For software I'm using PN3048 from k35fx's site together with an Aglient 82347b 
GPIB interface running thru Virtual Box. Other than not recognizing the 
instruments once and a while, its been pretty reliable. I've also gotten the 
original BASIC program to work under HT BASIC and surprisingly is more reliable 
in terms of communicating with the devices.
I  ran the test at 10mhz against the builtin oscillator that comes with the 
hp11848a. I also ran a test against a Fluke 6070a at 100mhz and got almost the 
exact same results from 1khz on out. Maybe I got lucky with my 8660c. It would 
be interesting to see measurements from a few other units.
David

> From: [email protected]
> To: [email protected]
> Date: Sun, 17 Jan 2016 04:21:40 +0000
> Subject: Re: [time-nuts] Phase Noise Comparisons
> 
> Hi Dave,
> 
> Very good.  Interesting you data is better than any typical data I could find 
> on the web which was -105 to -107 dBc/Hz at 10 kHz. 
> 
> I need N0QO to get his 3048A up and running again so we can test some other 
> equipment.
> 
> What are you using for software?
> At what frequency was this plot made?
> 
> I tried to send a moderate size PDF, but it at least temporarily has been 
> blocked. 
> 
> Rob
> NC0B
> 
> -----Original Message-----
> From: time-nuts [mailto:[email protected]] On Behalf Of Bob Bob
> Sent: Saturday, January 16, 2016 8:37 PM
> To: Discussion of precise time and frequency measurement
> Subject: Re: [time-nuts] Phase Noise Comparisons
> 
> 
> 
> 
> 
> 
> 
> 
> 
> Actually, I just acquired a HP3048a phase measurement system and I'm 
> measuring all of the my signal generators, one of which is an 8660c with the 
> 86603 plugin.  Attached (ignore the caption) is the result compared against 
> the 3048a internal 10mhz low noise reference. 
> 
> 
> David
> > From: [email protected]
> > To: [email protected]
> > Date: Sat, 16 Jan 2016 20:05:18 +0000
> > Subject: [time-nuts] Phase Noise Comparisons
> > 
> > I did some interpolation to compare the specs for phase noise of the 8660 
> > the way it was done then to the present method. I think it is going to be 
> > something like -105 dBc/Hz to -110 dBc/Hz at 10 kHz offset at 14 MHz 
> > carrier frequency.
> > 
> > 
> > 
> > As a comparison, the 8662A is about -135 dBc/Hz and the 8642A is better 
> > than -150 dBc/Hz.
> > 
> > Actual measurements on multiple generators in my lab.  Unfortunately I have 
> > never used or measured an 8660.
> > 
> > 
> > 
> > 3335A -128 dBc/Hz
> > 
> > 3336C -120 dBc/Hz
> > 
> > 3325A -115 dBc/Hz (original version.  Later improved by 4 pr 5 dB 
> > after the 3336C came out.)
> > 
> > 
> > 
> > As a comparison a Rigol DG4162 is -115 dBc/Hz All at 10 kHz offset and on 
> > the 20 meter band.
> > 
> > 
> > 
> > It depends what you are trying to measure. Sensitivity or noise floor, 
> > anything will work within the range of specified output accuracy and 
> > leakage.  On the other hand, if you are trying to measure the dynamic range 
> > of a receiver, or its phase noise characteristics, you cannot do that with 
> > most generators.
> > 
> > 
> > 
> > Rob, NC0B
> > 
> > 
> > 
> > 
> > 
> > 
> > 
> > _______________________________________________
> > time-nuts mailing list -- [email protected] To unsubscribe, go to 
> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> > and follow the instructions there.
>                                         
> 
> 
> --
> If this email is spam, report it to
> https://support.onlymyemail.com/view/report_spam/ODExMjI6MTg0MzU3MjgwMDpyb2JAbmMwYi5jb206ZGVsaXZlcmVk
> _______________________________________________
> time-nuts mailing list -- [email protected]
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
                                          
_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to