Hi Tom,

My experience is the manufacturer's, of hp or fluke quality, reported error band (their spec sheet) is the worst case (or at least 3 sigma) provided that the device is operated according to the recommended environment, etc.

Nonetheless, my statement you quoted is still accurate as to the purpose of the specification. Although I probably should have qualified a bit more. It goes without saying that one should be careful when it comes to specmanship no matter who it is.

Bill....WB6BNQ

Tom Van Baak wrote:

The value of 5E^-11 refers to the resolution that the precision can be relied upon after taking into account all the factors that influence it. It means that there is an error that can be as much as +/-5 parts per e^-11.

No, it can be way more. A 5e-11 spec value is likely just RMS, or 1-sigma. 
Actual measurements will show significantly larger 2-sigma, 3-sigma, etc. value.

Your "as much as" wording sounds more like a 6-sigma, or peak-to-peak spec.

One must be very careful to match the measurement used to make the spec with 
how the device is intended to be used. For example, Magnus will likely tell us 
about MTIE, which is a perfect-storm, worst-case time error spec. It's very 
different from rough 1-sigma specs we usually talk about.

/tvb