I wrote:

The LT1720** works on a 3.3v single supply, and has internal hysteresis. Just ground one input, ground the other through a resistor, and attach your AC-coupled signal (best if you can use equal resistors to ground on the two inputs, but if the signal termination is 50 ohms the other input can be 0 ohms to ground; even 1k ohm would probably be OK).

Note that the LT1720 is specified for inputs to 100mV below V- (ground, in this case) and your 200mVpp signal pushes it right to the limit. A bit of attenuation is probably a good idea.

** The 1720 is a dual comparator, so you'd have an extra to use or ignore. Alternatively, the 1719 is a single with extra features that you could ignore. Either should work well in your application.


See attached schematic.

Best regards,

Charles