John,

On 01/17/2016 10:49 PM, John Miles wrote:
Therefore, talk-only mode is a big advantage in terms of decoupling
on RS-232 and makes almost no difference on GPIB.

That's not the case when it comes to counters.  By timing issues, I wasn't 
referring to layer-1 handshaking, but rather the interplay between the GPIB 
software application, the network or bus connectivity between the app and GPIB 
controller, the controller itself and its firmware, and an addressable counter 
that returns each measurement only in response to a command from the app.  The 
difference in performance between a talk-only connection and a two-way 
conversation can be substantial.  Yes, everything runs in lockstep, but the sum 
of the delays and latencies in each of those stages can easily exceed 0.1 
second.  In Attila's case there are also VM crossings in both directions to 
worry about.

The counter, unlike any of the other participants in the conversation, is working on a 
deadline.  Everything in a counter tends to happen in the "foreground," so to 
speak.  If the counter takes too long to interpret and execute a GPIB command, it may 
fail to rearm itself in time for the next trigger.  That's never a problem in talk-only 
mode, at least not at rates we're concerned with here, because the counter never has to 
do anything but send out data.

It was time that we started to talk about this total picture.

Add that some GPIB interfaces uses USB-to-serial chips and then serial-CPU-GPIB. Many instruments uses a GPIB chip having the low-level logic to off-load the CPU. The SR620 for instance uses the TMS9914 GPIB controller chip, which does much of the hand-shaking, but does not offload the CPU much in regards to FIFO-buffering, which is typical of its age. Still, much of the lower level stuff gets the boost it needs. The SR620 has maybe not the powerhouse processor, so that should be a limitation by itself. Interestingly the SR620 uses a Z80 as a IO processor to draw on XY scopes.

There is more handshake in the GPIB than pure flow-control, but I've not seen it used and honestly I haven't tried it myself. It seems like the support is there and documented for many instruments. Shouldn't we try to use that to aid the ping-pong between instruments?

Building up the system, we should think about how the system maintain its properties, data flow, arm-trigger mechanisms.

Cheers,
Magnus