Corby,

It might be easier to just make an adapter for Lecture Bottles for the H2
supply.

Also, the H2 probably does not have to be Research Grade. The inflow is
through a "Palladium Valve" which is a diffusion device ans only passes H2
at any significant rate.

Best,

-John

=============


> Paul,
>
> Yes it is operating.
>
> With the exception of some down times to replace each of the ion pumps
> when they failed (at separate times) it has been running nonstop for 3
> years.
>
> I also had to replace the disassociator oscillator with an
> oscillator/amplifier of my own design.
>
> Also replaced the original 1.4Ghz RF amp (2.8db NF) with a modern amp
> having a .4db noise figure.
>
> Had to refill the little hydrogen bottle once so far. (via a homebrew
> manifold and a tank of ultra pure H2)
>
> The local gas co. does provide the H2 but the adaptor for refilling is
> homebrew as the vintage European hydrogen bottle fittings are no longer
> available.
>
> I'll see if I can post a PIX or two of the Vacsorb pumping station and
> the ion pumps.
>
> Corby Dawson
> [email protected]
> ____________________________________________________________
> Obama Urges Homeowners to Refinance
> If you owe under $729k you probably qualify for Obama's Refi Program
> http://thirdpartyoffers.juno.com/TGL3141/4c7ebc25c1b6fdd40em04duc
>
> _______________________________________________
> time-nuts mailing list -- [email protected]
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>



_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to