Antonio,

I suspected something like that.

I realized that I've never done a similar test so I'm correcting that now. Both the oscillator and the counter have been off for some time - days for the counter due to equipment rearrangements and months for the oscillator. I replaced the 04E oscillator with the 04A oscillator to see if the 04A's lower power drain would help a power problem I'm having. It didn't.

I installed the 04E oscillator and set up my 1992 to measure the frequency of my Efratom FRK rubidium which has been running continuously for about 3 months. I'm using a gate time of 10 seconds.

The first reading after turn-on was about 370 Hz (yes, Hertz) low. After 20 minutes the initial warm-up was complete and the reading was 238 counts (e.g. 0.238 Hz) above 10 MHz - well within the 1e-7 specification. Now, after about 13 hrs, the reading is 78 counts high. It's now dropping at about 1 count per hour, but there was a recent drop of 4 counts in one hour as it works out the kinks.

It'll take a while to match the drift rate of your graph. Then I'll check the standby drift again.

Ed


On 12/19/2011 10:49 AM, [email protected] wrote:
Hi Ed,

my mistake, the test started from cold, but the curve I've posted starts at a
certain midnight. For completeness, the very initial warm-up is missing in the
curve. Checking my archive, I see that the test started the previous day around
02:22 pm and the reading was 3 mHz higher, and I might have lost a few minutes
since turning on the oscillators and the beginning of the record. The test was
aimed at seeing how many days the Racal takes to stabilize around the 1E-10
level, and the answer was one week. And this is the overall sense of my curve.
Hence a few minutes less or more were not important.
But indeed this test is not a good indication of the very first warm-up
minutes (intentionally not from stby).

Antonio I8IOV

Hi Antonio,

Between you and Charles, I'm getting the feeling that my 04E oscillator
is defective.  Yours only moved 15 counts over a period of days from a
cold start.  Mine moved 40 points in two hours from standby.  I think I
have some testing to do.

Thanks,
Ed


On 12/18/2011 6:01 PM, [email protected] wrote:
As a contribution to this discussion, I attach a 17-day trace recorded with one 
of my 1992s on Feb 2008. The counter was counting a LPRO Rb. The test started 
from cold for both sets. Gate time was 10 sec.
The 1992 is the 02M/04E military.

Antonio I8IOV

_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to