If you haven't already done so, download the manual - there are multiple copies on the net - and run through the Operation Verification Tests in Table 4-1. All you need for that is a BNC cable. If it passes those tests you can be reasonably sure that everything is alive.

Next, I suggest running a noise floor test. Take the signal from any decent oscillator and route it to the Start input (high impedance) via a T-connector then through maybe a 3 to 5 meter (length not important) cable to the Stop input (50 ohm). You could also use a splitter and unequal cable lengths to the Start and Stop inputs (both 50 ohms). Measure the time from Start to Stop. Collect the data via GPIB if possible and see what the ADEV curve looks like by using something like TimeLab, Plotter, or Stable32. I can't find any 5370A results on the net, but they should be similar to the 5370B results shown here:

http://www.febo.com/pages/hp5370b/

Ed

On 8/28/2013 10:35 AM, cfo wrote:
I bought a 5370A on *bay last week , and the unit is supposed to arrive
tomorrow.

I have never used such a beast , and as the seller wrote it is fully
working , and passes self test. I was wondering if anyone could give some
hints to test that it's working ok.

I mean the input stages , known defects etc. ?

I have a Tbolt ,a PM6680 Counter , a cheap Rigol DG1022 Func Gen, a
working Linux-Gpib installation , and a Prologic GPIB-Ethernet adapter.

And for clock sources: A Morion MV89,some 10Mhz *bay #310380778466 5v
square out i think, some Pletronics 26Mhz OCXO's, a X72 Rubi, some FEI
Rubi's and some Jupiter GPS boards to test with.

But what do i do for a start ?
Besides opening it up and vaccum clean it ,set it for 240V, and check if
there's a OCXO in there ?
        


Also what are the input voltage limits ?
I'd rather not start by killing the input stages my self.
                                                        
I think i read around 1volt for the 5370A , and 3.3v for a 5370B ?

But i also think Bruce wrote that (maybe only the B) would accept 5v , as
the clamping diodes would then be active.

Somewhere i also read that a 5v signal via a 270R resistor would be ok on
the 5370A

Thank you for any hints/tips,warnings or help.

CFO - T-Nut Beginner Denmark

_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to