Hi

I believe that you have a problem with that probe. A broken ground is one 
possibility, a significant mismatch between the probe output and the scope 
input is another possibility. 

Bob

On Apr 17, 2014, at 1:32 AM, Claude Fender <lab...@yahoo.fr> wrote:

> Hello List,
> I've recently bought a Quartzlock 10A R (rubidium). Does anyone has 
> experience with this unit? 
> 
> My
> unit has two outputs with no options. The problem is the shape of the 
> waveform (see attachments), it's the same shape for the two outputs. 
> When I
> connect an output directly to the oscilloscope, the shape is nearly 
> "squared" and when I use a probe, the shape is nearly "sine" !
> 
> When I send a 10 MHz signal from a generator, the waveform is a clean sine 
> with or whithout the probe.
> 
> So what signal should I except from this rubidium ? Or most probably, what is 
> wrong with my use of the oscilloscope ?
> (The frequency is ok by the way (+2E-10) and I didn't open the case yet) 
> 
> Thanks for any advice
> 
> Claude
> <10AR-sm.jpg><noprobe.gif><withprobe.gif>_______________________________________________