which is close enough for anything I need to do. I suppose if I really needed to get accurate, I'd take a file to a 50 Ohm resistor like we did in the good old days. I've never been sure why the 3586B always seems to be less expensive than the A version other than possibly because of the A being a 50 Ohm instrument.

I have a pdf service manual for the A,B & C instruments. Do you need a copy for working on your B? What kind of problems does your B have? I've repaired a few and have a little knowledge in repairing them.

Burt


From: "Robert Benward" <[email protected]>
Subject: Re: [time-nuts] HP-3586B receiver...


Burt
I believe the voltmeter is calibrated for 75 ohms, not 50, otherwise they
are very similar.  You can apply a dB offset to correct for the measurement
error,  run the input at 10Kohms, and use a 50ohm feedthrough termination.

I own both (the B is has problems now), but I use the voltage measurement
capability extensively.  The B is much cheaper on Ebay.

Bob


  ----- Original Message -----
  From: Burt I. Weiner

  Sent: Saturday, July 24, 2010 4:05 PM
  Subject: [time-nuts] HP-3586B receiver...


  I prefer the 3586-B version.  I think it's more user friendly - once
  you get the hang of it, it does more.

Burt I. Weiner Associates
Broadcast Technical Services
Glendale, California  U.S.A.
[email protected]
www.biwa.cc
K6OQK

_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to