I am maybe using the wrong word "discernible". By that I mean that you cannot 
discern phase noise from AM noise at the real low levels. You can certainly 
measure, or see the contribution of the total noise power, but do not 
necessarily know if it is phase noise or AM noise or how much of each is 
included. Regards - Mike 

Mike B. Feher, EOZ Inc.
89 Arnold Blvd.
Howell, NJ, 07731
732-886-5960 office
908-902-3831 cell


-----Original Message-----
From: time-nuts [mailto:time-nuts-boun...@febo.com] On Behalf Of Bob Camp
Sent: Thursday, January 15, 2015 7:48 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Current state of optical clocks and the definition of 
the second

Hi

I guess the question becomes how low is low. 

If it’s a 50 ohm system 

If the power level is rational

If you are at room temperature 

There are some limits on how low low can be. 

You have a -174 dbm  / Hz thermal floor. AM or PM noise can only be 3db better 
than the thermal floor. At a power level of 1 watt, that’s a -204 dbc / Hz 
limit. You will spend some time correlating to that level. You also may need to 
play a bit with the input circuits to handle the 1W without damage. At a 
somewhat more common 100 mw, the limit is -194. People have been measuring 
phase noise in the > -190 dbc / Hz range for at least 20 years now. Correlation 
may take a week at some offsets. Time will be longer or shorter at other 
offsets. As with anything else, the more money (correlation channels)  you 
throw at the problem, the quicker it will go.  Numbers in the -180 vicinity 
with normal gear, offsets, and FFT windows are an overnight run sort of thing. 