Hi

Alll of the GPSDO’s that NIST is looking at are 1 PPS based. What’s different 
about them:

1) The OCXO’s (or Rb’s) are stable to 0,001 ppb at 1 second
2) The GPS is designed for timing (and spec’d by an outfit that understands 
timing — see below) (uBlox maybe)
3) The GPS puts out a sawtooth correction and the loop uses it.
4) The GPS PPS and OCXO are compared at the sub nanosecond level to feed the 
loop
5) The loop parameters are optimized such that the GPS does not feed though 
much at < 1,000 seconds
6) The OCXO is stable enough that it can chug along without help from the GPS 
for hundreds of seconds 

That’s the short list, there is probably a much more detailed list, but that 
gets into a lot of “that depends” 
sort of stuff. With $25 eBay “treasures” popping up from time to time the cost 
target is very fluid on these 
things.

One thing to take a look at: NIST has papers on L1 ionosphere correction (= 
lack of). Note from the papers how it must 
impact your module. After you dig a bit into that 10 to 50 ns per day variable, 
take a look again at the 
10 ns manufacturer’s accuracy claim in your module...

Bob

> On Aug 16, 2015, at 2:47 PM, Nick Sayer via time-nuts <[email protected]> 
> wrote:
> 
> I’ve designed and make and sell a GPSDO on Tindie 
> (https://hackaday.io/project/6872-gps-disciplined-tcxo). It’s brand new - 
> I’ve sold a handful of them so far. So as to make this post not *entirely* 
> self-serving, what I would like is some further guidance on how I can better 
> characterize its performance.
> 
> The GPS reference is a 1 pps signal (It’s the Adafruit Ultimate GPS module - 
> a PA6H). The manufacturer claims an accuracy of ±10 ns, but that's accuracy 
> relative to the true start of the GPS second. They don’t make any claim for 
> stability.
> 
> The oscillator itself (Connor Winfield DOT050V 10 MHz) has a short-term 
> (though they don’t say how short that term is) stability of 1 ppb. The 
> absolute accuracy of it is (I assume) irrelevant, because it’s a VCTCXO and 
> the control voltage is steered by GPS feedback.
> 
> The feedback loop takes samples over a 100 second period. That gives me an 
> error sample with a granularity of 1 ppb. I keep a rolling sample window of 
> 10 samples to get an error count over 1000 seconds. I've kept track of both 
> of these values for extended periods (days) as well as logging the DAC value 
> (the number that's proportional to the control voltage). The 1000 second 
> sample window error averages zero, and it almost never exceeds ±7 (every once 
> in a while if I physically move it, it will show a momentary error glitch, 
> but that shows up in the short term feedback sampling too). The 100 second 
> samples are almost all 0 or ±1, with an occasional ±2 showing up. As I said 
> before, if I bonk the oscillator, it may briefly show a ±6 or so for one 
> sample.
> 
> If I pit two of them against each other on a scope and take a time lapse 
> video (http://www.youtube.com/watch?v=9HkeCI90i44), you can see that they 
> stay mostly locked with occasional periods of drift. I sort of assume that 
> that represents periods where the two GPS receivers disagree as they decide 
> differently how to select among the available satellites.
> 
> I've been saying out loud that the oscillator is ±1 ppb from GPS over the 
> 1000 second window. I know of Allan variance, but I don't have anything else 
> handy I can use for comparison. I also can't really afford to send one off 
> for testing to a proper lab. In looking at 
> http://tf.nist.gov/general/pdf/2297.pdf, it suggests that my results are 
> relatively poor compared to what a GPSDO can achieve (more like 10^-12 rather 
> than 10^-9), but I assume that they’re able to use a higher frequency GPS 
> reference than just 1 PPS (and they’re a lot pricier).
> 
> What else can I do to try and characterize the performance? If mine is 
> performing far more poorly than the same price ($175) can buy elsewhere, then 
> what am I doing wrong?
> 
> _______________________________________________
> time-nuts mailing list -- [email protected]
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.

_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to