Karen wrote:

All OCXOs have been tested in free-running mode after 40-50 min of
switch-on/warming.

All OCXOs have been installed in isolated close box to exclude external
airflow.

Cases of all OCXOs had not any contact with external metallic parts of the
box.

Then it appears that something is wrong with your method of measuring the ADEV. I would suggest trying a few of the oscillators again (no need to try them all), using the Isotemp OCXO as the time base for the Pendulum by connecting it to the Pendulum's REF input and selecting External Reference from the front panel menus. This will tell you if there is a problem with the internal time base in the counter. [NOTE that the REF input needs a sinewave between 0.1 Vrms and 5 Vrms. If the Isotemp has a square-wave output, choose another OCXO that fulfills the Pendulum specification.]

If those tests produce ADEV measurements of 1e-11 or better at 1 second, the internal time base is too noisy for what you are trying to do. I expect that the internal time base will prove to be OK, but you need to verify that before proceeding.

However, I note that in the manual, the internal time base of the OPT 19/90 counter is rated for a Root Allan Variance of 1e-10 at tau = 1 second and 10 seconds (see p. 8-15 of the 2014 User's Manual, Rev.18). OPT 30/90 is rated at 1e-11, and OPT 40/90 is rated at 5e-12. So, you may be getting about all of the performance the OPT 19/90 can deliver, if its OCXO just meets specification. (HP5370Bs are also specified at 1e-10, but they often deliver ADEV = 1e-11 or 1e-12 at 1 second. I do not have enough experience with Pendulum counters to know what kind of stability is typical.)

Assuming that the test above (with the external REF) gives similar results to what you posted today, there is something wrong with your measurement setup. Are the input signals triggering the counter very stably near the middle of their peak-to-peak voltages? Is the counter adjusted to give maximum resolution for the sampling you are doing? Does your sample interval give the counter time to process each new sample? Have you read Chapters 3, 4, and 5 of the User's Manual carefully, and set up the counter properly for your measurements? Are you sure your TimeLab setup is correct? (I'll leave it to John to comment on that.)

Best regards,

Charles