Hi Attila,

I quite agree, I should have remembered that.... but I tend to use fairly small 
coax anyway at those frequencies as a matter of course and it didn't register 
as something that needed saying. My fault!
It can be quite difficult to get people to notice the difference between SMA 
and 3.5mm - I have often had to repeatedly remind some of our test technicians 
of the difference as they tend to think they are interchangeable with the 
unfortunate results you describe :-(    It is just as well that the 3.5mm cal 
kit for the HP8510 (still a fine VNA, if lacking a few of the new bells and 
whistles) is a work of art with DTIs for checking the mechanical setup of the 
connectors - they are not just for show!

Regards,

Paul Reeves

-----Original Message-----
From: time-nuts [mailto:[email protected]] On Behalf Of Attila Kinali
Sent: 23 January 2017 19:33
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Thermal effects on cables

On Mon, 23 Jan 2017 15:13:39 +0000
REEVES Paul <[email protected]> wrote:

> Surely the impedance of the cable is only affected by the ratio of the 
> inner conductor and outer conductor diameters modified by the internal 
> dielectric constant, nothing to do with the frequency of operation. 
> You might well have problems converting the larger diameters down to a 
> suitable size for the connectors at the higher frequencies though....

If you are only looking at the impedance, then yes. But once you get to high 
frequencies, you get also multi-mode behaviour of the coax cables and 
connectors, which leads to dispersion. That's why people hardly use N 
connectors for GHz frequencies, even though the connector itself would allow 
it. For VNAs where even small phase shifts/instabilities due to multi-mode 
behaviour/dispersion are a no go, the connectors are usually 3.5mm (basically a 
precise version of the SMA), 2.9mm, 2,4mm etc and go down to even 1.0mm which 
can be spec'ed up to 110GHz.

BTW: the 3.5mm connector is one that you will find on many instruments that go 
beyond 1-2GHz. Unfortunately it looks exactly the same as an SMA connector and 
will mate with one. Even more unfortunately, mating it with an SMA connector 
will scar the connector and most likely move it out of spec (ie degrade it to a 
simple SMA connector). If it's just an adapter, you've only lost a bit of money 
(in the order of a few €100).
If it was the connector of your VNA/oscilloscope/..., you might need to send it 
in for an expensive repair.

                        Attila Kinali

--
You know, the very powerful and the very stupid have one thing in common.
They don't alters their views to fit the facts, they alter the facts to fit the 
views, which can be uncomfortable if you happen to be one of the facts that 
needs altering.  -- The Doctor _______________________________________________
time-nuts mailing list -- [email protected] To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to