The 2nd Harmonic is the only one showing variation

Battery         0
Supply          38
Lamp Oven       25
Cell Oven       23
Osc Oven        36
Photo I         31
5 Mhz           18 (loaded)
Control         0.5
Error           0
2nd Harm        31
1 Mhz           38 (no load)
100 khz         38 (no load)

Brian - KD4FM





On 10/29/2010 5:20 PM, John Miles wrote:
There are a couple of 20 uF axial tantalum electrolytics across the B-C
junctions of the pass transistors for both the lamp and cell heaters.  They
were open in my 5065A when I got it, and there was some overheating damage
on the lamp driver PCB.  Correlation != causation, but since I'm not sure
what those capacitors were actually supposed to do in the first place, I
think they're worth checking.

-- john, KE5FX

-----Original Message-----
From: [email protected] [mailto:[email protected]]on
Behalf Of paul swed
Sent: Friday, October 29, 2010 6:02 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] HP5065A Rubidium 2nd Harmonic Readings


John,
Interested. How might you connect the oven controller to what he
is seeing?
Bad caps causing it to oscillate, 60 hz modulation, ......
I have a 5065 so am always interested in tid bits that I read.
Thanks

On Fri, Oct 29, 2010 at 12:34 AM, John Miles<[email protected]>  wrote:

How's the photo current?  Does it vary as well?  If not, I
don't think I'd
be too worried about the physics package.

I would, however, check the electrolytics on the oven controller board.
  You
don't want an oven controller failure.

-- john, KE5FX

-----Original Message-----
From: [email protected] [mailto:[email protected]]on
Behalf Of Brian Kirby
Sent: Thursday, October 28, 2010 9:15 PM
To: precise time
Subject: [time-nuts] HP5065A Rubidium 2nd Harmonic Readings


I recently had a long term power failure and ran out of UPS time.  The
power was out over a day and when I returned home, it had
been restored.

The rubidium recovered, I hit the logic reset button and start and got
the green light.

I made some meter readings, and noticed something I have not seen
before.  The 2nd Harmonic reading is showing 30 and its bouncing up to
31 about every 2 to 3 seconds.

Otherwise, the unit seems to be performing nominal.

Anybody ever seen this ?  Hope its not trying to tell me
failure of the
cell is on the way...


Brian - KD4FM



_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to