>Luke, the HP5335A was introduced in around 1980 and has an early >(pre-IEEE-488.2) HP implementation of IEEE-488 which requires a >terminator character at the end of each string. According to the manual, >this terminator can be a comma, semicolon, space, carriage return, or >line feed character.
>The delays are extremely important. The 5335A can and will hang if you >don't wait long enough. In keeping with that, *do not* set your adapter to >do an automatic read after sending data over the GPIB. Bill, Orin, Thank you! You solved my problem! It's working normally now! I was not sending a terminator at the end of the command and I had automatic read after send turned on. Setting to manual read, sending <CR> and then reading manually stopped the strange behavior. With a few lines of BASIC I'm now saving frequency readings to a file. I've already found that my home-built adapter won't work with Timelab. It has served it's purpose. With less than $10 worth of parts I was able to verify that the GPIB on my HP5335A is working. Now it's time to get a better GPIB adapter. I like the Prologix because It's easy for me to write code to access a com port. Do you know if the 82357 type adapters can be accessed as a com port? Any suggestions on what adapter to use? _______________________________________________ time-nuts mailing list -- [email protected] To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.