If you have tube-life and not other issues, it's about the same.
Also works for rubidiums, as the loop aspect here is essentially the same.

There can be *other* issues. For the 5060A for instance, you might need to also adjust the crystal filter of the OCXO, as that too drifts out of range, so you get no signal out.

What I write is not a fix-it-all but rather addresses that one issue.

Cheers,
Magnus

On 08/28/2014 06:35 PM, Bob Bownes wrote:
Is there a similar 'bring it back to life' procedure for the 5061A?


On Thu, Aug 28, 2014 at 12:32 PM, Magnus Danielson <
[email protected]> wrote:

Chris,

Do you have a GPS clock?

First turn the operational mode setting from off to second step (ocxo +
ion pump) and let it stay there for a day or so.

Then, as the oven have stabilized, switch it over to third step, the open
loop mode, and tune the OCXO up against a GPS reference so that it is very
near 5 MHz. You use the calibration whole on the front of the clock for
that.

Then, you turn the operation mode switch to the fourth and last step, the
closed loop step, and see if it locks up. Let it just sit there and lock
up, as it takes some time.

It's quite common that OCXOs have drifted outside the capture range of the
analogue loop, so loosing lock and not being able to attain it again is a
typical response.

Cheers,
Magnus


On 08/28/2014 04:33 PM, Chris wrote:

On 08/28/14 05:03, Javier Herrero wrote:

Hello,

Here is the manual I've. I have also some other documentations, and some
Oscilloquartz software for the OSA-5585, but I don't know if they are
very useful.

Regards,

Javier


Hi Javier,

Thanks for that and for the other replies. The 3210 looks like quite an
early design, with no sign of microprocessors at all. There's a 8 slot
card cage with a load of discrete analog circuitry, 741 op amps etc and
a couple of boards full of 14 / 16 pin ssi cmos / ttl devices, which I
guess would be the synthesiser logic and perhaps a state machine style
startup sequencer. Apart from that, the rest is power supply related and
what looks like an alarm board with optoisolator discete outputs to a 25
way D connector. The step recovery diode (?) multiplier into the
microwave cavity is a really neat gold plated assembly with what looks
like a 50r termination (setup tap for spectrum analyser ?) and an
adjustment trimmer, but am not touching that or the many trimpots on the
boards or any adjustments until I have more info. The tube is from FTS,
part number / model 7101.

It seems strange that the 2nd harmonic, meter #9, is zero, since even
with a tube approaching eol, one would expect at least some indication,
which is why I think there may be an electronic fault. Perhaps the hv
power supply module feeding the electron multiplier. Will try to measure
that, but the area around the tube is really heavily rivetted and
screwed down in all directions. Looks like a lot of the left hand side
of the case will need to be disassembled just to get at the tube
connections. It also had the battery backup option, with 4 sets of 3 x
cyclon type cells, but with a date code of 1984, are seriously dead and
have been removed.

This time nuts things seems to be a growing interest and wonder if there
is a cure ? :-). Recently bought a 1970's era Tracor 304D rubidium
standard. Again, no lock, but a very well engineered and screwed
together piece of kit and should be fixable. Collection now includes the
Z3816, from Ebay US around 7 years ago, a Z3815 currently being
repackaged, an HP103 with open circuit oven heater elements and the
3210...

Regards,

Chris



  _______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


  _______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.

_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to