Hello Mike:
Many thanks for useful information. The problem is I have no more
information of my USB-GPIB controller, so I don't know any command of this
device. I used stupid way to send my command, if I want send "DI" to increase
the display digis, I will convert "DI" to HEX code, it's 0x44 and 0x49. I add
0x36 (0x36 is HP5386's address) and 0x0A in the begin and end of my command
sequence separately. So my command looks like "0x3644490A0D", the command was
worked.
but I don't know what command can set my counter to "talk" status, I am
also confused on how to "initiate its read cycle".
Hui
At 2012-06-17 23:45:07,"Michael Blazer" <[email protected]> wrote:
>Hui,
>The older instruments have a very simple command structure. You don't
>need the "?" to query. Typically just doing a GPIB Read command (ibread
>for NI controllers) will address the counter to output and it will send
>the current reading. This can cause a problem if you request data too
>soon after a setup command. You'll get the previous reading.
>The "ENTER S ; N$" command just reads from address "S" into the variable
>"N$".
>It sound like you need the command to your USB-GPIB controller to get it
>to read. The counter is probably ready to send the data, the controller
>just needs to initiate its read cycle.
>
>Mike
>
>On 6/17/2012 10:17 AM, Hui Zhang wrote:
>> I have a old HP5386A counter and a OEM USB-GPIB controller(no tech support),
>> I use a serial port debug program to send command via USB-GPIB to HP5386A,
>> when I send "FU2"(Measure period) or "DI"(increment display digis), the 5386
>> response the command and worked well, the other control command in the HP's
>> manual also worked well.
>>
>>
>> But I don't know how to read the measure value from 5385A, in my HP3478A
>> mutilmeter case, I just send a "?", it will return voltage value to
>> software, but at 5386A, the “?” not work, I got error message "52" on LCD
>> panel.
>>
>>
>> I read the HP5386's manual, the demonstrates used "ENTER" command to get
>> measure value, but the "ENTER" is a HP-BASIC command of vintage HP-85
>> computer. I don't know how to send similar command to control my counter,
>> can someone give me some advice? Thank you!
>>
>>
>> Hui
>> _______________________________________________
>> time-nuts mailing list -- [email protected]
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>
>
>
>_______________________________________________
>time-nuts mailing list -- [email protected]
>To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>and follow the instructions there.
_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.