Thanks Bob;
It would be great if these are all cal issues. The offset is most likely 
calibration related, but I will be surprised if the jitter in a cal issue. It 
is certainly possible. I am thinking a weak/noisy cap. 
Thomas Knox



> From: [email protected]
> Date: Sat, 9 Jun 2012 12:58:12 -0400
> To: [email protected]
> Subject: Re: [time-nuts] SR620 flicker.
> 
> Hi
> 
> There is a fairly detailed calibration procedure for the SR620. I don't 
> remember all the steps, but it does indeed reduce the reading jitter.
> 
> Bob
> 
> On Jun 9, 2012, at 11:49 AM, Tom Knox wrote:
> 
> > 
> > I have a SRS SR620 that when an input channel is connected to the Ext Ref 
> > out has "Flicker" of about 3 digits of jitter/noise/drift. My experience is 
> > that most counter monitoring their external reference will be stable to one 
> > count with only an occasional larger jump. Magnus told me he has a SR620 
> > that has the same instability problem, and that he heard this is a common 
> > problem for these units. Has anyone else encounter this problem? And of 
> > course the 2 dollar question, does anyone know how to repair this issue?
> > THANKS;
> > Thomas Knox
> > 
> > 
> >                                       
> > _______________________________________________
> > time-nuts mailing list -- [email protected]
> > To unsubscribe, go to 
> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> > and follow the instructions there.
> 
> 
> _______________________________________________
> time-nuts mailing list -- [email protected]
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
                                          
_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to