Anders,

On 04/03/2015 11:49 AM, Anders Wallin wrote:
Hi all,

I ran some tests on a new 53230A counter that just arrived:
http://www.anderswallin.net/2015/04/keysight-53230a-noise-floor-test/

The time-interval single-shot noise and resulting ADEV seems easy enough to
explain.

I didn't yet have time to look closely at Enrico Rubiola's notes over here:
http://rubiola.org/pdf-slides/2012T-IFCS-Counters.pdf
There is probably a good explanation for the ADEV-level in standard
(pi-counting?) reciprocal frequency counter mode as well as the roughly
1/sqrt(10) enhancement in ADEV when increasing the gate-time 10-fold.

When using new counters with Delta (HP53132A) or Omega (Pendelums Least-Square) estimators for refined frequency resolution, they work by doing a filtered weighing which has a lower bandwidth than the brute force counter, and this will give a lower reading for white and flicker phase noise (see their formulas), and this causes the deception that the ADEV is lower, just because you get the reading, that the noise-type is lower, but all which is happening is that you lower the reading for that noise-type, not that you have a better noise situation.

on another note:
The discussion on phase-noise <-> ADEV calculations earlier was
interesting. Datasets (synthetic or real) with known correct ADEVs and
phase-noise-spectra would be a good addition to allantools and a motivation
to add code for the phase-noise <-> ADEV conversion.

Do look at the NIST SP 1065 for a noise-generator with known properties for a number of algorithms.

Cheers,
Magnus
_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to