Mark Spencer wrote:

I'm curious in knowing if there are any rules of thumb as to how much better a
reference oscillator needs to be than the device under test when measuring ADEV
?
To provide a bit of context to my question I'm using a HP 5370B to repeatedly
measure the time interval between the 10 Mhz signals from a FTS 1050 and
rubidium oscillator. Having compared the FTS 1050 to another OCXO in the past I'm confident that it
is meeting or exceeding it's spec of an ADEV of 1X10-12 or better at tau's of
say 100 thru 1000 seconds.   I'm also confident that at tau's of 100 seconds or
more the noise introduced by the 5370B is less than 1x10-12. The underlying question that I have is how much confidence could I put in
measurements that show the rubidium of having an ADEV of say less than 2.0X10-12
at a tau of 100 sec ?

This is just for a hobby (: and I'm not looking for any degree of certainty but
I'm curious in understanding how accurate the measurements might be and if
changes in the measurements might not be caused by changes in the device under
test. Thanks in advance for any replies. Regards
Mark Spencer


The reference oscillator needs to be at least a factor of 3 (preferably 10x) better for all Tau of interest. Unless of curse one uses a 3 (or N) cornered hat technique in which case the ADEV of the oscillators being compared may be similar for the Tau range of interest. However one needs to guard against correlations between oscillators due to injection locking, environmental fluctuations etc.

For Tau= 100s relatively simple low cost three cornered hat setups (with an instrumentation ADEV noise floor below 1E-12) are feasible (provided on has a suitable offset oscillator) using inexpensive digital logic plus a suitable microprocessor.
No complex interpolators are required.

Bruce

_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.