On 03/02/14 01:19, Azelio Boriani wrote:
Magnus, please, can you elaborate this:
> A common mistake is to assume you can average it out, but that gives you a 
different measure which does not
> represent the ADEV values you are comparing with. The time between samples 
will scale down the relative impact of
> the time-noise, but not really average it.
That is, have I to increase the time between samples or take samples
and average them? For example, go from samples every second to samples
every 2 seconds or average two samples every second to obtain a sample
every 2 seconds. The last method will (apparently) increase the
resolution...

What I am saying in the first sentence is that taking the average of samples does not give you improved limit, but just add filtering which makes low-tau ADEV values which is biased to look better than the real ADEV for that measurement is.

In the second sentence, I make the point that when you take two sample values, at various taus, you really do not average them but rather make their time stability contribution (trigger jitter and resolution) less important relative to the tau between them. This is by itself not an averaging effect.

Cheers,
Magnus
_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to