Mark Spencer wrote:
Seasons greetings.
I was pleasantly surprised to find that a new to me HP 5328A had an 10811 in
it.    (I was expecting the older OCXO..)     After leaving it running for a
week I adjusted the frequency and while doing so I noticed that the out put
frequency changes by a third of a hz or so if I add or remove a 50 ohm load. I'm measuring the frequency at the timebase output connection of the
5328A.  Is this to be expected ?

Many thanks
Mark Spencer



_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

No, that seems to be a couple of orders of magnitude larger than the specs for the 10811A would lead one to believe.
Maybe there's a problem with the output buffer amp in the 10811A?

Bruce


_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to